Least count thumb rule - Stone walling SPC

  • Thread starter Thread starter cyberspider
  • Start date Start date
C

cyberspider

While performing capability analysis, we hit a stone wall.

Problem is, as per thumb rule of SPC, measurement system should have [(USL-LSL)/10] as the least count or resolution.

We are operating process with very narrow tolerance eg. in some cases [USL-LSL] = 0.1 and 0.1/10 = 0.01. In most cases, we found that the required resolution (0.01 in this case) is inadquate.

Now what to do ?

Any opinion or innovative ideas ?
 
Elsmar Forum Sponsor
cyberspider said:
While performing capability analysis, we hit a stone wall.

Problem is, as per thumb rule of SPC, measurement system should have [(USL-LSL)/10] as the least count or resolution.

We are operating process with very narrow tolerance eg. in some cases [USL-LSL] = 0.1 and 0.1/10 = 0.01. In most cases, we found that the required resolution (0.01 in this case) is inadquate.

Now what to do ?

Any opinion or innovative ideas ?
Do you mean "inadequate" or "not possible"?
If it's not possible, there's nothing you can do about it, so you have to do the best you can. What are you measuring, and what are you using to measure it?
 
Clarification

Measurement System least count is inadquate.

It is however possible to change the system but at a premium cost and I am not in favour to go for changing the entire measurement system. Hence, I am looking for some statistical idea or strategy.

Thanks,
 
cyberspider said:
Measurement System least count is inadquate.

It is however possible to change the system but at a premium cost and I am not in favour to go for changing the entire measurement system. Hence, I am looking for some statistical idea or strategy.

Thanks,
If the measurement system lacks the requisite sensitivity, there is no magical statistical method that will help.
 
cyberspider said:
Measurement System least count is inadquate.

It is however possible to change the system but at a premium cost and I am not in favour to go for changing the entire measurement system. Hence, I am looking for some statistical idea or strategy.

Thanks,

If the system is inadequate then you shouldn't be using it for the measurement. It would seem you have no choice but to change the system to one that is suitable. The obvious solution would be to use the same system but with greater resolution. Is that possible without incurring great expense?

Dave
 
Other metrics you can use to evaluate the Measurement Process.

cyberspider said:
While performing capability analysis, we hit a stone wall.

Problem is, as per thumb rule of SPC, measurement system should have [(USL-LSL)/10] as the least count or resolution.

We are operating process with very narrow tolerance eg. in some cases [USL-LSL] = 0.1 and 0.1/10 = 0.01. In most cases, we found that the required resolution (0.01 in this case) is inadquate.

Now what to do ?

Any opinion or innovative ideas ?

Cyberspider:

What I would do before any capability analysis, and before any SPC, is a Gage R&R, calculating the following metrics:
ndc : number of distinct categories , which tells into how many different categories you can divide your process output, using your current measurement system (MS). This value should exceed 5
PTR : precision to tolerance ratio= 6*sigma(MS) / Tolerance. There is consent that this value should not exceed 30% (some authors use 5.15 in the numerator of the formula)

ndc relates the MS to the Voice of the Process.
PTR relates the MS to the Voice of the Customer (specs)

I would not use the resolution as the only mean to decide on the adequacy of the MS, because there could be other sources of uncertainty that make the measurement system even less capable.

If you already have control charts on the process, it is easy to verify the incapability of the instrument in the Range chart, the points are "discretized" in just a handful of values. This situation invalidates the usability of the control chart.

Daniel
 
Thanks

Thanks for right direction on NDC , I think , i got the direction
 
Literature quotes resolution upto 1/6 of tolerance as permissible

Keki Bhote, author of many books on quality mentions that resolution upto 1/6 of tolerance is permissible
 