Hi Beloved Engineers,

I have a question regarding IEC60601-1 requirement for Wireless ECG device.
-> This device will be internally powered. (Li-Po rechargeable battery)
-> Device will be connected to PC wirelessly (Bluetooth)
-> Also recharged Wirelessly (qi 2.0).
-> Defibrillator protected with transient suppression diodes

Which clearly means device will never be connected to mains supply in its entire life cycle..
So do I really need MOOP (Means of operator protection) and MOPP (Means of patient protection)
rough block diagram is attached.

Thanks in advance.

Winston
 

Attachments

Elsmar Forum Sponsor

Benjamin Weber

Trusted Information Resource
#2
You maybe also need means of protection (operator and patient) for secondary circuits. For operator protection I see a good chance, that the voltage and stored energy in the battery will be below the limits of cl. 8.4.2 c). But for patient protection there is no scuh thing as "safety extra voltage" at all in 60601. You have to make sure, that the leakage and patient auxiliary current limits are not exceeded. And therefor MOPP should be implemented.
 
#3
Thanks Benjamin,

Can you please look at attached block diagram and tell me if that isolation is enough for MOPP ?
Or do I need isolation between defib protection and ADC input ? Apologies if this question is noob.

Thanks,

Winston
 

Attachments

#4
Hi Winston,
The only isolation barrier I see is between the battery terminals and the enclosure (2MOOP for 3.3Vdc), that should be ensured by the plastic enclosure.
However, you have to check that the auxiliary leakage currents between the different connections to the patient, and also between the patient connections and the other parts (enclosure) remain below the limits in both normal and SF conditions. Normally this should be the case thanks to the high input impedance of the ECG amplifier. But still you have to check it by putting in fault the relevant components and barriers.

Regards

Souheil
 
Thread starter Similar threads Forum Replies Date
F USB powered handheld medical device - Isolation requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E Isolation requirements for USB/Ethernet IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W Body-worn brain stimulator (MOPP isolation requirements) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A IEC 60601-1 SIP/SIO Isolation and Touch current requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S Class II PSU (200W) BF Power Supply Configuration Isolation Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Isolation of secondary grounded circuits IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R IEC 60601-1 - Different methods of achievement of the isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Q Isolation/Insulation Diagrams - IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
T IEC60601-1: Isolation for an applied part with working voltage of 1500Vp. IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
N Triple insulated wire for transformer isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
J Medical Device Electronics - Secondary to output isolation for 1 MOPP IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
A MOPs needed after Isolation Transformer IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F Ambulance Battery Isolation - IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Isolation of Internally Powered (3.3 Vdc) ME Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R How to Determine Working Voltage & Tolerance for Isolation Barriers? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Suitable Protocol for Isolation of Clostridium perfringens Food Safety - ISO 22000, HACCP (21 CFR 120) 2
F Additional Isolation Required? Doppler Probe Desktop Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
C Medical device worn on body with USB, isolation, 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
B IEC 60601-1 2nd Clause 17 and 3rd edition - Electrical Isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Ethernet Isolation Transformer for Medical Grade PCB IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
S High Voltage inside Ultrasonic Probes - Do SIP/SOP ports require isolation? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A USB Port Isolation as per IEC60601-1 - Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R USB Port (Signal I/O) Isolation as per IEC60601-1 - Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 40
D UL Certified Isolation Transformer Various Other Specifications, Standards, and related Requirements 3
apestate Vibration, Measurement and Isolation - Good ways to isolate M&TE from vibration? General Measurement Device and Calibration Topics 9
G What differentiates a "bridging transformer" from an ordinary "isolation transformer" General Measurement Device and Calibration Topics 5
P UDI-PI requirements on reusable surgical device, do we need serialisation? ISO 13485:2016 - Medical Device Quality Management Systems 0
Sidney Vianna Interesting Discussion Should ISO 9004 be changed from a guidance standard to a requirements standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
J WAIVED ON Q1 - We Don't have to comply with FORDS customer specific requirements IATF 16949 - Automotive Quality Systems Standard 2
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
Techilady Ford Special Characteristics Communication and Agreement Form (SCCAF) provides production requirements? FMEA and Control Plans 4
P Requirements for being an European Representative EU Medical Device Regulations 3
C ISO 14001:2015 6.1.3 Compliance Obligations - Legal requirements monitoring ISO 14001:2015 Specific Discussions 0
lanley liao How to understand correctly the requirements of API monogram products Oil and Gas Industry Standards and Regulations 0
A GPS Regulatory requirements? Manufacturing and Related Processes 6
J SSQR-01 Vision Requirements- Color Blindness AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
P ISO 14644 Class 8 Cleanroom Air Filter Requirements Other Medical Device Related Standards 4
J Requirements as a Distributor for Incoming Inspection of Purchased Finished Medical Device Medical Device Related Regulations 0
R Weld Penetration Standards and Requirements Manufacturing and Related Processes 2
S Registration of Medical Device in Hong Kong - labeling requirements Other Medical Device Regulations World-Wide 0
L Requirements for Cables used for Low-Voltage DC Mains (12-36VDC) IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M AS9100D Registrar pre-audit requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 15
S Determining the requirements for the products and services ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
D LiPo battery minimum requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
V MDSAP Requirements for Device Designer Canada Medical Device Regulations 1
J Biocompatbility testing on Class 1 device requirements Other US Medical Device Regulations 12
Le Chiffre Online training available for ISO/IEC 17021-1: Requirements for bodies providing audit and certification of management systems Training - Internal, External, Online and Distance Learning 3
S What are the International Forging Requirements for Nitrogen Strengthened Stainless Steel? EU Medical Device Regulations 3
M Supplier requirements - Major supplier is a Non-Profit registered with ICCBBA (FDA UDI) Supply Chain Security Management Systems 12

Similar threads

Top Bottom