Justification for Not Having a Validation Plan for a Critical Process
medwise
Dear Experts,
We are a manufacturer of a class III medical device. There are several stages to manufacturing which includes chemical synthesis, lyophilization, packaging, gas chromatography, gamma radiation to name a few.
We have validation protocols and validation reports for all the above mentioned processes. However, do not have validation plans for chemical synthesis, lyophilization and packaging. Lyophiliser for example is qualified and process validation is covered as part of the entire process validation of the device we manufacture. For gas chromatography, we do 100% inspection on the outcome therefore we have got away without having a specific validation plan for this process.
My question is how can we justify to our NB that our critical process such as lyophilization and chemical synthesis is validated as part of the product process validation but does not have a validation plan.
Ronen E
Problem Solver
Moderator
Does "the entire process validation of the device you manufacture" have a plan? If so, it should be mentioning and covering every one of its sub-processes, either directly or by reference. If it does, your problem is solved.
medwise
For the entire process validation of the device we only have validation protocol and process validation reports unfortunately. Validation Protocol however covers responsibilities, traceability, brief description of the process, validation strategy, test and acceptance criteria, deviations. Is this sufficient to justify that protocol acts as validation plan?
Ronen E
Problem Solver
Moderator
Sounds good to me. I'm not sure how you perceive the difference between a protocol and a plan. What you describe as the protocol's contents is roughly what I'd include in the plan.
lawcch
in any critical process or work, there are some form of validation of processes or critical control point (CCP) as in the CCP Plan in HACCP Plan in Food safety management system, ISO 22000.
Therefore, there should be NO justification to exclude any critical processes with validation of processes if the standard or product spec. requires them.
B Exclusions or justification for non-applicability of IEC standards Reliability Analysis - Predictions, Testing and Standards 1
B Per 21CFR 820.198 (b) - Justification for no investigation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 22
M CER Adverse events - Justification of databases searched EU Medical Device Regulations 5
P IATF 16949 Clause 8.4.2.3 - Justification for non-certified suppliers IATF 16949 - Automotive Quality Systems Standard 14
O Statistical justification of sampling size in V&V tests ISO 13485:2016 - Medical Device Quality Management Systems 5
V Statistical basis and justification while comparing / changing sampling plans Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 11
J Calibration Justification - Requirements for Measurement with a Calibrated Device General Measurement Device and Calibration Topics 1
SATHYABALARAMAN Exclusion of IATF 16949 Clauses - 8.5.5.1 & 8.5.5.2 & Justification IATF 16949 - Automotive Quality Systems Standard 9
N Sample Size Justification for Medical Device Shelf-Life Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
B EMC Repeat Justification - Reducing NRTL testing required after modification of ME IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Biocompatibility omission justification for FDA 510(k) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
G Exemption Filing / Exemption Justification RoHS, REACH, ELV, IMDS and Restricted Substances 9
H Factory Dictated Sample Size Justification Statistical Analysis Tools, Techniques and SPC 9
J Justification for Exclusion of Design and Development ISO 13485:2016 - Medical Device Quality Management Systems 8
S Justification of Lifetime of a Medical Device including Component Aspects ISO 13485:2016 - Medical Device Quality Management Systems 13
C Justification for No CAPA in Complaint Files Nonconformance and Corrective Action 24
Q Sample Size justification for IEC 60601-1 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
Wes Bucey A justification to take HR out of hiring Career and Occupation Discussions 3
S Establishing an MSA Plan and Justification Suggestions wanted Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
G Sampling Plan - Sample Size Justification for a Bottling Cleaning Process Validation Qualification and Validation (including 21 CFR Part 11) 2
M Justification for Month End Calibration Date Calibration Frequency (Interval) 24
X Justification of Decrease in Receivables, Increase in Negative Cashflows Coffee Break and Water Cooler Discussions 2
J Sampling Plan Justification and Rationale Inspection, Prints (Drawings), Testing, Sampling and Related Topics 9
S MDR (Medical Device Reporting) Justification - Canned Statements Other US Medical Device Regulations 2
P Out of Tolerance Gage - Justification for not Recalling Product Misc. Quality Assurance and Business Systems Related Topics 7
Wes Bucey Justification for organization's tax-exempt status? Funny Stuff - Jokes and Humour 1
C Looking for Skills Matrix & Justification of skill levels. Document Control Systems, Procedures, Forms and Templates 5
A Justification for not performing overseas supplier audit General Auditing Discussions 6
Q Justification/Rationale for not investigating a nonconformance ISO 13485:2016 - Medical Device Quality Management Systems 19
P Reference Only Gages - "For Reference Only" Rationale (Reasoning/Justification) General Measurement Device and Calibration Topics 25
G Justification for Supplier Quality Engineer Career and Occupation Discussions 3
M Justification for not doing Gage R&R Studies Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
J Medical Device FMEA Occurence Rating Scale Justification FMEA and Control Plans 3
J What would be the justification for exclusion of ISO9001 clause 7.5.1? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
J Software Justification in a Lean Environment - Convincing Management Lean in Manufacturing and Service Industries 12
R Frequency of self control and costs - Justification for 4 Hours vs. Every Hour IATF 16949 - Automotive Quality Systems Standard 2
N CE Mark Auditing - All they had for CE was a written Technical Justification EU Medical Device Regulations 5
R Cost Justification Based Upon Intangibles - Cost of Quality Misc. Quality Assurance and Business Systems Related Topics 4
NDesouza No Quality Professional vs Having a Quality Professional Benchmarking 24
A EU MDR Annex I section 11.3 - Medical Devices labelled as having a specific microbial state EU Medical Device Regulations 7
O Inspecting paint and having workable standards in the supply chain Supplier Quality Assurance and other Supplier Issues 1
P Received a minor for not having good measureables/goals. Need help with KPIs. IATF 16949 - Automotive Quality Systems Standard 52
N GDP - Having to provide a reason when writing N/A? ISO 13485:2016 - Medical Device Quality Management Systems 1
O How do reduce the risk of suppliers having similar problems? Supplier Quality Assurance and other Supplier Issues 4
D Having got ISO 13485:2016 are you dropping ISO 9001? ISO 13485:2016 - Medical Device Quality Management Systems 8
Q Partial non applicability of ISO 9001 Cl. 7.1.5 (not having services)? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
A Not having an internal audit during the year ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
T Having Fun While Learning ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
S Having a hard time closing CAR/ PAR Nonconformance and Corrective Action 13