Hello all.
I have been engaging with chinease laboratory to test our medical RF device and come to a dead lock with them pertaining their interpretation of the patient leakage test , basically on how it should be done.
Our device has a BF type B polar RF .
Per standard, we need to connect the bpolar handpiece to a 200 ohms load, and measure the leakage current flowing to earth via another 200 ohms resistor.
We have done this test both in our internal lab and an official lab in our country and passed, by using a floating oscilloscope measuring voltage across the 200 ohms resistor connected in series to ground., hence found the V and calculated the I from ohms law.
The chinease however, insist, that since in standard it says HF current meter, they must use a current probes, like clamps...and with their method we dont pass in fractions...for example, if we measure 48mA they measure 53mA.
The standard says below 50 mA is OK.
All of our explanations to them has been blocked so far.
My questions:
1. Is there an addendum or some official paper that I can use to prove to them that the illustrations or wordings in the standard are not one sided and mandatory and it may happen that using ohms law is OK?
2. Any professional paper explaining that using o scope to measure a 1mhz signal is much more accurate than a current probe, that itself is converting magnetic field to voltage internally and then converts it to current has its tolerances as well?
3. Is there any formal paper dealing with acceptable tolerances within such measurements due to all tolerances comprising the dut? The standard stipulates that under 50 mA is acceptable, but I am sure there is also an acceptable tolerance ration
4. Any other suggestions on how to mitigate the situation, would be highly appreciated.
Udi
I have been engaging with chinease laboratory to test our medical RF device and come to a dead lock with them pertaining their interpretation of the patient leakage test , basically on how it should be done.
Our device has a BF type B polar RF .
Per standard, we need to connect the bpolar handpiece to a 200 ohms load, and measure the leakage current flowing to earth via another 200 ohms resistor.
We have done this test both in our internal lab and an official lab in our country and passed, by using a floating oscilloscope measuring voltage across the 200 ohms resistor connected in series to ground., hence found the V and calculated the I from ohms law.
The chinease however, insist, that since in standard it says HF current meter, they must use a current probes, like clamps...and with their method we dont pass in fractions...for example, if we measure 48mA they measure 53mA.
The standard says below 50 mA is OK.
All of our explanations to them has been blocked so far.
My questions:
1. Is there an addendum or some official paper that I can use to prove to them that the illustrations or wordings in the standard are not one sided and mandatory and it may happen that using ohms law is OK?
2. Any professional paper explaining that using o scope to measure a 1mhz signal is much more accurate than a current probe, that itself is converting magnetic field to voltage internally and then converts it to current has its tolerances as well?
3. Is there any formal paper dealing with acceptable tolerances within such measurements due to all tolerances comprising the dut? The standard stipulates that under 50 mA is acceptable, but I am sure there is also an acceptable tolerance ration
4. Any other suggestions on how to mitigate the situation, would be highly appreciated.
Udi