OK, ladies and gentleman, we've got a range of opinions on this one from our notified body, our distributor, and ourselves ranging from Class I (non-measuring), to Class IIa, and to Class IIb. What say you? Here is the device intended use:
DEVICE is intended to acquire, store, analyze, and report on ECG data collected from infants. DEVICE is intended to be used by trained operators under the direct supervision of a licensed health care practitioner in a hospital neonatal or pediatric ICU environment.

DEVICE is intended to be used for the analysis of the variability in RR Intervals (heart rate) and to report measurements of the variability of heart rate data (HRV). The HRV measurements reported by DEVICE are specialized in nature, and intended to identify periods of transient decelerations and/or reduced baseline variability in the heart rate.

DEVICE is intended to provide specialized HRV measurements and may detect arrhythmias, that when combined with other risk factors, can be used as an aid in assessing the risk of sepsis in infants. The use of the DEVICE monitoring improves survival in very low birth weight neonates.

DEVICE acquires data from a user-supplied ECG monitor, and requires a user-supplied local area network.

Installation, training, maintenance and support are to be performed by MANUFACTURER employees or contractors who have been certified by MANUFACTURER as qualified to perform such duties.

This case for Class I:
Rule 12 is the default for an Active Device for Diagnosis which does not fit into Rule 10. Page 45 of MedDEV 2.4/1 lists thermography, an active device for diagnosis, as Class I by Rule 12. On page 42, an electronic thermometer is listed as IIa under Rule 10 because it allows "direct diagnosis or monitoring of vital physiological processes".

[FONT=&quot]thermography [/FONT]
[FONT=&quot]a technique wherein an infrared camera photographically portrays the body's surface temperature, based on self-emanating infrared radiations; used as a diagnostic aid in the detection of breast tumors and the assessment of rheumatic joints; also used in the study of pain.
[/FONT]
[FONT=&quot]Miller-Keane Encyclopedia and Dictionary of Medicine, Nursing, and Allied Health, Seventh Edition[/FONT]

Hence, thermography must be not providing *direct* diagnosis, but rather indirect diagnosis (of breast cancer, for instance).

Further, thermography must be not monitoring a "vital physiological process", even though it is looking at patterns in body temperature (Note 1 on page 43 lists "body temperature as a vital physiological process".)

Similar to thermography, heart rate variability analysis is looking at the *patterns of* a vital sign in order to provide a risk assessment of sepsis, but not a direct diagnosis.

Hence, the device is Class I.


The Case for Class IIa:
By rule 10, we've got an Active Device for Diagnosis, but one where the nature of the variations is such that it could NOT result in *immediate* danger to the patient.

In the context of a Neonatal Intensive Care Unit, where babies stop breathing routinely and must be revived before they suffer results of hypoxia, harboring a case of sepsis that brews over days does not require *immediate* intervention.

Hence IIa.


The Case for IIb:
By rule 10, we've got an Active Device for Diagnosis where the nature of the variations DOES result in *immediate* danger to the patient. Sepsis is dangerous and it needs to be treated ASAP.

Hence IIb.


***
Looking forward to your opinions!
 
Last edited by a moderator:
Elsmar Forum Sponsor
S

Sarah Stec

#2
Re: Let's play... Classify My Device!!!

As a last resort, you could ask the competent authority, or have your notified body ask the competent authority. :bigwave:
 

Marc

Hunkered Down for the Duration with a Mask on...
Staff member
Admin
#3
Re: Let's play... Classify My Device!!!

True, but often getting feedback from others helps. Many times one asks "the competent authority" and one person there says one thing and another says another. Discussions here can help in understanding the different requirements so that when one speaks with someone at "the competent authority" one is as informed as possible.

To me it's sort of like calling the IRS (in the US) about a tax issue. You can speak with several different people and each often has a different "slant" on the interpretation of the rules. The more one knows about the rules, the better one can assess the differences between conflicting advice from different people in "the competent authority" when it occurs.

I would offer my opinion on this specific poll but its outside my field of expertise.
 
#4
My vote is Class IIa.

By using rule 10.
- Active device
- intended to ....monitor vital physiological process

The bit about "immediate danger" is, in my opinion, referring to cardiac arrest VF / loss of breathing etc.

But, having said that, it may not be important anyway. Depending upon your route to the CE mark (that is, which annex are you using to get approval) the difference between IIa and IIb can be minimal. If you are using Annex II (full QA) then compare paragraphs 7.2 and 7.3 - there is very little practical difference.

Getting the opinion of others can be helpful but it is your Notified Body that matters. If I remember correctly there is a committee of NBs to decide on controversial / borderline cases. If you can wait that long!
 
P

phloQS

#6
Also agree with pads38! The intend of use does not specify, that the diagnosis is used to gain information as to used for immediate treatment or something like that. For me it looks like longterm diagnosis to gain information on success of therapy or something like that. Mabe you can expand something like that in the intend of use, so you directly exclude "immediate danger".

Regards

phloQS
 

Marcelo

Inactive Registered Visitor
#7
The main rationale is on note 1 from Rule 10 on the MEDDEV:

Medical devices intended to be used for continuous surveillance of vital physiological processes in anaesthesia, intensive care or emergency care are in Class IIb, whilst medical devices intended to be used to obtain readings of vital physiological signals in routine check ups and in self-monitoring are in Class IIa.
 
Last edited:
W

wking

#8
You make an excellent point. However, let's go back to Rule 10 in the MDD. I believe the distinction between IIa and IIb lies in the word *immediate* in the following:

Active devices intended for diagnosis are in Class IIa:
...
– if they are intended to allow direct diagnosis or monitoring of vital physiological processes, unless they are specifically intended for monitoring of vital physiological parameters, where the nature of variations is such that it could result in immediate danger to the patient, for instance variations in cardiac performance, respiration, activity of CNS in which case they are in Class IIb.
So, I *believe* that the explanation of the rule in MEDDEV is based on the fact that anesthesia, intensive care, and emergency care environments nearly universally involve immediate threats to the patient, whereas routine checkups do not.

So, is the distinction between IIa and IIb based solely on the environment (ICU vs doctor's office), or is it based on the immediacy of the threat to the patient?

I would think the *intention* of the rule is the latter, but that 99% of the time, the latter and the former are the same. So that in *general*, we can use the environment to distinguish between IIa and IIb, but there are in fact some situations where it doesn't work.

Thoughts on that?

Moving further, I'm really curious as to why no one has voted for Class I. I can't see any reason why the MEDDEV would have explicitly listed thermography as a Class I by Rule 12 (while leaving temperature as a Class IIa/b (depending on environment) by Rule 10) that does not correspond to Heart Rate Variability as a Class I (while Heart Rate is a Class IIa/b).

What am I missing?


Thanks to everyone who has voted and weighed in. I think this is a really interesting topic-- I hope you do as well!
 

Marcelo

Inactive Registered Visitor
#9
I would think the *intention* of the rule is the latter, but that 99% of the time, the latter and the former are the same. So that in *general*, we can use the environment to distinguish between IIa and IIb, but there are in fact some situations where it doesn't work.
Agreed. The official distinction is on immediacy because that´s what written on the directive.

The note on the MEDDEV is guidance, however, I think the original rationale is exactly as you said.

Anyway, the way you describe your device, it seems to fit into the official and into the note as being class IIb.
 

Miner

Forum Moderator
Staff member
Admin
#10
One distinction that I see is that the thermometer is directly measuring temperature, a parameter considered a "vital physiological process". No interpretation is necessary. Thermography does not directly measure a vital physiological process. It measures indirect parameters, which then must be interpreted to make a diagnosis.

Your product is directly measuring heart rate, a vital physiological process. I believe this would move it into the Class II product. Now a classification must be made as to whether it is IIa or IIb.

Is the diagnosis of sepsis the only use to which this device will be applied? Or might it be used for situations that are clearly "immediate"?
 
Thread starter Similar threads Forum Replies Date
ScottK Let's Catch Up... what's happened to you since last June Coffee Break and Water Cooler Discussions 24
E Contacting FDA via phone to let them know a Submission is Imminent Other US Medical Device Regulations 6
S Is it alright to let the supplier do all measurements and inspection? Quality Manager and Management Related Issues 10
J How many Major Nonconformances could let the Audit be Terminated? General Auditing Discussions 9
T Let's share ideas about an inspection problem.... Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
Z Let us Discuss ISO 9001's Tomorrow ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
N TL 9000 Starter - Please let me know if anyone has worked on TL 9000 Software Quality Assurance 1
L It is fair and legal to let a customer to make an assessment audit based on other cus IATF 16949 - Automotive Quality Systems Standard 2
Hershal Congrats! Let's talk about top posters in the threads General Measurement Device and Calibration Topics 17
R Quality Management System not active for 6 months. Registration let it Lapse or Fail? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
R The Improvement in Sigma Level (from Let's fix six sigma) Six Sigma 4
T Let's fix Six Sigma! Six Sigma 197
S How do you let customers know you are certified on your literature? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
Q How to write my CV if I was let go? Career and Occupation Discussions 11
A Quality Policy - Please let me know your opinions and comments ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
tony s Let the Registrar audit the Internal Audit process Internal Auditing 45
Claes Gefvenberg Let's do age! How 'Old' will you be in 2007? A poll - Version 3 Coffee Break and Water Cooler Discussions 91
K What do you expect from a new position? Let?s compare benefits Career and Occupation Discussions 18
Hershal Football! (OK, we will even let soccer come in) - Fall 2006 Coffee Break and Water Cooler Discussions 162
H Benchmarks for Plastic Injection Molding - Let's Share! Benchmarking 18
Hershal MEMORIAL DAY 2007 - Let it be heard Coffee Break and Water Cooler Discussions 18
W Useful PC software utilities let's share some info? After Work and Weekend Discussion Topics 20
Jim Wynne A Modest Proposal: Let's Get Rid of "Quality." Philosophy, Gurus, Innovation and Evolution 15
Douglas E. Purdy 7.5.2 e) Revalidation - Let me run this by you for verification ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
Steve Prevette Let's nominate a Cove'r for ASQ Board! ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 13
Q And now, let's add software development - Medical Devices - Applying ISO9001:2000 ISO 13485:2016 - Medical Device Quality Management Systems 12
E The Chit Chat and 'Let's Meet' Thread - Meeting Other Elsmar Cove Participants Coffee Break and Water Cooler Discussions 186
E Let's do age! How 'Old' are you? A poll - Version 2 Coffee Break and Water Cooler Discussions 223
Marc Let The Lawsuits Begin! Lawsuit could set crucial Y2K precedent World News 3
M Informational EU – October 2019 update – State-of-play of joint assessments of Notified Bodies in the medical device sector Medical Device and FDA Regulations and Standards News 2
M Informational EU – State-of-play of joint assessments of Notified Bodies in the medical device sector – June 2019 Medical Device and FDA Regulations and Standards News 0
R How to SHOW the importance of checklists, through some team gaming or play? Training - Internal, External, Online and Distance Learning 37
M Informational EU – Update on state-of-play of joint assessments of Notified Bodies in the medical device sector Medical Device and FDA Regulations and Standards News 1
M Medical Device News State-of-play of joint assessments of Notified Bodies in the medical device sector EU Medical Device Regulations 0
B Where do CQI Assessment requirements come into play? IATF 16949 - Automotive Quality Systems Standard 2
Marc Over 2,300 MS-DOS games now completely free to play at Internet Archive Coffee Break and Water Cooler Discussions 0
AnaMariaVR2 3 Ways To Deal With People Who Play Office Politics Against You Coffee Break and Water Cooler Discussions 7
A Have you seen the Green Bay Packers play? Coffee Break and Water Cooler Discussions 26
C At what stage does Design Control play a role in Medical Device Development ISO 13485:2016 - Medical Device Quality Management Systems 2
Z Customer play significant role when Input or not? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
P What Video Games do you play? Coffee Break and Water Cooler Discussions 29
Marc Fowl Play on Highway Funny Stuff - Jokes and Humour 1
W Games people play...needed for basic ISO training ISO 13485:2016 - Medical Device Quality Management Systems 23
Y Russian Roulette - Interviewer suggests to play imaginary Russian Roulette with you Brain Teasers and Puzzles 9
S Where does least significant digit come into play? Measurement Uncertainty (MU) 8
K Work Station Setup - How can we put the 5S rule into play? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
Marc Thinking of Vista OS from Microsoft? Will it play HD-DVDs or Blu-ray? After Work and Weekend Discussion Topics 2
K Play this game. Try to kick the ball & beat me !! Coffee Break and Water Cooler Discussions 20
T When do the requirements of AS9102 come into play? First Article AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 9
Marc The 50 Year History of Play-Doh Coffee Break and Water Cooler Discussions 5

Similar threads

Top Bottom