IVD/Medical Device Registration (with MHRA DORS account in the UK)

fondantcookie

Starting to get Involved
#1
Hi all...

Noobie question again... (another...)

Do all classifications of IVD/Medical device need registering on the MHRA DORS account in the UK?

Or does this just apply to self-declared/general IVD; as the MHRA are made aware of higher classification devices through the notified body involvement?

Thanks in advance! I cannot tell you how helpful you folk are :) Massive thank you
 
Elsmar Forum Sponsor

dgrainger

Trusted Information Resource
#2
From MHRA IVD guidance: https://www.gov.uk/government/publi...ostic-medical-devices-guidance-on-legislation

4.2 Registration
Pursuant to Regulation 44, a manufacturer with a registered place of business in the UK who places a relevant device on the market (remember that for these purposes, market means the EU market) or who makes available a device for performance evaluation under his own name must register with the MHRA.
In addition, a person with a registered places of business in the UK who (a) places a relevant device on the UK market, or (b) who makes a device available for performance evaluation, on behalf of a manufacturer who does not have a registered place of business in the Community or in a state which is party to an Association Agreement, must register with the MHRA (see also ‘Authorised Representatives’ below).
Registration will not be required if the IVD was first placed on the market in another Member State (or if applicable in a state which is a party to an Association Agreement) and the manufacturer or his authorised representative has already registered with the Competent Authorities of that state (although note the transitional arrangements which apply pending the setting up of the European database, as to which see later).
Further guidance for registration in the UK is given on our website.
 

fondantcookie

Starting to get Involved
#3
From MHRA IVD guidance:

4.2 Registration
Pursuant to Regulation 44, a manufacturer with a registered place of business in the UK who places a relevant device on the market (remember that for these purposes, market means the EU market) or who makes available a device for performance evaluation under his own name must register with the MHRA.
In addition, a person with a registered places of business in the UK who (a) places a relevant device on the UK market, or (b) who makes a device available for performance evaluation, on behalf of a manufacturer who does not have a registered place of business in the Community or in a state which is party to an Association Agreement, must register with the MHRA (see also ‘Authorised Representatives’ below).
Registration will not be required if the IVD was first placed on the market in another Member State (or if applicable in a state which is a party to an Association Agreement) and the manufacturer or his authorised representative has already registered with the Competent Authorities of that state (although note the transitional arrangements which apply pending the setting up of the European database, as to which see later).
Further guidance for registration in the UK is given on our website.
Thanks Dgrainger.

I have heard a number of times though that we only need to register a device/IVD direct with the MHRA if it's Class I/General IVD (ie no involvement of the notified body).

I was thinking this must be because there is some kind of communication between the notified body and the MHRA for the higher classification devices (where a NB is involved)? Or is this not the case...?

Thanks for your help :cool: Slowly but surely getting an understanding :eek:
 
Thread starter Similar threads Forum Replies Date
M Informational TGA – European implementation of Medical Device and IVD Regulations – Implications for Australia Medical Device and FDA Regulations and Standards News 0
P Usability/human factors engineering requirement (standard) for IVD medical device Other Medical Device Related Standards 2
R Medical Device (IVD) - Selling RUO to Customers EU Medical Device Regulations 2
C ISO 13485 Medical Device File requirements for IVD ISO 13485:2016 - Medical Device Quality Management Systems 1
T Is an EEG (Electroencephalogram) an IVD or a Medical Device? Other US Medical Device Regulations 14
F CE and FDA Mark on same IVD/Medical Device Label EU Medical Device Regulations 12
N New MDR EU Medical Device/IVD Regulations - Definition of Importer EU Medical Device Regulations 1
F CE Marking Algorithm for an IVD use only Medical Device EU Medical Device Regulations 7
Ajit Basrur New Medical Device and IVD Registration Rules - Pakistan Other Medical Device Regulations World-Wide 5
E Republic Of Georgia Medical Device Regulations for IVD Other Medical Device Regulations World-Wide 6
R Shipping Stability for Class III IVD (Non Sterile) Medical Device Other Medical Device Related Standards 7
J FDA Compliant Quality System for Virtual Manufacturing (IVD Medical Device)? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
W IVD Medical Device Registration in Indonesia Other Medical Device Regulations World-Wide 2
E IVD Medical Device Product Registration in Myanmar Other Medical Device Regulations World-Wide 3
E IVD Medical Device Product Registration in Indonesia Other Medical Device Regulations World-Wide 2
K South Korea IVD Medical Device Classifications Other Medical Device Regulations World-Wide 13
P IVD Medical Device Authorized Representative Requirements in EU EU Medical Device Regulations 6
bio_subbu GHTF issues revised document "Definitions of ?Medical Device? and ?IVD? Other Medical Device Regulations World-Wide 1
S Medical Device/IVD Manufacturer Supplier Requirements - Compliant to ISO 13485?? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
G IVD (In Vitro Diagnostic) MDD (Medical Device Directive) Borderline Issues EU Medical Device Regulations 3
L Canadian Medical Device and IVD (In Vitro Diagnostic) Requirements US Food and Drug Administration (FDA) 3
S Single Medical Device Warranty card for both RUO and IVD!! ISO 13485:2016 - Medical Device Quality Management Systems 1
W Requirements for using Open Source Software in Medical Devices - IVD Medical Device IEC 62304 - Medical Device Software Life Cycle Processes 10
M IEC 60601 vs. IEC 61010 - CE Mark for our IVD medical device IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
E Clinical Evaluation of IVD Medical Devices CE Marking (Conformité Européene) / CB Scheme 8
B IVD Medical Devices Performance Assessment EU Medical Device Regulations 4
C 2012 Registration Tax for Medical Devices and IVD in France EU Medical Device Regulations 0
S IVD Medical Devices Training for 2010 Training - Internal, External, Online and Distance Learning 3
P Mexican Requirements for IVD Medical Devices ISO 13485:2016 - Medical Device Quality Management Systems 7
V Purchasing process needs for R&D department - IVD Medical Devices ISO 13485:2016 - Medical Device Quality Management Systems 3
H Software Interface Translation IVD Regulation EU Medical Device Regulations 0
I IVD - 98/79EC - class IIa - Annex to the EC from the NB states the CM as one of the facilities? CE Marking (Conformité Européene) / CB Scheme 2
G CE marking of MD and IVD set EU Medical Device Regulations 3
D List A IVD - Change to material supplier EU Medical Device Regulations 3
D IVD (in vitro diagnostic) Labelling Requirements - 98/79/EC CE Marking (Conformité Européene) / CB Scheme 2
J IVD instrument as an accessory EU Medical Device Regulations 3
C Where to place CE/label in Sterile IVD EU Medical Device Regulations 1
S A clinical performance evaluation study with an IVD product as Investagional Use product - Clinical Monitor requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
D Software User Interface Languages for LVD and IVD CE Marking (Conformité Européene) / CB Scheme 2
Watchcat ISO 13485 for IVD (In-vitro Diagnostic Device) Manufacturers? ISO 13485:2016 - Medical Device Quality Management Systems 9
H Incomplete IVD kit under IVD Regulation EU Medical Device Regulations 1
J 510(k) for a control kit for an external IVD test kit 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
P Do I need to get registered or have German entity to sell IVD products in Germany? CE Marking (Conformité Européene) / CB Scheme 2
E Sharepoint for ISO 13485 QMS for small IVD company ISO 13485:2016 - Medical Device Quality Management Systems 7
samarth IVD manufacturer and service provider ISO 14971 - Medical Device Risk Management 2
J Universal Power Cord with Adapters - IVD IEC 61010-2-101 CE Marking (Conformité Européene) / CB Scheme 6
M PMS template for FDA submission (De novo submission for a Class II IVD) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
J IVD CE transfer to another manufacturer EU Medical Device Regulations 4
V CE Mark under MDD or IVDR - IVD company doing Virtual Manufacturing Other Medical Device Regulations World-Wide 3
C CE marking for general IVD (self-certified) & ISO 13485 QMS requirements - auditing EU Medical Device Regulations 6

Similar threads

Top Bottom