Issue with CAPA system - Should Over Due CAs warrant a CAPA?
Maira O.
Hello,:bigwave:
So we have several CAPA's that are severely past due, and we want to assign a CAPA to address the current problem.
While some people in my company agree that this is good practice to show the FDA that we noticed a gap in our system and we are addressing it, others seem to think that if assigned, that CAPA might cause more harm rather than help us during an FDA audit.
What do you guys think? Should a CAPA be warranted? Or will that just expose us during an FDA audit?
Any comments or suggestions will be greatly appreciated and helpful!
:thanks:
Coury Ferguson
Moderator here to help
Re: Issue with CAPA system - Should it warrant a CAPA?
Maira,
My opinion, and I am not an expert in Medical Devices.
CAPA is a major part of the Quality Management System, that is in place. It is one of the greatest tools you can use. But, it must be a thorough CAPA program.
Since your company is under Regulatory Oversight (FDA), in my opinion, if I was the FDA and was reviewing the CAPA program in your company, I would ask questions. Questions like:
1. Delinquent CAPAs, why?
2. Why haven't these CAPAs been elevated to the next level in your company or the supplier for appropriate action?
3. The delinquent CAPAs are a reflection of the company's failure in their CAPA program. Now, what actions are you going to take internally, to eliminate the recurrence?
4. Since, the CAPA program has shown evidence, that the system isn't working, I will write a Major Non-conformance for a complete breakdown of the CAPA program.
Do you feel you would be able to answer these questions and explain the reason why the System failed?
So yes, the CAPA program is very important to a healthy Quality System.
Again, just my opinion on this.
JoShmo
What's the subject of these CAPAs? (how I hate that term) How were they found?
Many I've seen are written in a way that anyone could tell what needed fixed. Then, the system gets burden-down with so much noise no-one knows which to go after. Classic. Before you worry about what to do, I'm with those who say they can cause more harm than good. Don't foget it's the same as poainting yourself into a corner.
TWA - not the airline
I agree with Coury that a finding of "CAPA system not working" would be a very serious issue in any inspection. So you'll need to address this very carefully. If your CAPA system has a monitoring of open CAPAs and that has been working and now alerted you to this situation then you could raise a CAPA and after investigation of the issue address the root cause with a CA. In order to restore compliance you could also install a task force with increased resources to timely close out the open CAPAs as a kind of correction. This would show that even though something happened on the way your CAPA system actually is working...
If it is just you who noticed this and there is no formal, documented monitoring of open CAPAs then maybe you could start to address this out of the management review/responsibility or internal audit system (which may or may not feed this back into the CAPA system...). Those systems are not that in-depth visible during an inspection which may be an advantage but also raise more suspicion about your system.
The best way to proceed also depends on what you perceive as the root cause right now. If it is priorities and/or resources then you'll never succeed without the support of management...So maybe you do not only have to consider FDA but also politics...
Hope this helps...
Best,
TWA
Candi1024
A CAPA to fix CAPAs. That just sounds scary in and of itself.
I don't love the way CAPAs have progressed in the system I work in. The preventative action tends to have too wide of a scope, to the point that we are digging into way too many details to verify that problems don't exist elsewhere. But being that the CAPA is usually raised because of an audit finding, and not actual quality issues, we are wasting time on non-value added activities.
Does it have to be a CAPA in the formal sense? I mean if a tool is broken, you wouldn't use it to fix itself.
Candi, I'm with you and I totally understand what you mean. The CAPA system has become the sacred cow of quality management, sometimes totally blown out of proportion. In my experience this is what happens when quality is not supported by management; if it is, things tend to get addressed much earlier...
RoxaneB
Change Agent and Data Storyteller
Personally, I am not a fan of generating a finding because CAPAs are "overdue". These "overdue" CAPAs are not the issue...they are a symptom of something else such as an inefficient and/or ineffective CAPA process.
The first question to be asked, in my opinion, is WHY are they taking so long to close? Some possibilities include (but are not limited to):
  • Resources - Are there simply just not enough resources (be it time, money, people, etc.) to properly process, resolve, action and close the CAPAs? Some CAPAs require significant investment (again...time, money, people) and could stay open for a while. If you're waiting for technology to be developed, this could take some time.
  • Quantity - If you issue a full-blown CAPA for every issue, it will bog the system down. Develop triggers (based on an analysis of solid data) that will help people determine what is a correction versus a corrective action.
  • Failure to adequate resolve the issue - Are they still open because issues keep occurring (i.e., the root cause has not been properly identified/actioned)?
  • Management commitment - If leadership isn't committed to reviewing, actioning, supporting and resolving CAPAs, why should anyone else in the organization care?
Tyler C
Hello,:bigwave:
So we have several CAPA's that are severely past due, and we want to assign a CAPA to address the current systemic problem.
:thanks:
I have a few questions. Are there any time frames around your CAPA system, either in your procedure or your key performance indicators?
How is your CAPA set up? Is it for any and every product nonconformity, or process nonconformities?
For example, we have a KPI for our CAPA system, and that is measuring any validation issues. Basically, we can't have a CAPA in our system longer than 3 months unless it is specifically for continual improvement (preventive action). Since this is a KPI, if we do have any longer than 3 months, we have to initiate a CAPA to determine why they have not been resolved. This helps us determine if there is a lack of follow through, a lack of commitment to quality, or if we simply have not determined the root cause yet.
There is not always a corrective action to be taken, such as if we have worked diligently on the root cause and just have not been able to determine it yet.
I do see the value in initiating a CAPA for severely overdue CAPAs. I also understand not wanting to bring attention to it or wanting to 'paint yourself into a corner', but on the other hand, what value is the QMS if it is not effective? I do feel an auditor would understand if you were working on fixing your system as that is a part of continual improvement, but the auditor may not like the fact that you are trying to hide a broken system.
Ninja
Looking for Reality
As was mentioned briefly above...what you call the effort of fixing the problem is likely far less important than finding the root cause and fixing it...quickly.
Generating a CA for the CA system not working...then dumping it straight into the non-working system seems a little illogical. Forget about using the broken system, and just fix it by any means necessary.
Worry about terminology after the system is fixed.
"We can't solve problems by using the same kind of thinking we used when we created them."
"Insanity: doing the same thing over and over again and expecting different results."
- Both attributed to Albert Einstein -
Candi1024
Just for perspective, our goal for CAPAs is under one year.
There are many reasons they go over. The flow of information from the CAPA team to the CAPA engineer to the engineer working on the project is one issue.
Another issue is that the scope is just too broad. This is an exaggeration, but if we found a misspelling in one AI, the scope would include checking all 1400 AI's for spelling issues. And that would be pretty easy compared to most of our scopes.
We also routinely fail effectiveness checks.
If I was working on this, I would do an A3. Unfortunately (or fortunately) I am not a CAPA engineer :)
P When is an issue a CAPA (Corrective and Preventive Action) request? Nonconformance and Corrective Action 6
P Issue Tracking and CAPA (Corrective and Preventive Action) Management Nonconformance and Corrective Action 6
P When to issue a CAPA (corrective action preventive action)? US Food and Drug Administration (FDA) 11
M Issue a CAPA if there is a corrective action associated with a complaint? Preventive Action and Continuous Improvement 7
K Best practices on the criteria to issue CAPA & expected response time? Preventive Action and Continuous Improvement 2
I SaMD Software bug and issue tracking. Manufacturing and Related Processes 1
S Supply Issue Supplier Quality Assurance and other Supplier Issues 4
bzachary1968 Issue Log / Tracker with dashboard. Document Control Systems, Procedures, Forms and Templates 2
S PPAP- (When to issue ECN,ECR & PCN ,PCR) APQP and PPAP 2
S ISO9001 Major for repeat issue. ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 24
O Issue with free sales certificate for implants from China China Medical Device Regulations 1
L Unique Supplier Control Issue AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
S Reliability issue - Frequent component failure of the product Reliability Analysis - Predictions, Testing and Standards 2
S Issue No. & Revision No. Document Control Systems, Procedures, Forms and Templates 4
Ch00Ch00 Evaluate a process - Issue number Lean in Manufacturing and Service Industries 7
M MSA angle issue - Nylon parts using Aberlink 3D Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
P MedWatch Report from a hospital due to a quality issue US Food and Drug Administration (FDA) 5
Marc Media Gallery Issue? Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 1
R Problem solving activity - Three hours to fix the issue Manufacturing and Related Processes 15
I Reissuing documents due to missing version / issue numbers Document Control Systems, Procedures, Forms and Templates 3
Marc Fuel tank issue (FOD) hits Boeing’s 737 MAX - Feb 2020 World News 19
F Struggling with a root cause analysis - Customer Returns - Escape issue Problem Solving, Root Cause Fault and Failure Analysis 15
Ron Rompen 2D matrix issue - Parts (machined steel) returned from the customer Design and Development of Products and Processes 0
G PolyWorks Alignment Issue - How to align the part General Measurement Device and Calibration Topics 1
M Accredited Calibration Sevice Provider using computerized system to issue calibration certificate Qualification and Validation (including 21 CFR Part 11) 3
M Accredited Calibration Sevice Provider using computerized system to issue calibration certificate General Measurement Device and Calibration Topics 2
A External Auditor issue with Internal Audits Internal Auditing 7
V Every good documentation practice observation is an data integrity issue US Food and Drug Administration (FDA) 7
V What is the criteria to cite an good documentation practices observation as an data integrity related issue US Food and Drug Administration (FDA) 6
A Domain of the component (EASA CM No.: CM-SWCEH-001 Issue 01 Revision 02) Point 9.3.3 (Usage domain aspects) EASA and JAA Aviation Standards and Requirements 0
B How to reply NCR on ineffectiveness of corrective action during IATF external audit? This is repeated issue whereby some mistake was done. IATF 16949 - Automotive Quality Systems Standard 7
J Control of nonconforming product issue (AS91000) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
C TL-9000 Certifying Body Issue - Auditor failed to find an issue for 10 years TL 9000 Telecommunications Standard and QuEST 16
E Determining what is good and what is bad can be subjective - when is it a quality issue? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
Ed Panek Sensitive Supplier Issue - Single source supplier ISO 13485:2016 - Medical Device Quality Management Systems 6
B NB won't issue CE Certificate or return phone calls CE Marking (Conformité Européene) / CB Scheme 3
A Drawing Issue with Supplier - xb file did not match that of the pdf drawing Nonconformance and Corrective Action 4
J MAG 6 and EASA Part-145 Issue 005. EASA and JAA Aviation Standards and Requirements 0
A ISO/IEC 27001 - Issue during implementation of system IEC 27001 - Information Security Management Systems (ISMS) 3
M Saturated salts issue - Humidity Chamber General Measurement Device and Calibration Topics 5
N Do I need to approach galvanized steel sheet flatness issue with DOE? Problem Solving, Root Cause Fault and Failure Analysis 9
Sidney Vianna ASQ apparently doesn't know ISO does not issue ISO 9001 certificates? ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 4
D Can an AS9100 auditor issue an NCR a month after the audit? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
K Class II Medical Device Compatibility Issue - Removal or Market withdrawal? US Food and Drug Administration (FDA) 2
J Can you stock parts without an 8130 and issue one at time of sale? Federal Aviation Administration (FAA) Standards and Requirements 2
R How many Corrective Action Request can be Issued by DCMA for a single issue? Nonconformance and Corrective Action 2
K NCR or CA? We're taking any issue we have and saying its a nonconformance Nonconformance and Corrective Action 11
R DCMA QAR Service Issue - Resident Inspectors absent from the plant AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4