Isolation of Internally Powered (3.3 Vdc) ME Equipment

R

rob.lim

#1
Hi. All.
I have a few questions.
The stimulator is used Internally powered (3.3 Vd.c), and Type BF.
I think that between circuit in internally powered (secondary circuit) and patient connection must be satisfied 1MOPP.
So, I think that apply the table 15 according to subclause 8.9.1.12.
But, there is only MOOP in table 15, not MOPP.
In such a case, do I need to apply the table12? Or would not need insulation to patient connection between circuit in internally powered (secondary circuit)?
Thank you.
 
Elsmar Forum Sponsor
Thread starter Similar threads Forum Replies Date
R IEC 60601-1 - Different methods of achievement of the isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
W Isolation requirements for Wireless ECG device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Isolation requirements for USB/Ethernet IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Q Isolation/Insulation Diagrams - IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
T IEC60601-1: Isolation for an applied part with working voltage of 1500Vp. IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
W Body-worn brain stimulator (MOPP isolation requirements) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
N Triple insulated wire for transformer isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
J Medical Device Electronics - Secondary to output isolation for 1 MOPP IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
A MOPs needed after Isolation Transformer IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F Ambulance Battery Isolation - IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R How to Determine Working Voltage & Tolerance for Isolation Barriers? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Suitable Protocol for Isolation of Clostridium perfringens Food Safety - ISO 22000, HACCP (21 CFR 120) 2
A IEC 60601-1 SIP/SIO Isolation and Touch current requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
F Additional Isolation Required? Doppler Probe Desktop Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
S Class II PSU (200W) BF Power Supply Configuration Isolation Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C Medical device worn on body with USB, isolation, 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
B IEC 60601-1 2nd Clause 17 and 3rd edition - Electrical Isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Ethernet Isolation Transformer for Medical Grade PCB IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
S High Voltage inside Ultrasonic Probes - Do SIP/SOP ports require isolation? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A USB Port Isolation as per IEC60601-1 - Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R USB Port (Signal I/O) Isolation as per IEC60601-1 - Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 40
D UL Certified Isolation Transformer Various Other Specifications, Standards, and related Requirements 3
apestate Vibration, Measurement and Isolation - Good ways to isolate M&TE from vibration? General Measurement Device and Calibration Topics 9
G What differentiates a "bridging transformer" from an ordinary "isolation transformer" General Measurement Device and Calibration Topics 5
C Can You Internally Audit a Process You Own? Internal Auditing 25
C Method of Monitoring our Internally Produced Non-Conforming Products Quality Tools, Improvement and Analysis 7
T Are internally found Medical Device Software "Bugs" Complaints? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
D Sharing ISO Standards internally on company computers ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
S Internally powered ME with non-conductive applied parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Single Fault Condition for Internally Powered Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S Should Client Controlled Procedures be Controlled Internally by us as well? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
R Referencing and Controlling Labels & Stickers used internally throughout Production Document Control Systems, Procedures, Forms and Templates 1
H How do we use AS9101D internally? AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 2
A Internally Powered Medical Device and Dielectric Strength IEC 60601 - Medical Electrical Equipment Safety Standards Series 17
D Means Of Protection of the Video OUT of an Internally powered device IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M All Internally Generated Documents ? Required to be Controlled or Not All? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 19
C How will I Internally Audit my BPO Center for ISO 9001:2000 Internal Auditing 8
J What PPM (Parts Per Million) targets do you use internally? Quality Tools, Improvement and Analysis 9
B Calibration of "Z" class ring gauges internally using a CMM General Measurement Device and Calibration Topics 5
B Can I calibrate Optical Comparators internally? General Measurement Device and Calibration Topics 22
K Requirements for 2nd languages? I can't even audit them internally IATF 16949 - Automotive Quality Systems Standard 59
A Gauges on pressure tank - What is the best way to calibrate a psi gauge internally? General Measurement Device and Calibration Topics 6
L Posting portion of standards internally? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
B Why internally audit ISO clauses for ISO 9001:2000? (Small company) General Auditing Discussions 16
E Wall Internal Auditor Training - Done internally Training - Internal, External, Online and Distance Learning 7
J FDA wants electrical safety testing on battery powered medical device US Food and Drug Administration (FDA) 11
J Powered wheelchairs (Class IIa) in the process of CE marking and use in hospitals CE Marking (Conformité Européene) / CB Scheme 4
R ISO 10079-1:2014 on electric powered nasal aspirator Other US Medical Device Regulations 0
L DC Powered Medical Device Fuse / 15.4.3.5 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
T Battery Powered Device - MOP and Patient Auxiliary Currents IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Similar threads


















































Top Bottom