Justification for Month End Calibration Date

M

mentalcrew

#1
To all- Running a QMS for a 13485 company and have made all calibration due dates month end. Wrote the Quality Manual to stae:

The calibration due date method used to determine when a unit is due for scheduled calibration is determined by the BCS. The determined method is based on requirements, cost, and the ability of Calibration Service Supplier to perform calibration. Any of the following methods are acceptable and may be applied as necessary by the BCS according to process requirements:
1. Month/Year Calibration – Due date is established based on the month the unit is calibrated and the interval assigned to the unit. Calibration will be due no later than midnight on the last day of the month on the certificate and/or label. This method allows the unit to go beyond the stated interval but will not carry over into the next month.
2. Month/Day/Year – Due date is established based on the day of the month the unit was calibrated and the interval assigned to the unit. Calibration will be due no later than midnight on the date on the certificate and/or label.
3. These methods are not exclusive and a supervisor/manager can request in writing or via the BCI electronic mail system for modification of the method to be approved by BCS.

Pretty self explanatory. 15 sites in the US are using this new system of month end but I now have one internal QA auditor stating that this is not compliant for FDA and 13485. I know that only the customer can establish cal due dates and the QM needs to reflect. What I am looking for is hard written facts some where to "prove" that we are comliant for month endcal dates. :bonk:Big stink has brewed and a wildfire is rampant. :bonk: Any information
 
Elsmar Forum Sponsor

mdurivage

Quite Involved in Discussions
#2
I have used the mm/yyyy format for calibration due dates. Never have had any issues with the FDA or otherwise. This method helps prevent silly NCR's in the case the IMTE was due on a Saturday and the calibration was performed the following Monday.

This system worked well as we generally scheduled our calibration house for the 4th Monday of every month.

Hope this helps!
 
J

Jon Nonns

#4
To all- Running a QMS for a 13485 company and have made all calibration due dates month end. Wrote the Quality Manual to stae:

The calibration due date method used to determine when a unit is due for scheduled calibration is determined by the BCS. The determined method is based on requirements, cost, and the ability of Calibration Service Supplier to perform calibration. Any of the following methods are acceptable and may be applied as necessary by the BCS according to process requirements:
1. Month/Year Calibration ? Due date is established based on the month the unit is calibrated and the interval assigned to the unit. Calibration will be due no later than midnight on the last day of the month on the certificate and/or label. This method allows the unit to go beyond the stated interval but will not carry over into the next month.
2. Month/Day/Year ? Due date is established based on the day of the month the unit was calibrated and the interval assigned to the unit. Calibration will be due no later than midnight on the date on the certificate and/or label.
3. These methods are not exclusive and a supervisor/manager can request in writing or via the BCI electronic mail system for modification of the method to be approved by BCS.

Pretty self explanatory. 15 sites in the US are using this new system of month end but I now have one internal QA auditor stating that this is not compliant for FDA and 13485. I know that only the customer can establish cal due dates and the QM needs to reflect. What I am looking for is hard written facts some where to "prove" that we are comliant for month endcal dates. :bonk:Big stink has brewed and a wildfire is rampant. :bonk: Any information
Just my :2cents: but I would personally turn the question on him to show you where in ISO 13485 and/or the CFRs it states that month end due dates are NOT acceptable. Both just require that you have a calibration cycle defined and that items past calibration or found out of tolerance are quarantined and product impact assessed. Good luck.

Jon
 

BradM

Staff member
Admin
#5
I have to agree with the others. I have never seen an issue with end of month due dates; nor have I really ever heard the practice questioned.

I think its better to have a consistent (and compliant) calibration process. If you start setting to the day, you will have more headaches/ issues/ past due/ etc.

This job is hard enough, without making it more difficult. :)
 
M

mentalcrew

#7
I have rounded back several times and can not get an answer of what is actually in non compliant. The issue came about with a worse case scenario. Unit calibrated on the 1st of the months and then had a six month interval with month end date. Technically giving the unit a possible 7 month calibration date. The term "VALIDATED calibration frequency" was thrown around. Nor can I get the validation on the frequency is said report.
 
L

Lucasmf

#8
I have never heard of a validated calibration frequency requirement. If you calibration process always captures 'as found' data prior to calibration then there should be no issue with a theoretical worst case 7 month interval. It merely becomes a manufacturer's risk if the unit is found OOT at the end of seven months and now your product quality is in question.

The requirement is that calibration occur at "specified intervals". As long your interval is adequately specified in your procedure and your equipment is not routinely OOT, you should be well within the requirements.
 
M

mentalcrew

#9
Heck I have never heard of a validated calibration frequency. Think this was an auditor talking about something he is not familiar with. Unforntunatley, I can not find something in writing. So now its a he said she said who knows better.
 
I

isoalchemist

#10
:2cents:Tell your auditor to chill out, the standards only require you to define your process and follow it. I've used a MM/DD/YY format for all my cal's and stated only the month was important and have had no problems.

I did get challanged on a registration audit with a similar worst case as you stated, but the focus was not on calibration frequency as much as if the people they were interviewing were aware of why the procedure was written the way it was. They were seaching for the rational behind the development of our calibration process to ensure we did not pick numbers out of the air. I don't think your internal auditor is searching for that.
 
Thread starter Similar threads Forum Replies Date
M V&V phase: Justification of acceptance criteria (statistical method ) - (Medical Device) Design and Development of Products and Processes 2
B Exclusions or justification for non-applicability of IEC standards Reliability Analysis - Predictions, Testing and Standards 1
B Per 21CFR 820.198 (b) - Justification for no investigation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 22
M CER Adverse events - Justification of databases searched EU Medical Device Regulations 5
P IATF 16949 Clause 8.4.2.3 - Justification for non-certified suppliers IATF 16949 - Automotive Quality Systems Standard 14
O Statistical justification of sampling size in V&V tests ISO 13485:2016 - Medical Device Quality Management Systems 5
V Statistical basis and justification while comparing / changing sampling plans Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 11
J Calibration Justification - Requirements for Measurement with a Calibrated Device General Measurement Device and Calibration Topics 1
SATHYABALARAMAN Exclusion of IATF 16949 Clauses - 8.5.5.1 & 8.5.5.2 & Justification IATF 16949 - Automotive Quality Systems Standard 9
N Sample Size Justification for Medical Device Shelf-Life Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
M Justification for Not Having a Validation Plan for a Critical Process ISO 13485:2016 - Medical Device Quality Management Systems 4
B EMC Repeat Justification - Reducing NRTL testing required after modification of ME IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Biocompatibility omission justification for FDA 510(k) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
G Exemption Filing / Exemption Justification RoHS, REACH, ELV, IMDS and Restricted Substances 9
H Factory Dictated Sample Size Justification Statistical Analysis Tools, Techniques and SPC 9
J Justification for Exclusion of Design and Development ISO 13485:2016 - Medical Device Quality Management Systems 8
S Justification of Lifetime of a Medical Device including Component Aspects ISO 13485:2016 - Medical Device Quality Management Systems 13
C Justification for No CAPA in Complaint Files Nonconformance and Corrective Action 24
Q Sample Size justification for IEC 60601-1 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
Wes Bucey A justification to take HR out of hiring Career and Occupation Discussions 3
S Establishing an MSA Plan and Justification Suggestions wanted Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
G Sampling Plan - Sample Size Justification for a Bottling Cleaning Process Validation Qualification and Validation (including 21 CFR Part 11) 2
X Justification of Decrease in Receivables, Increase in Negative Cashflows Coffee Break and Water Cooler Discussions 2
J Sampling Plan Justification and Rationale Inspection, Prints (Drawings), Testing, Sampling and Related Topics 9
S MDR (Medical Device Reporting) Justification - Canned Statements Other US Medical Device Regulations 2
P Out of Tolerance Gage - Justification for not Recalling Product Misc. Quality Assurance and Business Systems Related Topics 7
Wes Bucey Justification for organization's tax-exempt status? Funny Stuff - Jokes and Humour 1
C Looking for Skills Matrix & Justification of skill levels. Document Control Systems, Procedures, Forms and Templates 5
A Justification for not performing overseas supplier audit General Auditing Discussions 6
Q Justification/Rationale for not investigating a nonconformance ISO 13485:2016 - Medical Device Quality Management Systems 19
P Reference Only Gages - "For Reference Only" Rationale (Reasoning/Justification) General Measurement Device and Calibration Topics 25
G Justification for Supplier Quality Engineer Career and Occupation Discussions 3
M Justification for not doing Gage R&R Studies Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
J Medical Device FMEA Occurence Rating Scale Justification FMEA and Control Plans 3
J What would be the justification for exclusion of ISO9001 clause 7.5.1? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
J Software Justification in a Lean Environment - Convincing Management Lean in Manufacturing and Service Industries 12
R Frequency of self control and costs - Justification for 4 Hours vs. Every Hour IATF 16949 - Automotive Quality Systems Standard 2
N CE Mark Auditing - All they had for CE was a written Technical Justification EU Medical Device Regulations 5
R Cost Justification Based Upon Intangibles - Cost of Quality Misc. Quality Assurance and Business Systems Related Topics 4
E Our company is planning to file MDD not MDR next month. Do we require to show chemical characterization report ? CE Marking (Conformité Européene) / CB Scheme 1
B IATF 16949 News Six month extension on all valid IATF 16949 certs IATF 16949 - Automotive Quality Systems Standard 10
A API Spec Q1 9th Edition - 12 month Internal Audit Schedule Audit Nonconformance Oil and Gas Industry Standards and Regulations 10
D QMS Performance Based on how many NCRs were generated month to month Nonconformance and Corrective Action 8
D Can an AS9100 auditor issue an NCR a month after the audit? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
J New position and AS9100 audit is 1 month away- need advice AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
Colin I am retiring at the end of this month - March 2017 Coffee Break and Water Cooler Discussions 19
C Calibration Dates - Exact day or end of month? General Measurement Device and Calibration Topics 24
O New Job 1 Month from Recertification Audit - Missing Documents, no Internal Audits ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 25
Randy Looky here - Check out my 9-month old Great-Grandson Hayden Coffee Break and Water Cooler Discussions 4
SATHYABALARAMAN Quality Month - November 2012 Imported Legacy Blogs 3

Similar threads

Top Bottom