Leakage current measurement with non-frequency-weighted device

M

maxruben

#1
Hi,

I hava a question about the measuring device, MD according to figure 12, when measuring with a non-frequency-weighted device.

Clause 8.7.3 Allowable values in IEC 60601-1 ed. 3.1 paragraph e) "Additionally, regardless of waveform and frequency, no leakage current shall exceed 10 mA r.m.s. in normal condition or in single fault condition when measured with a non-frequency-weighted device."

I am using a Fluke ESA612, electrical safety analyzer, to measure leakage currents wich includes the MD according to figure 12. However this has no setting for measuring non-frequency-weighted currents so I have to do this with another instrument. I assume that the MD that I need to use together with a true r.m.s. measuring DVM then only includes the resistor R2, 1000 ohm, without R1 and C1.

Am I assuming correct?
 
Elsmar Forum Sponsor
M

MedMartin

#2
Hi,

Correct, you can measure the voltage on R2 only.
We measure during development using the resistor R2 and monitor the waveform with an oscilloscope and check the measurement using the current measurement function of a calibrated true-RMS multimeter.

Best regards,
Martin
 
#3
On the same subject and paragraph 8.7.3 e) do the words "no LEAKAGE CURRENT" apply to all leakage tests under 8.7.3? LEAKAGE CURRENT includes EARTH LEAKAGE CURRENT, TOUCH CURRENT, and PATIENT LEAKAGE CURRENT by definition. Therefore in addition to applicable measurements made with the MD from Figure 12, should all of the same measurements be made a second time with the non-frequency weighted device? If I understand correctly EARTH LEAKAGE CURRENT (8.7.4.5), TOUCH CURRENT (8.7.4.6), and PATIENT LEAKAGE CURRENTS (8.7.4.7 a, 8.7.4.7 b, 8.7.4.7 c, 8.7.4.7 d, 8.7.4.7 h) should be performed twice, once with the MD from figure 12 and a second time with a non-frequency-weighted device. Am I interpreting that correctly?
 

Benjamin Weber

Trusted Information Resource
#4
Yes, you are correct. All leakage currents should be done twice - with and without frequency weighting. Only if you can show, that there are no higher frequency components >1kHz (e.g. by use of an oscilloscope), you can skip the unweighted measurement.
 

Benjamin Weber

Trusted Information Resource
#5
Hi,

Correct, you can measure the voltage on R2 only.
We measure during development using the resistor R2 and monitor the waveform with an oscilloscope and check the measurement using the current measurement function of a calibrated true-RMS multimeter.

Best regards,
Martin
The standard says you should measure the voltage drop over the MD and converting them to the corresponding currents by dividing the voltage with 1kOhm. This is acutally current measurement with a quite high shunt resisitance.

I am not sure if measuring the currents directly with the current measurement function will give equivalent results.
 

Benjamin Weber

Trusted Information Resource
#6
Hi,

I hava a question about the measuring device, MD according to figure 12, when measuring with a non-frequency-weighted device.

Clause 8.7.3 Allowable values in IEC 60601-1 ed. 3.1 paragraph e) "Additionally, regardless of waveform and frequency, no leakage current shall exceed 10 mA r.m.s. in normal condition or in single fault condition when measured with a non-frequency-weighted device."

I am using a Fluke ESA612, electrical safety analyzer, to measure leakage currents wich includes the MD according to figure 12. However this has no setting for measuring non-frequency-weighted currents so I have to do this with another instrument. I assume that the MD that I need to use together with a true r.m.s. measuring DVM then only includes the resistor R2, 1000 ohm, without R1 and C1.

Am I assuming correct?
That is why test labs usually use custom-made equipment for this measurement.
 

Peter Selvey

Leader
Super Moderator
#8
There are a few technical side points here: measurements above say 10kHz are unlikely to be practically accurate nor legally traceable, even if using an oscilloscope, even if using an accredited test lab. There are no off the shelf systems, including DMMs, scopes that are OK up to 1MHz, and even the parts used to create the filter are unlikely to be stable up to 1MHz, e.g. stray capacitance, capacitor ESR, inductance and other high frequency effects such as leakage paths through the equipment to ground, common mode effects (which are a nightmare at higher frequencies). With some care measurements up to 100kHz could be traceable, but above 100k is not realistic. Fortunately, it's rare that higher frequency measurements come anywhere near the limits, so it's not a critical issue.

I personally prefer a while box approach where the source of the leakage is understood, including higher frequency components. For example if the ECG leads also have a 60kHz ac signal for measuring respiration, the test lab (guided by the designers) should be familiar with the circuit that creates this signal, the expected currents in normal and single fault condition. Blindly measuring leakage with a custom filter and Fluke 87 DMM isn't going to cut it, and potentially overlooks the single fault analysis.

Leakage associated with switching power supplies often have frequency components above mains frequency, but in my experience they are never anywhere near the limits. For example, you might measure 0.12mA earth leakage with the filter, and 0.14mA without the filter. Even if that second measurement is wildly inaccurate, it's nowhere near 10mA.
 
#9
Thanks for your response and time on this Peter. It is fortunate that higher frequency measurements rarely come near the limits. I appreciate your comments on tracing the source of the leakage to make sure it is understood and make any necessary modifications. It seems like the non-frequency-weighted measurements are rarely vastly different from the regular measurement, but still always good to check.
 

Benjamin Weber

Trusted Information Resource
#10
There are a few technical side points here: measurements above say 10kHz are unlikely to be practically accurate nor legally traceable, even if using an oscilloscope, even if using an accredited test lab. There are no off the shelf systems, including DMMs, scopes that are OK up to 1MHz, and even the parts used to create the filter are unlikely to be stable up to 1MHz, e.g. stray capacitance, capacitor ESR, inductance and other high frequency effects such as leakage paths through the equipment to ground, common mode effects (which are a nightmare at higher frequencies). With some care measurements up to 100kHz could be traceable, but above 100k is not realistic. Fortunately, it's rare that higher frequency measurements come anywhere near the limits, so it's not a critical issue.

I personally prefer a while box approach where the source of the leakage is understood, including higher frequency components. For example if the ECG leads also have a 60kHz ac signal for measuring respiration, the test lab (guided by the designers) should be familiar with the circuit that creates this signal, the expected currents in normal and single fault condition. Blindly measuring leakage with a custom filter and Fluke 87 DMM isn't going to cut it, and potentially overlooks the single fault analysis.

Leakage associated with switching power supplies often have frequency components above mains frequency, but in my experience they are never anywhere near the limits. For example, you might measure 0.12mA earth leakage with the filter, and 0.14mA without the filter. Even if that second measurement is wildly inaccurate, it's nowhere near 10mA.
Peter I have to disagree at some point with you ;-)

Frequency response of the measuring network at >10kHz can be an issue. But if you take care during the development of the network, you can achieve quite good results. So we did not just take "any" resistors or capacitors. First we chose parts with very good nominal tolerances (10 times better than recommended in fig. 12, IEC 60601-1). Then we purchased a certain number (10 to 20) for each part and selected the best fitting manually. Additionally we did not take just one single part per component. We actually used multiple single resistors/capacitors to achieve the required values of R1, R2 and C1 - let's say 4 x 2.5 kOhms in series for R1. By clever selection of each single part, you can actually level out the remaning deviations from the nominal values. This leads to very good results also at higher frequencies during calibration at an accredited callibration lab (<5% overall deviation at 1 MHz).

Regarding the actual voltage measuring device: You are correct, standard DMMs are usually limited to 100 kHz. But oscilloscopes have much higher bandwiths. We use a fully isolated/floating scope with 60 MHz bandwith. We can assume, that the frequency response up to 1 MHz is linear and not an issue. But AC measurement uncertainty is bit of a problem. All scopes are specifed by the bandwith (-3 dB deviation) and mostly by the DC gain accuracy. But you will not find a value for AC accuracy. So the conservative approach is to use the -3 dB (appr. -30%) also for low frequencies. But you can try to perform AC callibration for the whole measurement chain (network + scope) at discrete freqnuencies, e.g. 50Hz, 60Hz, 100Hz, 500Hz, 1kHz, 500kHz and 1MHz.

Regarding those aspects it is possible to have a good measurement setup also for higher frequencies.
 
Thread starter Similar threads Forum Replies Date
B Leakage current measurement without isolating transformer IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E IEC 60335-1 leakage current measurement Other ISO and International Standards and European Regulations 3
A Medical Device Leakage Current Measurement with External Power Supply IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B Leakage Current AC-DC discrimination per cl. 8.7.3 b) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Earth (ground) leakage current routine test IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Issues with leakage current testing for a CLASS II device since no PE and FE is present IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W Regarding Requirement for Hi pot & Leakage Current test Station Set-up IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
A Earth leakage current test for Permanently installed Large scale MEE IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
A Touch current in single fault conditions test and earth leakage current in normal conditions test, are they really different tests? IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
T Tests IEC 60601-1 - Grounding Resistance, Leakage Current and Dielectric Rigidity in an automated way IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
S Calibration - Isolated current meter - Measuring patient auxiliary leakage current Calibration and Metrology Software and Hardware 4
S Leakage Current test and Dielectric strength for Battery operated ECG device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
P Humidity Pre-Conditioning before leakage current and dielectric strength tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
W How can I test the Patient Leakage Current of an Infusion Pump? IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
A Accuracy Calculation of Leakage Current for an Electrical Safety Analyzer General Measurement Device and Calibration Topics 5
J Exceeding the DC Leakage Current Limit IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
L Building your own Leakage Current Tester IEC60990 Other US Medical Device Regulations 4
M Leakage Current Measuring Instrument - IEC 60601-1 Requirement IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R IEC 60601-2-49 - Touch Current (Enclosure Leakage Current) requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Why DC limits more stringent than AC limits for Leakage Current Test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J Pre-Conditioning Leakage Current IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
M Operator/Patient Leakage Current IEC 60601 - Battery Charger IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
P Leakage Currents and Patient Auxiliary Current under IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Leakage current from battery - IEC 60601-1 Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
U Leakage test Standard for Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
N 60601-2-2 Ed 6 Clause 201.8.8.3.102 - HF instrument leakage IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M ECG lead leakage currents - How to specify ECG leads during electrical safety testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
J Measuring Leakage Currents with IEC 60601 defined Measurement Device (MD) IEC 60601 - Medical Electrical Equipment Safety Standards Series 15
M ISO 13485 - Are Earth Leakage Tests Required? ISO 13485:2016 - Medical Device Quality Management Systems 2
D Patient Leakage ISO 7176-14 vs. IEC 60601-1:2005 Reporting Requirements Other US Medical Device Regulations 1
M IEC60601-1 Battery Powered Body Worn Device Leakage Currents IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
B Defining Expected Oxygen Leakage for Safety Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Understanding DC Patient Leakage Failure from 0.3VDC IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Whether some paths exist for leakage to get outside the LCD? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
C USB, Clearances and Earth Leakage - Medical Devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
K Calibration method for Torque Wrench & AC gas leakage Tester General Measurement Device and Calibration Topics 4
K Alternatives to Metal Foil for touch current measurement of enclosures made of insulating material. IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Help Understanding Current Role; Distributor, Manufacturer, or Importer EU Medical Device Regulations 4
C IEC60601-1: Impedance and current-carrying capability (8.6.4) IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
F IEC 60601-2-54:2009 Accuracy of CURRENT TIME PRODUCT IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B Precision Current Loop for Current Probe Calibration Reliability Analysis - Predictions, Testing and Standards 2
N Auxiliary current limit with coil cell powered device IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
T How can I determine the current level of exposure to talc in the workplace? Occupational Health & Safety Management Standards 6
H Prodigit DC Electronic Load - Drawing current but showed 0 on the screeen Manufacturing and Related Processes 0
D Add new device to current license Canada Medical Device Regulations 3
W FMEA - Current control and occurrence rating FMEA and Control Plans 3
S Forced ServiceNow validation - No change in our current user and functional requirements IT (Information Technology) Service Management 4
W Need for current design or process control FMEA and Control Plans 2
E Best way to keep up with pharmaceutical quality knowledge and current affairs Service Industry Specific Topics 1
T Iec 60601 Impedance and current-carrying capability IEC 60601 - Medical Electrical Equipment Safety Standards Series 1

Similar threads

Top Bottom