I have been involved in the quaification of a good number of fridges and freezers in 2011 for a medical device manufacturer.
There are some standard tests that need to be covered, many of which have been mention earlier.
Door open alarms, door open recovery tests, power failure, recovery tests, how the unit responds to a power failure event i.e. are the settings maintained. Alarms (high and low)
In terms of mapping, the internal volume of the unit will determine the number of points that need to be logged. Typically 9 points!!!
I tend to do the empty map as part of the OQ and the full map as part of the PQ.
I have included two attachments that I have found very useful.
One is from the Irish National Metrology lab, the other is a white paper from a US temperature logging vendor (which I have no affiliation with).
Hope this is of some use.
Here is the link to the previously attached document: CEN-LSH-8 Steps to Executing the Perfect Thermal Validation Project-B211103EN-A Low.pdf
Pat