Issues in Uncertainty Estimation
fahimk
Involved In Discussions
#1
Folks,
I need your help in tracing the errors and in further refining the Uncertainty Budget (sheet attached).
I wanted to keep it simple and therefore for Type B, I just considered the following three flavors:
UB1= due to calibrator inaccuracy
UB2 = calibration certificate
UB3 = resolution of the UUT
The confusion part: The Uncertainty (UB1) is causing a whooping increase in the final expanded uncertainty calculations.....
(In the Tab UB1, the column V shows the uncertainty after applied the essential coefficients (coverage factor, and the distribution coefficient etc).
Now, how i got it...
1) I got the concept after having a look on the Fluke's webinar: http://download.flukecal.com/pub/literature/webinar-uncertainty-presentation-Dec%202011.pdf (Page 37)
2) Then, i consulted the specification table from the Fluke's Operator Manual http://download.flukecal.com/pub/literature/5520a___omeng0600.pdf (Section: 1-12. Specifications)
I applied the method from (1) on (2) @ 50 Hz and eventually it lead me to the uncertainty in column V.
Can anybody please help me identify the blunder(s) i made?
Thanks a lot in advance...
Elsmar Forum Sponsor
#2
I am not certain (I'm just learning about making budgets myself), but hasn't the values in UB1 already been factored in the reference standard's uncertainty value? Hopefully someone with more uncertainty budget expertise than myself will chime in on this.
dgriffith
Quite Involved in Discussions
#3
I'm at work (playing with your spreadsheet... :geek: ) and I think you should look at your formulas. Example; Line 17 on UB1; S17 references A17, which is empty. Col. O is percent of output. In cell S17 you do math but it doesn't look like you account for the percent in O17. I think it's less confusing to have it as ppm, i.e. 12ppm of output. Then from S17, O17 * A17 (which is now 10V) is 0.000120uV, plus P17/1000000 (which is 20uV).
You now have [email protected]%CL. I'm almost certain this calibrator distribution is Gaussian, not rectangular. Removing the *Sqrt(3) from formula in V17 gets 0.000050, or 50uV.
fahimk
#4
Thanks mate for the response. I will be reviewing the work thoroughly!
Just a quick question: for writing final standard uncertainty, for category B, am i right if put two numbers in the denominator (one to represent the confidence level and another for the distribution)? Secondly, if no distribution scheme is provided then i should assume a rectangular distribution (i.e. denominator as root of (3)?
Looking forward....
Wolf.K
#6
Looking forward....
Hi, could you update your spreadsheet and post it here again? Elsmar is full of alpha version documents, but people unfortunately often forget to post the final result of the discussions here! :)
Thread starter Similar threads Forum Replies Date
D Regrading Test Blocks: Measurement Uncertainty (MU) Issues? Measurement Uncertainty (MU) 4
L 3rd party audit issues - No audit agenda received a week before the audit Registrars and Notified Bodies 7
K ISO 2859-5 (single run production) control plan issues FMEA and Control Plans 2
S Calibration Label issues General Measurement Device and Calibration Topics 14
H Relevant issues for the Context of the Organization IATF 16949 - Automotive Quality Systems Standard 5
S Powder coating issues Lean in Manufacturing and Service Industries 2
C Issues with UDI-DI and LAA/LUA registration in Eudamed EU Medical Device Regulations 2
S Biggest issues when producing MDR compliant Technical Files / DD EU Medical Device Regulations 4
K Powder coating issues Manufacturing and Related Processes 14
S ASTM F86 x Biocompatibility Issues Medical Device and FDA Regulations and Standards News 1
M Issues with leakage current testing for a CLASS II device since no PE and FE is present IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
Sidney Vianna Informational ESG - Environment, Social & Governance issues gaining ground in the Business World Sustainability, Green Initiatives and Ecology 22
A OQ and reporting issues Qualification and Validation (including 21 CFR Part 11) 2
A OEM On-Site Calibration issues during Covid19 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
G Issues/Roadblocks with APQP implementation APQP and PPAP 10
G Resistance to solve Issues Customer Complaints 8
G Engineering retest in line - Machine stability issues IATF 16949 - Automotive Quality Systems Standard 8
R Robotic weld cell - maintenance issues - expertise needed Manufacturing and Related Processes 0
K Implant Card - Full Manufacturer's Address - Space issues EU Medical Device Regulations 4
A Labeling - 2017/745 - EUMDR Requirements and Trademark Issues EU Medical Device Regulations 0
Q Connecting AS9100 D Clause 4.2 to 9.3.2 b - Interested parties and relevant issues AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 24
R Identifying internal issues.. at what level? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
Ronen E FDA News FDA issues guidance on bench testing in premarket submissions Other US Medical Device Regulations 0
dhuot1990 Issues with REES Environmental Monitoring System Manufacturing and Related Processes 1
P Why does the standard clause use the term Issues in place of Context - ISO 27001 4.1 IEC 27001 - Information Security Management Systems (ISMS) 3
Ashok sunder How to determine the internal and external issues in ISO 45001 standard? Occupational Health & Safety Management Standards 2
CCaantley Issues with Faro Quantum S and LLP General Measurement Device and Calibration Topics 0
M FDA News The FDA Issues Final Rule on Medical Device Classification Procedures Medical Device and FDA Regulations and Standards News 0
S Risk Approach doesn't address External Issues (Auditor's Comment) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 30
qualprod Closing Out Negative Issues in Context and Interested Parties ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
O Private label issues For EU - Customer does not have CE Mark CE Marking (Conformité Européene) / CB Scheme 7
O Examples of the external and internal issues and their risks and opportunities IATF 16949 - Automotive Quality Systems Standard 2
Y ISO 9001:2015 Cl. 4.1 Note 1 - External and Internal issues ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
bio_subbu Indian government issues guidance on Grouping Medical Devices in a Single Submission Other Medical Device Regulations World-Wide 1
qualprod External/internal issues removed as well? ISO 9001:2015 Cl. 4.1 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
bio_subbu India’s CDSCO Issues FAQ's on Medical Device Rule, 2017 Other Medical Device Regulations World-Wide 0
M Capturing Issues of Devices Past Lifetime Nonconformance and Corrective Action 3
A ISO 9001 Help - New Quality Manager - Upper Management Support Issues ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 23
A EMC Emission Issues - Communication between PC and Monitor IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Looking for an example of Issues Log ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
A One site of multi-site company closing - Quality transfer issues Quality Manager and Management Related Issues 7
Ronen E FDA issues draft guidance on form & content of UDI Other US Medical Device Regulations 0
Ronen E FDA issues draft guidance on additive manufacturing Other US Medical Device Regulations 1
L Packaging Issues Resulted CAR by customer Manufacturing and Related Processes 13
M What is the value of documenting in-process rework for easily-detectable issues? Quality Manager and Management Related Issues 4
bio_subbu USFDA issues draft Guidance on the Acceptance of Medical Device Clinical Data Other US Medical Device Regulations 1
M Issues with UDI additions to 21 CFR 820 requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
bio_subbu USFDA issues final guidance on "Reprocessing Medical Devices in Health Care Settings" Other US Medical Device Regulations 1
V What tool do you use for systemic issues? Problem Solving, Root Cause Fault and Failure Analysis 2
J CFDA issues Guiding Opinions on Enhancing the Construction of Food and Drug Inspectio China Medical Device Regulations 1
Similar threads
Top Bottom