Labeling Responsibility - Manufacturer vs. Repacker? Proper Component Use
G
Gert Sorensen
Is there an existing or new clause in the CFR which addresses the question of labeling for medical kits assembled from finished products made by other manufacturers, especially when the assembler is performing a sterilization process on the assembled kit? Specifically, I'm interested in what input the third party manufacturers could or should have on the labeling to assure proper use of the components they have manufactured which are in the kit
Al Rosen
Is this a kit for use by a physician or healhcare professional?
G
Gert Sorensen
Hi Al,
The kits are used by doctors during surgery. They can contain various items such as stents, catheters, guidewires, dilators, needles etc.
Al Rosen
Gert Sorensen said:
Hi Al,
If they are prescription devices, you should read Guidance for Industry and FDA on Alternative to Certain Prescription Device Labeling Requirements.
I don't think the issue is what regulations exist, but what agreement you have with your customer. You can look at all the guidance from the FDA, but I believe there is nothing other than the information I posted here.
You can find all the labeling guidance on the FDA
Labeling Requirements web page.
I hope that someone else can add some more to this if there is something that I might have missed or that I'm not aware of.
W
wrodnigg
I can just give you an european view:
It depends, if the parts of the kit are "adulterated" according to their primary intended use.
If the products are used as intended in the original product, no further labeling is needed, the original labeling must not be removed or changed.
If the intended use of at least one part of the kit differs from the original one, the whole kit is a new product and the repacker becomes a manufacturer.
The concept of "finished device" does not imply that a device when
reaching the final user is already in a state ready for use. Prior to use
further preparatory processing, preparation, configuration,
installation, assembling, adaptation or fitting to the needs of the user
or patient may be required. Examples :
- sterilisation of medical devices supplied non-sterile
- assembling of systems
- configuration of electronic equipment
- preparation of a dental filling
- fitting of contact lenses
- adaptation of prosthesis to the needs of the patient.
The aforementioned activities are normally not manufacturers
activities if they are carried out by the final user as part of the use or
preparation for use. In this context a distinction needs to be made
between a typical professional activity performed by a healthcare
professional and processing and assembling activities done by a
specialist for such processing. In the latter case relevant activities may
become proper manufacturing or assembling activities relevant within
the meaning of articles 11 and 12 MDD.
~ghw
O
ontheopenroad
Are these custom kits or are they standard inventory items for your company? That may make a difference. I used to work for a custom tray division of a medical device manufacturer, and the way we handled depended on whether they were custom or stock.
Are you buying bulk components (the stents, etc.), or are they in an original manufacturer's package inside your package?
Do you get/have instructions for use from the OEM?
Wes Bucey
You may have missed the main theme here. The original poster is an OEM making medical devices. An affiliated kit maker is buying his components to assemble into kits for sale. He wrote about it in http://elsmar.com/Forums/showthread.php?t=17302
It appears Gert wants to make sure he and his company :ca: on labeling. It's still not clear whether ONLY Gert's products are in the kit or whether there are products from other manufacturers included in the kits.
It would seem a pretty straightforward resolution if ONLY Gert's products are included in the kit. Things get dicey when other manufacturers are involved (could there be some incompatibility between the products from different manufacturers?)
How about some more info, Gert?
G
Gert Sorensen
More info, certainly.... We deliver only part of the kits, the repacker may also supply items as well as buy items from other sources. That's the tricky part of it. The repacker have been assembling these kits for years, and we have not received any information that there is any problem regarding neither our product nor the kit itself.
So, where does our responsibility for the use of our item stop? :confused:
L Responsibility for Labeling in Multi-Location Group Document Control Systems, Procedures, Forms and Templates 5
A Purchasing Controls on Change Management of Literature/Labeling ISO 13485:2016 - Medical Device Quality Management Systems 3
S Registration of Medical Device in Hong Kong - labeling requirements Other Medical Device Regulations World-Wide 0
D Hazard symbols and labeling ISO 13485:2016 - Medical Device Quality Management Systems 1
M IEC 62304 Software changes - Minor labeling changes on the GUI IEC 62304 - Medical Device Software Life Cycle Processes 3
D Labeling requirements for reagents, etc ISO 13485:2016 - Medical Device Quality Management Systems 4
N FDA UDI - Label vs. Labeling - Does the insert need to include UDI? Other US Medical Device Regulations 1
Anonymous16-2 Labeling Question (Dietary Supplements/Food) Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
L Medical-grade power adapter labeling Other US Medical Device Regulations 2
Y Labeling difference for storage and operating conditions EU Medical Device Regulations 1
H MDD Article 12 Labeling for Class IIa Medical Device - Please Advise EU Medical Device Regulations 3
K China Medical Device Labeling requirements - Language China Medical Device Regulations 3
S Special 510(K) - Labeling change from Rx to OTC Other US Medical Device Regulations 11
N Device Labeling - Medtronic Ventilator Files (Risk Management documents) Coffee Break and Water Cooler Discussions 2
JoCam Labeling to Customer Requirements EU Medical Device Regulations 1
Z Labeling for Medical Device Samples - Surgical instruments 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
A UDI and Design Controls - Labeling change via the Design Control process 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
D Change Control - Product Name Change and its associated labeling EU Medical Device Regulations 7
D Do User Manuals (Not Device Labeling) HAVE to be in their national languages for these countries? Will English Hardcopy not do? EU Medical Device Regulations 12
S How to decide which medical device symbols go on label and labeling? US Food and Drug Administration (FDA) 5
M MDR Legacy Medical Device Labeling compliance timeline EU Medical Device Regulations 3
M Informational US FDA Draft Guidance – Breast Implants – Certain Labeling Recommendations to Improve Patient Communication Medical Device and FDA Regulations and Standards News 0
M Labeling for off-label use - Prevention of off-label use of a medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
M Informational US FDA Final Guidance – Coronary, Peripheral, and Neurovascular Guidewires – Performance Tests and Recommended Labeling Medical Device and FDA Regulations and Standards News 0
M Informational US FDA Final Guidance – Intravascular Catheters, Wires, and Delivery Systems with Lubricious Coatings – Labeling Considerations Medical Device and FDA Regulations and Standards News 0
P UDI labeling for power module IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
A Labeling - 2017/745 - EUMDR Requirements and Trademark Issues EU Medical Device Regulations 0
DitchDigger UDI, Labeling Accessories, General Insanity, Etc. US Food and Drug Administration (FDA) 1
K Labeling requirements - Date of manufacture for software devices EU Medical Device Regulations 5
E Annotated US Prescribing Labeling US Food and Drug Administration (FDA) 0
T Product Labeling - Modified the name of one of our products EU Medical Device Regulations 5
M Informational US FDA draft guidance – Testing and Labeling Medical Devices for Safety in the Magnetic Resonance (MR) Environment Medical Device and FDA Regulations and Standards News 0
T Labeling Requirements - Creams in tubes EU Medical Device Regulations 3
K EU MDR Art. 22 - Device + insertion pack - User manual and Labeling EU Medical Device Regulations 4
M MDR 2017/745 sterile and non-sterile labeling EU Medical Device Regulations 12
M Professional Use Medical Software French Labeling for Canada -- Not Considered Medical Device Canada Medical Device Regulations 2
CPhelan Labeling template on QMS - Kitting and labeling documentation Manufacturing and Related Processes 9
G FDA UDI markings "label" vs. "labelling" (labeling)? US Food and Drug Administration (FDA) 4
G How is the labeling of units cleared for both Rx and OTC uses handled? US Food and Drug Administration (FDA) 4
E FDA Supplier Labeling Requirements ISO 13485:2016 - Medical Device Quality Management Systems 1
L Class IIa Medical Device, CE Mark & 510k - Labeling Content CE Marking (Conformité Européene) / CB Scheme 4
M Scrubbed version of a labeling procedure that is MDR compliant wanted EU Medical Device Regulations 3
M UDI labeling MDR requirement for reusable medical device accessories EU Medical Device Regulations 12
T Change to Country of Origin on Medical Device Labeling Other Medical Device Regulations World-Wide 0
L EU-MDR-Annex I - 10.4 (CMR & Endocrine-disrupting properties) - Labeling Content EU Medical Device Regulations 5
I Medical Device Packaging Multilingual Labeling EU Medical Device Regulations 3
T FDA proposed labeling standalone software cloud based US Food and Drug Administration (FDA) 4
M Putting AR's name and address on label - Authorized Representative Labeling Requirements CE Marking (Conformité Européene) / CB Scheme 0
N IFUs with promotional labeling Other US Medical Device Regulations 0
I Requirements for different Labeling Levels 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2