Hi everyone.
My company is developing a medical device and we are looking to comply to IEC 60601-1. We've seen a lot of mention of an "isolation diagram" or insulation diagram being required as a first step but we can not find a reference to it in the standard, could anyone give me the clause which requires or can be interpreted to require a isolation diagram?
Thanks in advance!
Benjamin Weber
Hi everyone.
Thanks in advance!
You're right, 60601 does not require an insulation diagram explicitely. But since a lot of test reports use the TRF (test report form) from IECEE this a a common thing to start with.
There are a lot of clauses relating to the electrical safety, most obvious section 8, and there in section 8.5 you shall implement means of protection (MOP) against electrical shock. See also cl. 3.43 for the definition of "insulation co-ordination".
The insulation diagram is is very helpful tool for the manufacturer and the tester, to identifiy, where such MOP are implemented in the device and which requirements apply to each single MOP.
So don't just look for an explicite reference in the standard, but also look for your benefit during development and testing of your device. Wiothout an proper insulation concept and diagram, it is hard for you to fulfill all requirements.
W Isolation requirements for Wireless ECG device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Isolation requirements for USB/Ethernet IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC60601-1: Isolation for an applied part with working voltage of 1500Vp. IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
W Body-worn brain stimulator (MOPP isolation requirements) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
N Triple insulated wire for transformer isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
J Medical Device Electronics - Secondary to output isolation for 1 MOPP IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
A MOPs needed after Isolation Transformer IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F Ambulance Battery Isolation - IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Isolation of Internally Powered (3.3 Vdc) ME Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R How to Determine Working Voltage & Tolerance for Isolation Barriers? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Suitable Protocol for Isolation of Clostridium perfringens Food Safety - ISO 22000, HACCP (21 CFR 120) 2
A IEC 60601-1 SIP/SIO Isolation and Touch current requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
F Additional Isolation Required? Doppler Probe Desktop Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
S Class II PSU (200W) BF Power Supply Configuration Isolation Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C Medical device worn on body with USB, isolation, 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
B IEC 60601-1 2nd Clause 17 and 3rd edition - Electrical Isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Ethernet Isolation Transformer for Medical Grade PCB IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
S High Voltage inside Ultrasonic Probes - Do SIP/SOP ports require isolation? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A USB Port Isolation as per IEC60601-1 - Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R USB Port (Signal I/O) Isolation as per IEC60601-1 - Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 40
D UL Certified Isolation Transformer Various Other Specifications, Standards, and related Requirements 3
apestate Vibration, Measurement and Isolation - Good ways to isolate M&TE from vibration? General Measurement Device and Calibration Topics 9
G What differentiates a "bridging transformer" from an ordinary "isolation transformer" General Measurement Device and Calibration Topics 5
K Doubt on insulation class of a device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W Insulation requirements of BF type infant skin temperature probe IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M 2xMOPP insulation for Applied Part B. IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
V PCB internal insulation treated as insulation formed by insulating compound IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W HF surgery device accessory insulation requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
E Creepage distance accross insulation of step up transformer EU Medical Device Regulations 2
S Who compiles the Critical Component List, Insulation Diagram etc.? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
P Question on Table 6 "Test voltages for solid insulation forming a MEANS OF PROTECTION IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
P MOOP and insulation coordination IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K IEC60601 3rd Edition Insulation Thickness for electrosurgical devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N Cable Connectors - Insulation does not go all the way into the back of the connector Manufacturing and Related Processes 4
A Insulation between Secondary Accessible Parts and Earth IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
P IEC 60601-1-11 - Conversion to Double Insulation IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Is insulation between F type applied part and other parts a must? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
N Insulation Diagram for Water Tub IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W Is anybody familiar with IEC61010-1 Insulation Requirements? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
V External power supply with reinforced insulation as a simple secondary circuit? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W Insulation requirements between F-type applied part and enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
P Understanding IEC 60601-1:2005, Clause 8.1 b) - Short Circuiting Double Insulation IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
R Determining Working Voltage - Drawing a Medical Device Insulation Diagram IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
W Insulation Grade for In Vitro Diagnostic (IVD) Device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Do insulation layers in double insulation have sequence requirement? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S Can enclosure of semiconductor or electrolytic capacitor provides basic insulation? IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
N Insulation Barrier Testing - According IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 22
T Insulation diagram for a medical electrical equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 4