I am new to IVDs from a background in medical devices and am a little confused how meteorological traceability can work in IVDs when there are no reference measurement procedures for particular analyte in matrix concentrations and how to deal with assigned values which are not lengths and weights.
Let's assume hypothetically that we are an IVD manufacturer which produces an IVD kit for measuring NaCl in serum which requires an assigned value to a calibrator of 10 mg Sodium Chloride (NaCl) per 1 ml serum.
We are fortunate that there exists a reference standard material for Sodium Chloride (NIST Standard Reference Material 919b) with a described reference measurement procedure how this standard reference material was determined to have a known concentration of pure NaCl.
Let's assume further, that unfortunately there seems to be missing a reference measurement procedure for NaCl in serum and thus no reference measurement standard for NaCl in serum can be bought from any reference measurement laboratory or a measurement be ordered as a service. In this hypothetical world only the manufacturers IVD kit itself exists for which we want to create the calibrator.
What shall the IVD manufacturer do in this situation?
Is it allowed to maintain a meteorological traceability by weighting the analyte NaCl (for which we can buy traceable secondary standards) and weighting the matrix serum and manufacturing a calibrator of known concentration NaCl in serum by just mixing those two known quantities? We can verify the concentration then with the IVD kit, but it cannot measure with a higher order measurement procedure.
In other words does ISO 17511 allow any deviation from measurement of a concentration of analyte directly by arguing about how the lower level calibrator was produced by mixing analyte in matrix?
Elsmar Forum Sponsor
dwperron
Trusted Information Resource
#4
Let's just say that I don't want to get involved in dealing with hypothetical equipment and processes.
I find it hard enough dealing with actual problems.
#5
I was hoping this would be easier to understand.
Analyte = Iohexol (e.g. European Pharmacopeia 1141)
Matrix = Human Serum
Manufacturer Method = HPLC-based analysis
John Broomfield
Staff member
Super Moderator
#7
European Pharmacopoeia 1141 is an internationally recognized reference standard.
Along with their use of a standard method, labs provide comparable results.
Given these controls what exactly is the problem you want to solve?
#8
Dear John,
The challenge is to measure the concentration of the analyte (Iohexol) in a matrix (Serum) for which there isn't a reference measurement procedure.
We can use our own method (3% bias) using our own manufactured calibrants, but there is no higher order measurement method using which we could trace it to the EP 1141.
John Broomfield
Staff member
Super Moderator
#9
So, you will have organized inter-laboratory comparison studies (ISO/IEC 17043) and proficiency testing (ISO Guide 43).
Did either or both of these deliver the required degree of confidence?
#10
We haven't done this and haven't found any ISO 17043 accredited proficiency testing laboratory which offers our combination of analyte / measurand.
But did I understand you correctly, that this would be one way forward to find somebody accredited to ISO 17043 and let them develop a method or use a method from literature for assessing it? If I look to the offerings of the proficiency laboratories I find only well-established markers/chemicals and not novel markers.
Thread starter Similar threads Forum Replies Date
B Metrological Traceability and IVD Reference System ISO 13485:2016 - Medical Device Quality Management Systems 1
S CE Marking components within IVD CE Marking (Conformité Européene) / CB Scheme 0
B IVD - GSPR checklist Other Medical Device Related Standards 0
E Does IVD need to integrate with hospital IT infrastructure? Medical Information Technology, Medical Software and Health Informatics 2
T CE-IVD mark on packaging EU Medical Device Regulations 1
A Is computing Hardware part of the IVD CE Marking (Conformité Européene) / CB Scheme 3
B Looking for SSCP file of IVD devices EU Medical Device Regulations 7
B Research Use Only (RUO) IVD Medical Devices in China China Medical Device Regulations 2
B What is the difference btw RUO vs IUO for IVD in FDA guidance ? US Food and Drug Administration (FDA) 12
Y Handbook recommended lightcycler for IVD Risk Management Principles and Generic Guidelines 1
S IVD Device History Record ISO 13485:2016 - Medical Device Quality Management Systems 3
J IVD sampling sizes for buyer Reliability Analysis - Predictions, Testing and Standards 0
D Low risk IVD study in the UK, do I need MHRA approval? UK Medical Device Regulations 1
A CE-IVD Technical File with Multiple IFUs EU Medical Device Regulations 4
A Ethics Committee Review Process for IVD Products EU Medical Device Regulations 2
F IVD registration in EU - Northern Ireland based company EU Medical Device Regulations 0
J IVD registration renewal in China China Medical Device Regulations 6
S IVD risk class II devices for Brazil and MDSAP Other Medical Device Regulations World-Wide 0
A IVD in MDR procedure pack EU Medical Device Regulations 2
H Software Interface Translation IVD Regulation EU Medical Device Regulations 0
I IVD - 98/79EC - class IIa - Annex to the EC from the NB states the CM as one of the facilities? CE Marking (Conformité Européene) / CB Scheme 2
G CE marking of MD and IVD set EU Medical Device Regulations 3
D List A IVD - Change to material supplier EU Medical Device Regulations 3
D IVD (in vitro diagnostic) Labelling Requirements - 98/79/EC CE Marking (Conformité Européene) / CB Scheme 2
J IVD instrument as an accessory EU Medical Device Regulations 3
C Where to place CE/label in Sterile IVD EU Medical Device Regulations 1
S A clinical performance evaluation study with an IVD product as Investagional Use product - Clinical Monitor requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
D Software User Interface Languages for LVD and IVD CE Marking (Conformité Européene) / CB Scheme 2
Watchcat ISO 13485 for IVD (In-vitro Diagnostic Device) Manufacturers? ISO 13485:2016 - Medical Device Quality Management Systems 10
H Incomplete IVD kit under IVD Regulation EU Medical Device Regulations 1
J 510(k) for a control kit for an external IVD test kit 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
P Do I need to get registered or have German entity to sell IVD products in Germany? CE Marking (Conformité Européene) / CB Scheme 2
E Sharepoint for ISO 13485 QMS for small IVD company ISO 13485:2016 - Medical Device Quality Management Systems 11
samarth IVD manufacturer and service provider ISO 14971 - Medical Device Risk Management 2
J Universal Power Cord with Adapters - IVD IEC 61010-2-101 CE Marking (Conformité Européene) / CB Scheme 6
M PMS template for FDA submission (De novo submission for a Class II IVD) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
J IVD CE transfer to another manufacturer EU Medical Device Regulations 4
V CE Mark under MDD or IVDR - IVD company doing Virtual Manufacturing Other Medical Device Regulations World-Wide 3
C CE marking for general IVD (self-certified) & ISO 13485 QMS requirements - auditing EU Medical Device Regulations 6
M Informational TGA – European implementation of Medical Device and IVD Regulations – Implications for Australia Medical Device and FDA Regulations and Standards News 0
S How to consider the relevant standards during development of ISO13482:2016 for IVD manufacturing Blood grouping Other Medical Device Related Standards 6
M EU IVD Risk Determination of sample collection EU Medical Device Regulations 0
J Deciding between Professional Use or Lay Person IVD Test - Human fecal stool specimens Other Medical Device and Orthopedic Related Topics 1
D RUOs in IVD products Other Medical Device and Orthopedic Related Topics 2
P Usability/human factors engineering requirement (standard) for IVD medical device Other Medical Device Related Standards 2
M IVD Registration - Use of Prospective Study Data Other Medical Device Regulations World-Wide 1
R Medical Device (IVD) - Selling RUO to Customers EU Medical Device Regulations 2
I IVD prototypes in EU market ISO 13485:2016 - Medical Device Quality Management Systems 2
N Own brand labelling/virtual manufacture of IVD's EU Medical Device Regulations 2
M IVD - Generating clinical data for IVD’s in the EU EU Medical Device Regulations 2
Similar threads
Top Bottom