Justification of Lifetime of a Medical Device including Component Aspects

S

Sherif Amer

Hi all,

We are about to release our medical product but we are having a difficulty determining the lifetime of the device. We have looked at the datasheets of the components the device is built of, found some information about most of them, not all of them. What should we do with the rest of the components that I can't find their lifetime?

Sherif
 

Wes Bucey

Prophet of Profit
Re: Justification of lifetime

Of course, the first question in my mind is whether the device is external or implantable.

There is a guideline
Lifetime Advice – ISO/TR 14969:2004
ISO/TR 14969:2004 Medical devices – Quality management systems – Guidance on the application of ISO 13485:2003, clause 7.1.3 offers advice on setting the lifetime of a device.

The guidance recommends documenting the defined lifetime of the medical device as well as the rationale for determining the lifetime. Some considerations in the determination include:


  1. shelf life of the medical device
  2. expiration date of the device or its components
  3. the number of cycles or periods of use of the medical device
  4. anticipated material degradation
  5. stability of the packaging material
  6. for implantable devices, the residual risk that results from the entire period of residence of the device inside the patient’s body
  7. for sterile medical devices, the ability to maintain sterility
  8. the organization’s contractual or regulatory obligation to support and service the device
  9. spare parts cost and availability;
  10. legal considerations including liability

Risk Management Considerations
The defined product lifetime can help control residual risks that can increase to unacceptable levels over extended device periods of use.1
ISO 14971:2003 Clause 2.7 define the device life-cycle as all phases in the life of a medical device, from the initial conception to final decommissioning and disposal.

The Risk Management Plan needs to include the scope of the planned risk management activities, identifying and describing the medical device and the life-cycle phases for which each element of the plan is applicable.

Because risks appear throughout the product life-cycle, the life-cycle definition is important. Risks that appear at one point in the life-cycle may be managed by action or controls at a different point. This argues for the application of risk management throughout the life-cycle from design to disposal.
Typically, if the device has been designed so individual components are replaceable, then the primary consideration is whether continual repair and replacement may be feasible, extending the useful life of the device. In the absence of "easily" replaceable components, the lifetime devolves to the estimated life of the shortest lived component.

In an auto, windshield wipers are easily replaceable and don't affect the lifetime of the vehicle. Engines or transmissions "may" be replaceable, but at what point is the useful life of the original vehicle affected before it becomes "rebuilt?"
 

AndyN

Moved On
Re: Justification of lifetime of a Medical Device

<snip> We are about to release our medical product but we are having a difficulty determining the lifetime of the device. We have looked at the datasheets of the components the device is built of, found some information about most of them, not all of them. What should we do with the rest of the components that I can't find their lifetime? <snip>
You may wish to get some expert advice about the specifics of the devices etc, since you haven't given us that information. Sadly, quoting a guide - even an ISO one - isn't going to help you it seems to me. If you don't have specific information and you release the product - wouldn't that put the patient and your organization at risk?
 

Marc

Fully vaccinated are you?
Leader
Re: Justification of lifetime of a Medical Device

<snip> What should we do with the rest of the components that I can't find their lifetime? <snip>
Help with this aspect from medical device experts here will be appreciated.
 

Ajit Basrur

Leader
Admin
Hi all,

We are about to release our medical product but we are having a difficulty determining the lifetime of the device. We have looked at the datasheets of the components the device is built of, found some information about most of them, not all of them. What should we do with the rest of the components that I can't find their lifetime?

Sherif

Welcome to the Cove, Sherif :bigwave:

Couple of questions -

1. When you refer as a medical product, is it a "finished medical device"?

2. What does "release" mean? Is the device approved by the regulatory agencies? If yes, the lifetime of the device would have been already submitted as part of label.

3. Which markets are you targeting for commercial release of the product?
 

somashekar

Leader
Admin
Hi all,

We are about to release our medical product but we are having a difficulty determining the lifetime of the device. We have looked at the datasheets of the components the device is built of, found some information about most of them, not all of them. What should we do with the rest of the components that I can't find their lifetime?

Sherif
Also look down this page in the "Similar Discussion Threads". You may find some useful information and your question may be answered perhaps.
 
M

MIREGMGR

You as a regulatory person need to have a talk with senior management about the extent to which your product design department understands the company's regulatory obligations. No product is properly designed without a design life as a design input, and an appropriate documented engineering analysis, probably backed by testing, to show that the design output meets the various aspects of that design input.

That might include product shelf stability before use, product functionality in use, sterilization stability of the product's terminal packaging if applicable, and other factors.

Sterilization stability, for one, might be validated via accelerated aging studies, backed up by post-release-completed real time studies. There's a considerable body of knowledge regarding how to do this.

The comment above about risk is very much on point. If your design hasn't been validated to conform to its requisite life requirements, you don't have any idea whether it's risk-acceptable.
 
S

Sherif Amer

Thanks a lot guys. The device is external that doesn't need any sterilization. We're after the CE mark. I think I'm going to ask for someone's help around me.
 
R

Roland Cooke

I believe the cool kids talk about "product life cycle" these, on the basis that any particular device could have a bunch of different "lifetimes" depending on its phase of use.
 

Marcelo

Inactive Registered Visitor
"Product life cycle" is generally related to the technology, not each product manufactured. But normally a device has only one defined lifetime.

In fact, I usually says there?s two lifetime:

- the "true" lifetime, meaning, how long the device can perform (tied to reliability)

- and the "regulatory" lifetime, which is how long the manufacturer wants to be responsible for the device (tied to other considerations, and can include reliability).
 
Thread starter Similar threads Forum Replies Date
adztesla Medical device Drawing related change justification Other Medical Device Related Standards 0
rob3027 ISO 13485 justification for design and development exclusion for medical packaging ISO 13485:2016 - Medical Device Quality Management Systems 18
J Justification for visual inspection ISO 13485:2016 - Medical Device Quality Management Systems 3
T Justification of use of non-harmonized standards for MDR conformity Other Medical Device Related Standards 12
somashekar Justification for Non-applicable clause at Stage 2, in clause 4. ISO 13485:2016 - Medical Device Quality Management Systems 11
T Justification of "Design Excluded" in IATF 16949 Design and Development of Products and Processes 5
M V&V phase: Justification of acceptance criteria (statistical method ) - (Medical Device) Design and Development of Products and Processes 2
B Exclusions or justification for non-applicability of IEC standards Reliability Analysis - Predictions, Testing and Standards 1
B Per 21CFR 820.198 (b) - Justification for no investigation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 22
M CER Adverse events - Justification of databases searched EU Medical Device Regulations 5
P IATF 16949 Clause 8.4.2.3 - Justification for non-certified suppliers IATF 16949 - Automotive Quality Systems Standard 14
O Statistical justification of sampling size in V&V tests ISO 13485:2016 - Medical Device Quality Management Systems 5
V Statistical basis and justification while comparing / changing sampling plans Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 11
J Calibration Justification - Requirements for Measurement with a Calibrated Device General Measurement Device and Calibration Topics 1
SATHYABALARAMAN Exclusion of IATF 16949 Clauses - 8.5.5.1 & 8.5.5.2 & Justification IATF 16949 - Automotive Quality Systems Standard 9
N Sample Size Justification for Medical Device Shelf-Life Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
M Justification for Not Having a Validation Plan for a Critical Process ISO 13485:2016 - Medical Device Quality Management Systems 4
B EMC Repeat Justification - Reducing NRTL testing required after modification of ME IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Biocompatibility omission justification for FDA 510(k) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
G Exemption Filing / Exemption Justification RoHS, REACH, ELV, IMDS and Restricted Substances 9
H Factory Dictated Sample Size Justification Statistical Analysis Tools, Techniques and SPC 9
J Justification for Exclusion of Design and Development ISO 13485:2016 - Medical Device Quality Management Systems 8
C Justification for No CAPA in Complaint Files Nonconformance and Corrective Action 24
Q Sample Size justification for IEC 60601-1 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
Wes Bucey A justification to take HR out of hiring Career and Occupation Discussions 3
S Establishing an MSA Plan and Justification Suggestions wanted Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
G Sampling Plan - Sample Size Justification for a Bottling Cleaning Process Validation Qualification and Validation (including 21 CFR Part 11) 2
M Justification for Month End Calibration Date Calibration Frequency (Interval) 24
X Justification of Decrease in Receivables, Increase in Negative Cashflows Coffee Break and Water Cooler Discussions 2
J Sampling Plan Justification and Rationale Inspection, Prints (Drawings), Testing, Sampling and Related Topics 9
S MDR (Medical Device Reporting) Justification - Canned Statements Other US Medical Device Regulations 2
P Out of Tolerance Gage - Justification for not Recalling Product Misc. Quality Assurance and Business Systems Related Topics 7
Wes Bucey Justification for organization's tax-exempt status? Funny Stuff - Jokes and Humour 1
C Looking for Skills Matrix & Justification of skill levels. Document Control Systems, Procedures, Forms and Templates 5
A Justification for not performing overseas supplier audit General Auditing Discussions 6
Q Justification/Rationale for not investigating a nonconformance ISO 13485:2016 - Medical Device Quality Management Systems 19
P Reference Only Gages - "For Reference Only" Rationale (Reasoning/Justification) General Measurement Device and Calibration Topics 25
G Justification for Supplier Quality Engineer Career and Occupation Discussions 3
M Justification for not doing Gage R&R Studies Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 13
J Medical Device FMEA Occurence Rating Scale Justification FMEA and Control Plans 3
J What would be the justification for exclusion of ISO9001 clause 7.5.1? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
J Software Justification in a Lean Environment - Convincing Management Lean in Manufacturing and Service Industries 12
R Frequency of self control and costs - Justification for 4 Hours vs. Every Hour IATF 16949 - Automotive Quality Systems Standard 2
N CE Mark Auditing - All they had for CE was a written Technical Justification EU Medical Device Regulations 5
R Cost Justification Based Upon Intangibles - Cost of Quality Misc. Quality Assurance and Business Systems Related Topics 4
R ISO 13485:2016 “Lifetime” of a product Manufacturing and Related Processes 4
B System lifetime for SaMD Medical Device and FDA Regulations and Standards News 8
M Lifetime of medical device EU Medical Device Regulations 16
K Definition and Example of Shelf life, Expiry date, Expected service life and Lifetime EU Medical Device Regulations 6