Laboratory Equipment Quality System - Calibration, Maintenance and Inspection
AnaMariaVR2
Could anybody share your experience about what kind of calibration, maintenance and inspection is required for laboratory equipment for a small biotech company.
under ISO 9001:2008 clause 7.6 "Control of monitoring and measuring equipment".
What quality systems should I develop, evaluate, apply and/or monitor in order to have a robust quality system that controls equipment in our facility?
Does anybody have any ideas or experience on this area? I'm doing this from scratch. I started with the equipment database, wrote a general equipment SOP and now writing specific equipment/instrumentation SOPs. Now, what else?:confused:
What approach would Elsmar Cove recommend?
Should I develop this system based on the 6 activities that ISO must document? As it follows:
4.2.3 Control of Documents
4.2.4 Control of Records
8.2.2 Internal Audits
8.3 Control of NC products
8.5.2 Corrective Action
8.5.3 Preventive Action​
Thanks in advance to all!
:thanx:
Ana
Ajit Basrur
Re: Equipment Quality System
Hi Ana,
Firstly, I have moved your question to the FDA section since it focuses on a Biotech organization and you could get more focused answers :)
All the laboratory equipment / instruments are Qualified as per IQ, OQ and PQ before they are put into use. On a regular basis, these are calibrated and maintained as per supplier recommendations, and could vary from one instrument to another based on the usage and complexity.
With regards to the 6 procedures of ISO 9001, my 2 cents are -
4.2.3 Control of Documents - yes, these are in the form of Work Instruction on the usage of the instrument / equipment.
4.2.4 Control of Records - whenever, there is a revision, these WIs would be superceded.
8.2.2 Internal Audits - yes
8.3 Control of NC products - as part of Laboratory Investigations.
8.5.2 Corrective Action - built in Lab Investigation
8.5.3 Preventive Action - built in Lab Investigation
Take a look at the attachment from ISO for detailed info on Clause 7.6
U
Umang Vidyarthi
:thanx:
Ana
Ajit has provided sound advise. There is another thread concurrently going on similar lines The best way to quickly review the quality situation, and another here. You may benefit from the suggestions provided therein.
Umang :D
AnaMariaVR2
Thank you Ajit & Umang for your prompt response!:)
...& thanks for suggesting & moving the thread to the FDA section. I was unsure where to posted since yes FDA regulations apply but we are also pursuing ISO certification. I appreciate both of your help & insight.
:thanks:
Ana
U
Umang Vidyarthi
:thanks:
Ana
You are Welcome AnaMaria:bigwave:
Umang :D
yodon
A little late to the party, but...
One thing that's apparently often overlooked in procedures is what to do if you find something that is out of calibration. You will need to assess the impact and take actions accordingly. This means that anything that was measured using the equipment since the last known calibration date is suspect.
So think beyond the procedures to keep things in calibration when you're developing the procedures.
AnaMariaVR2
Thank you Yodon...don't worry, you are not too late for the party...the party is on & goes on at ELSMAR!:notme:
You are right, if there is no calibration schedule, inspections or routine preventive maintenance going on, how can I ensure that the data generated by those instruments is real?! And the impact to studies that used that uncalibrated equipment?
Maintenance, calibration, inspection and even simple cleaning activities are so crucial to prove that equipment is fit for use.
I'll follow your suggestion and I'll add a procedure to follow when equipment is out of calibration.:agree1: Good point :agree1:
:thanx:
If you have more suggestions for me, I have a patient ear or better saying, a grateful eye to read it in this forum.
Does anybody has an equipment malfunctioning or out of calibration SOP?
BradM
I guess since the others covered the big stuff, I'll hit some of the general basic stuff.
First, give some thought to what you are purchasing. Buy just what you need in terms of accuracy and use. Try to stay as consistent as possible.
Develop some unique, high quality identification tags and start from the beginning correctly identifying the instruments.
Develop some kind of master record for each instrument. It should have the make, model, serial, procedure, required tolerance, test points, etc. on it. This will always be a good reference to specify why you purchased the item, who is going to use it, and what it is to be used for.
Also, develop a good calibration sticker, a warning/do not use sticker, and possibly, a limited calibration sticker.
I might say also add a For Reference Only sticker. But honestly, if you're small, if you don't calibrate it, you can probably get rid of it.:D
I might spend the money calibrating instruments in a smaller interval, to assure you minimize the out of tolerance conditions. Once stability is established, you can open it up.
Nothing really earth shattering that I've said.:D Just start from the beginning with good documentation and labeling of your equipment/instrumentation.
D Definition of "Critical Laboratory Equipment"? General Measurement Device and Calibration Topics 1
D Does every piece of equipment used in a laboratory need to have an IQ protocol written and executed? ISO 13485:2016 - Medical Device Quality Management Systems 1
S Calibration of Laboratory Equipment in accordance with the manufacturer's procedure ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
M ISO17025 Testing Laboratory Equipment Calibration Requirements ISO 17025 related Discussions 4
L How to set up a basic electronic test equipment calibration laboratory General Measurement Device and Calibration Topics 3
V For Analytical Laboratory Equipment, is PQ = Calibration Criteria? Qualification and Validation (including 21 CFR Part 11) 2
M New Calibration Laboratory Set Up Measurement Parameters and Equipment General Measurement Device and Calibration Topics 4
J TS 16949 requirements for an internal laboratory equipment list IATF 16949 - Automotive Quality Systems Standard 2
R Comparing Laboratory Scope of Accreditation against internal equipment General Measurement Device and Calibration Topics 3
C Does FDA cGMP apply to Laboratory Equipment? ISO 13485:2016 - Medical Device Quality Management Systems 2
6 Fuel Testing Laboratory Test Equipment - Methods and Calibration General Measurement Device and Calibration Topics 2
T Requirements for marking Laboratory Equipment Electrical Products Manufacturing and Related Processes 3
6 Acceptance Criteria For pH Meters and other Laboratory Equipment Measurement Uncertainty (MU) 16
N General Microbiological Laboratory Layout and Required Equipment General Measurement Device and Calibration Topics 6
J Equipment required for a Dimensional Calibration Laboratory Document Control Systems, Procedures, Forms and Templates 1
AnaMariaVR2 Need Laboratory Equipment Maintenance Program Training Material US Food and Drug Administration (FDA) 3
S Standard Equipment List for Stocking a Dimensional Metrology Laboratory General Measurement Device and Calibration Topics 8
S Calibration of equipment in an A2LA accredited testing laboratory General Measurement Device and Calibration Topics 7
B IQ, OQ, PQ - Laboratory Equipment Qualifications? (eg. incubator) General Measurement Device and Calibration Topics 8
S Calibrating Biomed equipment in an ISO 17025 accredited calibration laboratory ISO 17025 related Discussions 5
G PMEL (Precision Measurement Equipment Laboratory) Calibration Course General Measurement Device and Calibration Topics 1
G PMEL (Precision Measurement Equipment Laboratory) Calibration Course Professional Certifications and Degrees 10
Antonio Vieira 7.6 Calibration - We can?t find any laboratory to calibrate our measuring equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
D 17025 - I am looking for someone to come in and certify our laboratory equipment ISO 17025 related Discussions 1
L Measuring Equipment, Which Manufacturers? Buying All New Laboratory Gages - Devices General Measurement Device and Calibration Topics 4
D Test Equipment at Work Station - Are Operator Daily Checks part of Laboratory Scope? General Measurement Device and Calibration Topics 7
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
B Laboratory KPIs ISO 17025 related Discussions 13
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
T Laboratory Verification after validation ISO 17025 related Discussions 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
D Using Laboratory Notebooks in R&D and Design and Development ISO 13485:2016 - Medical Device Quality Management Systems 3
G ITAR Registration for Metrology Laboratory? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
S Laboratory Environmental Conditions Outside of limits ISO 17025 related Discussions 4
I Laboratory Fridge / Freezer Inventory - Requirement(s) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 16
DMLqms Material for laboratory use or Medical Device? EU Medical Device Regulations 2
N IATF 16949:2016 7.1.5.3.2 External Laboratory - How to approve the Testing Laboratory without accreditation scope IATF 16949 - Automotive Quality Systems Standard 2
A Single use non-sterile syringe used in the oral cavity - Laboratory test advice US Food and Drug Administration (FDA) 7
B Accredited laboratory or not under MDD 93/42/EEC IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
Casana IATF 16949 7.1.5.3.2 External Laboratory - On Site Calibration IATF 16949 - Automotive Quality Systems Standard 8
A Information on good laboratory practices (GLP) for the non-medical or food industry wanted Manufacturing and Related Processes 1
D Internal audit forms or checklists for a medical/veterinary laboratory General Auditing Discussions 5
N IATF 16949 7.1.5.3.2 External Laboratory - Calibration of Nikon machine in factory IATF 16949 - Automotive Quality Systems Standard 24
Q Calibration laboratory scope - Include software such as Gagetrak or Mini tab used? General Measurement Device and Calibration Topics 3
B IATF 16949 7.1.5.3.2 FAQ #7 Audit Finding - External Calibration Laboratory IATF 16949 - Automotive Quality Systems Standard 7
P Procedure for maintaining Confidentiality And Impartiality in a Laboratory Quality Management System (QMS) Manuals 2
sswaim IATF 16949 7.1.5.3.2 External Calibration Laboratory Requirelents IATF 16949 - Automotive Quality Systems Standard 15
J Good Laboratory Practice: online training course recommendations? US Food and Drug Administration (FDA) 0
E 2 participants in inter-laboratory comparison tests ISO 17025 related Discussions 3