Maintaining of Performance during the Lifetime of a IIa Medical Device

G

Gawelle

#1
Dear all,

I am new on this forum and I am working in the quality department of a company which manufactures dental rotary files made with nickel titanium.

It's a class IIa device and it's sold non-sterile.

I need help with the maintaining of performance during the lifetime of the medical device.

As far as I know, there is no guidelines regarding that matter and I have trouble defining what to do. :(

Thank you in advance for your help,

:thanx:

Gawelle
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Moderator
#2
Dear all,

I am new on this forum and I am working in the quality department of a company which manufactures dental rotary files made with nickel titanium.

It's a class IIa device and it's sold non-sterile.

I need help with the maintaining of performance during the lifetime of the medical device.

As far as I know, there is no guidelines regarding that matter and I have trouble defining what to do. :(

Thank you in advance for your help,

:thanx:

Gawelle
Hello Gawelle and welcome to the Cove :bigwave:

I'm not sure what you are referring to in "maintaining of performance". Are you talking about a technical aspect of the device itself or about some documentation / regulatory aspect? If you could describe the situation in more detail and maybe explain how you came to deal with this issue, it might help in directing us.

Thanks,
Ronen.
 
G

Gawelle

#3
Hello Gawelle and welcome to the Cove :bigwave:

I'm not sure what you are referring to in "maintaining of performance". Are you talking about a technical aspect of the device itself or about some documentation / regulatory aspect? If you could describe the situation in more detail and maybe explain how you came to deal with this issue, it might help in directing us.

Thanks,
Ronen.
Hi Ronen,

Thank you for your answer. I am sorry for the lack of clarity of my question, I will try to make it more understandable.
This issue comes from a remark made at the last audit. In the essentiel requirements of the 93/42/CEE directive, section 4 specifies that the characteristics and performances must not be affected during the lifetime of the device in normal conditions of use.
I am wondering how to prove it. Apparently, there is a link between clinical and pre-clinical evaluation and this matter.
The goal is to claim a 2 years lifetime. Technically speaking, I can have samples from 2 years ago for testing but the way to proceed isn't clear for me.
I hope this will clarify my question.

Thank you,

Gawelle
 

chris1price

Trusted Information Resource
#4
Hi Gawelle, I presume the rotary files are used more than once and then cleaned or sterilised betwen uses. I think you need to show that the performance of the file meets its requirements at the end of its life. This might not be simply time based, but on number of uses, or you might estimate how many uses occur in 2 years, and use that as a time justification.

You will need to decide what testing is required to show this, it might be clinical or bench testing, but essentially repeat the testing performed originally to show the file was effective at day 1.
 
G

Gawelle

#5
Hi Gawelle, I presume the rotary files are used more than once and then cleaned or sterilised betwen uses. I think you need to show that the performance of the file meets its requirements at the end of its life. This might not be simply time based, but on number of uses, or you might estimate how many uses occur in 2 years, and use that as a time justification.

You will need to decide what testing is required to show this, it might be clinical or bench testing, but essentially repeat the testing performed originally to show the file was effective at day 1.
Hi Chris,

Thank you for your answer.
The file is sold as a single use device. In our case, the lifetime in use is within 15 minutes that is why we consider the lifetime of the instrument stored, before use.
I think I should perform the validation tests (mechanical) which currently permit to release a batch.
Is there any other tests I should realise? Such as bioburden or biocompatibility after 2 years of storage.
Regarding clinical testing, we established in the clinical evaluation that clinical investigations don't have to be carried out for this file. Is it advised to realise clinical testing to prove the maintaining of performance during the lifetime anyway ?

Thank you,

Gawelle
 

Ronen E

Problem Solver
Moderator
#6
Hi Chris,

Thank you for your answer.
The file is sold as a single use device. In our case, the lifetime in use is within 15 minutes that is why we consider the lifetime of the instrument stored, before use.
I think I should perform the validation tests (mechanical) which currently permit to release a batch.
Is there any other tests I should realise? Such as bioburden or biocompatibility after 2 years of storage.
Regarding clinical testing, we established in the clinical evaluation that clinical investigations don't have to be carried out for this file. Is it advised to realise clinical testing to prove the maintaining of performance during the lifetime anyway ?

Thank you,

Gawelle
Hi Gawelle,

You are on the right track. Repeating you validation (bench) testing on 2-years old devices can support your shelf-life claim, you just need to ensure that the storage conditions the test specimens were exposed to are representative of real life, including worst-case scenarios. On top of that a risk analysis would tell you whether other characteristics need testing as well. For example, if bioburden is an issue (aren't the files cleaned right before use?) and there's a risk that it will exceed safe limits over time, test it. Biocompatibility testing is not likely to be required if you've done your initial biological evaluation properly - it should have already accounted for shelf-life deterioration. You can hypothesise and evaluate the specific risks better than us because you are more familiar with the device and the application.

I don't think that a clinical investigation is required to support your shelf-life claim as long as the literature review establishes that equivalent devices carry similar life claims and are handled (shipped, stored etc.) in a similar manner.
 
Thread starter Similar threads Forum Replies Date
J Maintaining site specific ISO Compliance/Certification within a non-certified organization & centralized processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
I Maintaining Technical Documentation under MDD and MDR EU Medical Device Regulations 1
S Maintaining Equipment Records - We use a manual system ISO 17025 related Discussions 1
B Main responsibility for Control Plans - creating and maintaining FMEA and Control Plans 15
P Procedure for maintaining Confidentiality And Impartiality in a Laboratory Quality Management System (QMS) Manuals 2
Crimpshrine13 Rules of achieving and maintaining IATF recognition - Determining audit days IATF 16949 - Automotive Quality Systems Standard 2
J Software for maintaining of Technical files and medical device documentation(RA need) Medical Information Technology, Medical Software and Health Informatics 2
B What is our requirement for maintaining the EMC series of standards? IEC 61000 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
JoCam Cost of maintaining CE Mark Regulatory Approval Other Medical Device Regulations World-Wide 3
A CE Mark - Maintaining Technical File and Regulatory Assessment CE Marking (Conformité Européene) / CB Scheme 1
W Maintaining Component Traceability in DHR (Design History Record) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
V Maintaining an E-Record for Equipment Data Qualification and Validation (including 21 CFR Part 11) 4
D Redundant BOMs - Maintaining a Reference Only BOM on our Top Level Assembly Drawings Document Control Systems, Procedures, Forms and Templates 7
D Procedure for Maintaining Industry Standards and Identifying Revisions Document Control Systems, Procedures, Forms and Templates 1
H Source for Compiling and Maintaining an EHS Legal Register Miscellaneous Environmental Standards and EMS Related Discussions 4
S Advice on moving QMS Documents and Maintaining Revision History Quality Manager and Management Related Issues 9
M AS9100 Rev C - Determining, Providing and Maintaining" Infrastructure AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
S Maintaining ISO 9001:2008 when Automotive Business is Slow ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
T Requirements - Controlling and Maintaining Sheet Metal Templates for Production Use ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
N ISO 9001 based Audit Schedules: Creating and Maintaining - Template wanted General Auditing Discussions 5
S What's the logical difference between maintaining & retaining records of review? Management Review Meetings and related Processes 10
S A question on Maintaining the Integrity of the QMS (5.4.2) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
C API Q1 8th ed/ISO TS 29001 Maintaining Responsibility for Outsourced Processes Oil and Gas Industry Standards and Regulations 6
S Quality Manual - Possibility of maintaining only SOFT COPIES Quality Management System (QMS) Manuals 17
A Maintaining Dual Standards ISO 9001 and ISO 13485 - Contract Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 9
T How many hours a week do you devote to maintaining your QMS? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 15
A Rechargeable Battery in Medical Device - Controling and Maintaining Batteries Other Medical Device and Orthopedic Related Topics 4
G Calibration laboratory start-up - Maintaining the same ambient conditions General Measurement Device and Calibration Topics 7
ScottK Maintaining an SPC program with reduced staffing Statistical Analysis Tools, Techniques and SPC 8
B Specifications - Maintaining current revisions Document Control Systems, Procedures, Forms and Templates 6
Sidney Vianna Maintaining Accredited Certification During the Global Financial Crisis Registrars and Notified Bodies 9
Colin Cost of Maintaining Auditor Certification (IRCA QMS Lead Auditor) ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 1
A Review and Verification for Maintaining Software Software Quality Assurance 10
Z Creating and Maintaining a Top Ten Poor Quality Supplier Performer List Supplier Quality Assurance and other Supplier Issues 16
T Plug Gages - How people are maintaining calibration on plug gages General Measurement Device and Calibration Topics 3
Q Production on Plastics Drums - Process Flow and Maintaining Cleanliness Manufacturing and Related Processes 2
apestate Advice on maintaining surface plate condition - Starrett crystal pink surface plate Inspection, Prints (Drawings), Testing, Sampling and Related Topics 14
gard2372 Implementation and Maintaining Lean Guide book - Seeking recommendations and reviews Book, Video, Blog and Web Site Reviews and Recommendations 10
A Maintaining traceability through our manufacturing process - Protective suits - 110 Misc. Quality Assurance and Business Systems Related Topics 1
G How is everyone handling TS2 Clause 8.2.3.1 - Maintaining Process Capability IATF 16949 - Automotive Quality Systems Standard 2
S Maintaining and tracking laboratory voltage standards - A2LA audit General Measurement Device and Calibration Topics 3
H Maintaining Certification - Do you need to take the test again? Professional Certifications and Degrees 5
P TS 16949 Clause 8.2.3.1 - Maintaining Process Capability IATF 16949 - Automotive Quality Systems Standard 9
R Guidance for 'Maintaining Compliance to 820 - FDA and ISO 13485:2003 ISO 13485:2016 - Medical Device Quality Management Systems 8
R Maintaining manufactoring process capability IATF 16949 - Automotive Quality Systems Standard 6
E Procedure Numbering Change? Still the need for maintaining Level 1, 2, 3, documents? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
T Maintaining Process Control QS-9000 - American Automotive Manufacturers Standard 1
M 4.9.2 Maintaining Process Control - Maintaining ongoing process capability QS-9000 - American Automotive Manufacturers Standard 13
E Clause 15.4.2.1 d) - Loss of ESSENTIAL PERFORMANCE IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Dazzur Sharing Suppliers Performance Data with Supplier. Supplier Quality Assurance and other Supplier Issues 6

Similar threads

Top Bottom