Hi,
I hava a question about the measuring device, MD according to figure 12, when measuring with a non-frequency-weighted device.
Clause 8.7.3 Allowable values in IEC 60601-1 ed. 3.1 paragraph e) "Additionally, regardless of waveform and frequency, no leakage current shall exceed 10 mA r.m.s. in normal condition or in single fault condition when measured with a non-frequency-weighted device."
I am using a Fluke ESA612, electrical safety analyzer, to measure leakage currents wich includes the MD according to figure 12. However this has no setting for measuring non-frequency-weighted currents so I have to do this with another instrument. I assume that the MD that I need to use together with a true r.m.s. measuring DVM then only includes the resistor R2, 1000 ohm, without R1 and C1.
Am I assuming correct?
I hava a question about the measuring device, MD according to figure 12, when measuring with a non-frequency-weighted device.
Clause 8.7.3 Allowable values in IEC 60601-1 ed. 3.1 paragraph e) "Additionally, regardless of waveform and frequency, no leakage current shall exceed 10 mA r.m.s. in normal condition or in single fault condition when measured with a non-frequency-weighted device."
I am using a Fluke ESA612, electrical safety analyzer, to measure leakage currents wich includes the MD according to figure 12. However this has no setting for measuring non-frequency-weighted currents so I have to do this with another instrument. I assume that the MD that I need to use together with a true r.m.s. measuring DVM then only includes the resistor R2, 1000 ohm, without R1 and C1.
Am I assuming correct?