SBS - the best value in QMS software

Lesser expensive to calibrate various test weights

Dongzkie

Involved In Discussions
#1
I have 25 20kg in our lab, they were all calibrated annually.
my question is, can I only request 1 20kg to be calibrated this will be my working standard then I will do the verification for the remaining 24 test weights? because I have here available scale w/c can be use to verify those remaining test weights.
is this acceptable?
 
Elsmar Forum Sponsor

Jen Kirley

Quality and Auditing Expert
Staff member
Admin
#2
Good day Dongzkie,

I am wondering what are your tolerances? Secondly, what discrimination do your scales detect? Standards - including transfer standards - are often calibrated to an accuracy notable in discrimination x10 that of the instrument they are used to verify. As an example, the scale measures in 10ths and the test weight is calibrated to accuracy in hundredths. Unless your scales measure at a larger discrimination than the test weights are calibrated to, I would not agree to use them to verify test weight accuracy.

I hope this makes sense!

If costs are a very great concern you might question whether you need 25 of these test weights. Could you manage with fewer of them?
 

Dongzkie

Involved In Discussions
#3
Jennifer,
we use 0.5kg resolution on our process. we use bench scale on the production 150kg capacity and 1 500kg floorscale. As for example I have 1 20kg calibrated result is 20.000kg As found and has a uncertainty of +/-0.001kg. I have some available scale here which i can be use, and i can set this scale to 0.001kg readability, what if when I weigh the 20kg test weight to that scale and reads correctly to 20.000kg, is it okay to use that scale to verify the other 24 20kg test weights? thank you :rolleyes:
Good day Dongzkie,

I am wondering what are your tolerances? Secondly, what discrimination do your scales detect? Standards - including transfer standards - are often calibrated to an accuracy notable in discrimination x10 that of the instrument they are used to verify. As an example, the scale measures in 10ths and the test weight is calibrated to accuracy in hundredths. Unless your scales measure at a larger discrimination than the test weights are calibrated to, I would not agree to use them to verify test weight accuracy.

I hope this makes sense!

If costs are a very great concern you might question whether you need 25 of these test weights. Could you manage with fewer of them?
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#5
In theory yes. But to do a proper job of calibrating the other weights, you should understand the effects of local acceleration of gravity, temperature, essentricity, altitude compared to the external lab, and then these and other factors must be considered in your uncertainty calculations, so that you know whether the other weights actually meet your needs.
 
J

Jeff Frost

#6
A good question at this point is what has your review of all past calibration reports indicated. If the weights are not changing between annual calibration interval why are you not extending the calibration cycle? Your calibration system should have a method to extend or reduce calibration cycle based on results of calibrations.
 
Thread starter Similar threads Forum Replies Date
SteelMaiden Murphy's Lesser Known Laws Coffee Break and Water Cooler Discussions 3
B Murphy's Lesser Known Laws Funny Stuff - Jokes and Humour 9
F Looking for a Japan MAH recommendation that is not too expensive Japan Medical Device Regulations 5
A Measuring Countersink Depth without the use of an expensive gage General Measurement Device and Calibration Topics 3
B If the inspection and analysis of raw material is very expensive? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 19
B Customer Product or Material - Machining of very expensive material AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
Wes Bucey Is FDA Approval Too Expensive? The Cost of Doing Business in the Medical Industry ISO 13485:2016 - Medical Device Quality Management Systems 12
Claes Gefvenberg Talk about expensive mishaps... Nonconformance and Corrective Action 16
S Where can I buy the TS 16949 standard? The AIAG is expensive! IATF 16949 - Automotive Quality Systems Standard 9
M Less expensive way to calibrate multimeters such as a "Fluke"? General Measurement Device and Calibration Topics 2
A Looking for a third party to calibrate our measure equipment - South Carolina General Measurement Device and Calibration Topics 6
I Can I Calibrate On Demand? General Measurement Device and Calibration Topics 9
S Calibrating our own equipment, can we? Micrometers to calibrate vernier calipers General Measurement Device and Calibration Topics 13
F Crimper Calibration - How do you Control & Calibrate your crimpers? Calibration Frequency (Interval) 13
P How to Calibrate a HP 3458A Digital Multimeter with Fluke 5080A calibrator General Measurement Device and Calibration Topics 1
B Choosing not to calibrate (IATF 16949) IATF 16949 - Automotive Quality Systems Standard 6
K New Kilogram - The process we use to calibrate the standards General Measurement Device and Calibration Topics 6
D How to Calibrate any Digital Sphygmomanometer under use for 2 years in a Hospital Forum News and General Information 1
L Re-calibrate after software update? Calibration Frequency (Interval) 3
J To Calibrate or not to calibrate electronics rack components Manufacturing and Related Processes 5
T How to Calibrate Taper Gages General Measurement Device and Calibration Topics 1
N Calibrate temperature on refrigerated Centrifuges? General Measurement Device and Calibration Topics 2
B Can we calibrate a standard pin with a gauge? General Measurement Device and Calibration Topics 2
A How to Calibrate Laser Beam Tester & Radiation Monitor General Measurement Device and Calibration Topics 3
D How to Calibrate and Verify a Heating Oven General Measurement Device and Calibration Topics 3
P How do you calibrate a VNA (Vector Network Analyzer) cal kit? General Measurement Device and Calibration Topics 4
V How to Calibrate Ring Force Gauges Service Industry Specific Topics 1
bobdoering Who says you can't calibrate a steel rule! Funny Stuff - Jokes and Humour 14
S How can I Calibrate a Biosafety Cabinet? General Measurement Device and Calibration Topics 1
C How to Calibrate a Viscosity Tester General Measurement Device and Calibration Topics 3
0 Is it possible to use a threaded ring gage to calibrate the M8 gage? Calibration and Metrology Software and Hardware 4
C How to Calibrate a Medical Infrared Thermometer General Measurement Device and Calibration Topics 3
T How do I calibrate NPT Thread Plug Gages? General Measurement Device and Calibration Topics 9
E Do you need a Certification to Calibrate Gauges? General Measurement Device and Calibration Topics 2
B Calibrate Hole Presence Detection Device? General Measurement Device and Calibration Topics 5
J What do you use to Calibrate 40" Calipers General Measurement Device and Calibration Topics 10
A Guideline to Calibrate Chamber Temp. & Pressure of a Medical Steam Sterilizer General Measurement Device and Calibration Topics 8
S How to calibrate AOI (Automated Optical Inspection) for IOPQ Qualification and Validation (including 21 CFR Part 11) 3
S Laboratory to Calibrate Objective Micrometer needed General Measurement Device and Calibration Topics 4
C Tolerance Estimation of a Load Cell I want to Calibrate General Measurement Device and Calibration Topics 9
D Procedure on How to Calibrate a External Micrometer General Measurement Device and Calibration Topics 1
B ISO/TS 16949 Section 7.6.3 - Use of Manufacturer to Calibrate IATF 16949 - Automotive Quality Systems Standard 21
W How to verify/calibrate a Neptune Water Meter General Measurement Device and Calibration Topics 4
N Laboratory that is able to Calibrate Dew Point Meter up to 60?Cdp - Needed General Measurement Device and Calibration Topics 1
C How to Calibrate and Verify Torque Multiplier Gages General Measurement Device and Calibration Topics 6
C How to calibrate or control patterns for fiberglass General Measurement Device and Calibration Topics 4
N Scale with Internal Calibration - Do I need to calibrate? How to document this? Calibration Frequency (Interval) 8
I Need help finding an Accredited Lab that can Calibrate Brunson 76-RH Survey Transits General Measurement Device and Calibration Topics 5
M Controlling Temperature and Time - Do I need to calibrate the thermostat? IATF 16949 - Automotive Quality Systems Standard 4
D How much Tolerance do I use to Calibrate a 150kgs capacity Scale? General Measurement Device and Calibration Topics 5