Hi All,
I just wondered what you generally do when the lab environmental conditions are outside of limits.
We immediately stop work and then do not restart until the lab has stabilized back within limits.

Are there any documents that give guidance on the suspension and resumption of work? Anything that has recommended stabilization periods?
 
Elsmar Forum Sponsor

dwperron

Trusted Information Resource
#2
Unfortunately, none that I know of.

NCSLI RP-7 is a guide to setting up a Metrology lab. They recommend a 20°C area for dimensional calibrations and 23°C for other calibrations. They recommend relative humidity set to 0%, not to be allowed below 20% for ESD purposes.
These are the basic lab environmental requirements in the business, most labs model these parameters.
I have always set my lab tolerances based on my standards, keeping them in the acceptable environmental range so that I don't need to derate the tolerances on them. For the dimensional lab use usually keep the specification as tight as your HVAC system can maintain temperature. That cuts down on correction calculations for temperature. Stabilization is also based on the specifications of my standards.
 
#3
Hello,
New to the site, but as far I know there is no set standard. When one of our customer labs has a fault we usually approach it depending on How far the lab went out of spec and for how long. With CMM Labs it depends on the size of the CMM and granite. Once again depending on the equipment in the lab. Most of our Labs Designed to run +/- .5c and we design the system to shut down anything outside of a +- 1c for repair or inspection. If you have a large piece of granite that is set in 24c temperature for the weekend we would recommend 72hr soak in 20c before you start measurements. Many variables, no set standard that I know of unless it comes from your equipment manufacturer.
 

Steve Prevette

Deming Disciple
Staff member
Super Moderator
#4
Generally speaking, my experience is yes, they would stop work and figure out what is going on and fix it. My college roommate is now a Chief Toxicologist and always complaining the HVAC gives him fits. Where I work, if any work was accomplished on a test and it had not been noticed, we would likely write an NCR to track resolution of did it affect product.
 
Thread starter Similar threads Forum Replies Date
T How to meet the requirements of Environmental Conditions in a Laboratory General Measurement Device and Calibration Topics 11
H Analytical Environmental Laboratory ISO 17025 Registration Scope ISO 17025 related Discussions 3
B As-9100 - Environmental controls of our QC laboratory Imported Legacy Blogs 1
D Environmental standards for Gage and CMM laboratory General Measurement Device and Calibration Topics 1
R Laboratory Environmental Controls - Temperature and Humidity General Measurement Device and Calibration Topics 5
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
B Laboratory KPIs ISO 17025 related Discussions 13
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
T Laboratory Verification after validation ISO 17025 related Discussions 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
D Using Laboratory Notebooks in R&D and Design and Development ISO 13485:2016 - Medical Device Quality Management Systems 3
G ITAR Registration for Metrology Laboratory? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
I Laboratory Fridge / Freezer Inventory - Requirement(s) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
D Definition of "Critical Laboratory Equipment"? General Measurement Device and Calibration Topics 1
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 15
DMLqms Material for laboratory use or Medical Device? EU Medical Device Regulations 2
N IATF 16949:2016 7.1.5.3.2 External Laboratory - How to approve the Testing Laboratory without accreditation scope IATF 16949 - Automotive Quality Systems Standard 2
A Single use non-sterile syringe used in the oral cavity - Laboratory test advice US Food and Drug Administration (FDA) 7
B Accredited laboratory or not under MDD 93/42/EEC IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
D Does every piece of equipment used in a laboratory need to have an IQ protocol written and executed? ISO 13485:2016 - Medical Device Quality Management Systems 1
Casana IATF 16949 7.1.5.3.2 External Laboratory - On Site Calibration IATF 16949 - Automotive Quality Systems Standard 8
A Information on good laboratory practices (GLP) for the non-medical or food industry wanted Manufacturing and Related Processes 1
D Internal audit forms or checklists for a medical/veterinary laboratory General Auditing Discussions 5
N IATF 16949 7.1.5.3.2 External Laboratory - Calibration of Nikon machine in factory IATF 16949 - Automotive Quality Systems Standard 24
Q Calibration laboratory scope - Include software such as Gagetrak or Mini tab used? General Measurement Device and Calibration Topics 3
B IATF 16949 7.1.5.3.2 FAQ #7 Audit Finding - External Calibration Laboratory IATF 16949 - Automotive Quality Systems Standard 7
P Procedure for maintaining Confidentiality And Impartiality in a Laboratory Quality Management System (QMS) Manuals 2
sswaim IATF 16949 7.1.5.3.2 External Calibration Laboratory Requirelents IATF 16949 - Automotive Quality Systems Standard 15
J Good Laboratory Practice: online training course recommendations? US Food and Drug Administration (FDA) 0
E 2 participants in inter-laboratory comparison tests ISO 17025 related Discussions 3
T MDR on Good Laboratory Practice (GLP) - Pg 110, section 6.1 (b) EU Medical Device Regulations 1
D Laboratory Manual ISO/IEC 17025 Example wanted ISO 17025 related Discussions 2
D Help identify a Nationally Recognized Testing Laboratory (NRTL) (UL) certified lab General Measurement Device and Calibration Topics 3
L How to manage a Metrology Department or Calibration Laboratory General Measurement Device and Calibration Topics 3
J IATF 16949 Calibration Laboratory Scope IATF 16949 - Automotive Quality Systems Standard 22
H Internal laboratory scope requirements - IATF 16949 Cl. 7.1.5.3.1 IATF 16949 - Automotive Quality Systems Standard 1
S RSL (Remote Site Location) Laboratory - Does it need to be ISO 17025 IATF 16949 - Automotive Quality Systems Standard 5
S Jewelry vs. Normal Laboratory Balances - Accuracy and calibration General Measurement Device and Calibration Topics 2
rob73 Requirement for independent certification of device using an ISO 17025 laboratory EU Medical Device Regulations 2
G Calibration Laboratory Dimensional Temperature Question General Measurement Device and Calibration Topics 2
G ISO 17025 Calibration Laboratory Assessor's Common Questions and Procedures ISO 17025 related Discussions 11
H Need to Start an R&D Laboratory, what to do? Design and Development of Products and Processes 6
B IATF 16949 - 7.1.5.3.1 Internal calibration laboratory requirements IATF 16949 - Automotive Quality Systems Standard 9
A Pharmaceutical Manufacturer's obligations - Laboratory Test instruments General Measurement Device and Calibration Topics 0
S HVAC Thermostat for Calibration Laboratory General Measurement Device and Calibration Topics 8
V Temperature/humidity requirements in internal metrology laboratory General Measurement Device and Calibration Topics 3
C Looking for Testing Laboratory for Aluminum Products - CE Marking General Measurement Device and Calibration Topics 5
S Looking for a Threaded Ring Gage Calibration Laboratory General Measurement Device and Calibration Topics 13
J Internal Laboratory Scope and Measurement Uncertainty (MU) IATF 16949 - Automotive Quality Systems Standard 1
B IATF 16949 Cl. 7.1.5.3.2 - External Laboratory Sub-Supplier Requirements IATF 16949 - Automotive Quality Systems Standard 2

Similar threads

Top Bottom