Maintenance of Class 9/8 clean room in a medical devices factory
G
gereard_kgb
#1
Hi Everyone,
Just recently we have launched our medical devices factory and we are in our learning period now.
We are running class 9/8 clean room and I am afraid we are having some problems with the level of cleanness in the clean room.
Could you share your experience regarding the maintenance of the clean room? What disinfectants and other cleaning products to you use to clean the premises?
How about cleaning the hands of the personnel? How often do they do it and what disinfectants are used for this purpose?
do they work in gloves or with bear hands?
Do you use UV laps in the whole clean room to enhance the bacteria killing?
All hints are welcome.
:thanx:
Elsmar Forum Sponsor
CarolX
Super Moderator
Super Moderator
#2
Re: Maintenance of Class 9/8 clean room
Sorry I can't help you but I moved your thread to the medical device board - may get some response here.
G
Gert Sorensen
#4
Hi Everyone,
And what a learning period you are in for :D
Sounds about right :D
Cleaning done on a daily basis. Operators clean the specific premises before production starts and an external company comes in at night to clean the entire room. Don't know what specific products are used, but try Google for "clean room cleaning products", you may get lucky.
Operators are required to wash hands everytime they have been out of the clean room. No gloves used.
No. We sterilize our products :lmao:
C
cochranemurray
#5
Have you established and validated a cleaning method with associated frequencies? Most freely available cleanroom disinfectants will do a job for you if you have validated your method. Have the cleanroom staff been instructed in the necessary hygiene rules? As well as a bacterial soap, we have hand sanitizer stations where staff can "top up".
When you say you have problems with cleanliness, have you established the possible source of the contamination (is it human or is it raw materials.
I presume when you started up the cleanroom, it was tested at rest and active. After a few monitoring sessions you will have set limits for levels of contamination and corrective action? Don't get too hung up on these limits as they will inevitably come under pressure as you build up staff numbers and increase production levels.
If you monitor the staff (usually the main source of contamination) and establish suitable cleaning frequencies, things should settle down, but the improvement may not be immediate.
What products are you making?
G
gereard_kgb
#6
Cochranemurray thank you very much for your valuable comments.
I will take this into account.
First of all I need to tell that our factory is located in China where all the standards differ a bit to those in Europe... :)
As we are just starting the project we are teaching our people how does the system works and what are the required hygiene measures etc.
We started with some connecting tubes etc but intend to launch class IIb products soon.
I am afraid the disinfectants we use are just not good enough.
People wash their hands on entering the clean-room but only with normal soap. What anti-bacterial soaps are used?
The other idea was that the disinfectants used for cleaning of the premises are not good enough. Right now we are launching tests and validation for chlorine dioxide as it seams to be quite promising - have you got any experience?
Of course every 2 hours people are spraying some other disinfectant on their hands while working to reduce the bacteria.
As for now I am not sure what is the main source of the contamination... we are working on it.
Gert, we also sterilize our products but I was wondering if it was not a good idea to use some UV lamps (just like they do in the hospitals) to enhance bacteria killing in the clean-room while people don't work there. Any comments to this?
Considering that hands of the workers could be some source of contamination I was wondering to make them use gloves... how do you do it at your facilities?
:thanx:
Ajit Basrur
Staff member
Admin
#7
Do you use UV laps in the whole clean room to enhance the bacteria killing ?
Hi Gereard,
When you say that you are having problems with level of cleanliness, do u mean that the number of particulate contamination is exceeding the limits ? This could be due to the air quality / personnel / materials / process itself. You may also want to have a look at the number of air changes etc. You need to investigate the reasons.
On cleaning hands, apart from cleaning hands with soap (any soap will do. Bactericidal soap not required; a sales gimmick. Soap being alkaline automatically kills bacteria). Ensure that all personnel before entering the cleanroom should sanitise their hands and wear clean attire. Personal hygiene is also of utmost importance.
They can work with bare hands but ensure that they wear gloves while handling the product. At any given time, direct contact by bare hands to the product should be avoided.
I am not in favour of UV lamps as UV is carcinogenic and is harmful to skin and eyes.
G
gereard_kgb
#8
Thank you for further valuable comments.
In fact we discovered that the products before sterilization are pretty contaminated.
I do not know yet what is the source of this contamination. I am looking for it.
Thank you for the comments on the soap.
As far as UV is concerned I was thinking of it as an additional measure that could be labour-free and could enhance bacteria killing.
Since I remember that UV can cause cancer I am sure it could only be switched on when the premises are empty.
The gloves - this is of some issue to me. In the majority of clean-rooms I've been people used to work with bear hands with appropriate disinfection measures.
Yet I thought this could be of some use to reduce any potential contamination of hands origin :)
As for all the other measures we are OK. People wash their hands before entering the premises and they disinfect them while working.
The cloaks - do you wash them everyday or people wear them for some longer period before washing?
:thanks:
G
Gert Sorensen
#9
I'm with Qualityalways on this one. Any standard liquid soap should be sufficient. There are so many things that cause more trouble than soap...
I would dig a bit more into how the desinfectants are used. Do you follow the suppliers instructions?
Hmmmmmm, I don't like that. How about allergies etc.?
I would not get into that. It might be overkill. Look into the air supply - along with hygiene. Do you use the correct HEPA-filters? How good is your over-pressure? Do you have sufficient air-changes?
Remember, your goal is not to have sterile products during production, it is to reduce the contamination.
Once again, allergies spring to my mind. You may want to have the operators wear full body suits that are being cleaned regularly (and of course hair nets).
Good luck on the challenge :bigwave:
G
gereard_kgb
#10
Thanks for the next insight.
We are investigation the chlorine dioxide now. We are doing some validation righ now. From the information we got from different manufacturers it should be fine. I hope to know it soon...
The allergies resulting from using the disinfectants too much? that could be a point but for now we have not discovered anything like this. perhaps it would be a good idea to use various and exchange them... In this case risk of allergies should be reduced.
The filters and over-pressure seams to be OK. The clean-room has been constructed just recently so I trust all should be fine here.
Overkill? Hm... is it a problem?
Thread starter Similar threads Forum Replies Date
E Software maintenance Process Software maintenance Process to IEC 6204? IEC 62304 - Medical Device Software Life Cycle Processes 3
D Preventive Maintenance vs. General Cleaning ISO 13485:2016 - Medical Device Quality Management Systems 2
J Help settle a disagreement: Should external providers of preventive maintenance be on your ASL? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
F Component maintenance qualification - Regulation (EU) No 1321/2014 Article 5 EASA and JAA Aviation Standards and Requirements 6
K Generic preventive maintenance form wanted Document Control Systems, Procedures, Forms and Templates 2
S Quality management system (Well head installation & maintenance) Quality Management System (QMS) Manuals 5
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
A Feedback of typical maintenance problems with LMI digital probes Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
S Are there ISO9001:2015 Requirements for Preventive Maintenance? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 36
Q Software SOP - Use and maintenance of an ERP system Software Quality Assurance 6
R Robotic weld cell - maintenance issues - expertise needed Manufacturing and Related Processes 0
J Requirements for Layered Process Audits including Maintenance, Shipping, Receiving IATF 16949 - Automotive Quality Systems Standard 4
D Preventive Maintenance Implementation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
D Preventative Maintenance for Gages, handtools etc Other Medical Device Regulations World-Wide 1
C Determining if Maintenance Contractor is an External Service subject to ISO 9001 Clause 8.4 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 43
C Maintenance of Product Technical Documentation EU Medical Device Regulations 3
R IATF 16949 Clause 8.5.1.6 a) maintenance and repair facilities - Production tooling management and personnel IATF 16949 - Automotive Quality Systems Standard 4
qualprod Best criteria to measure Corrective Action effectiveness - Poor Maintenance ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
M Minitab tool to evaluate PM (Preventive Maintenance) process Using Minitab Software 6
B Total Productive Maintenance - Company Transport IATF 16949 - Automotive Quality Systems Standard 2
P Minor Non-Conformance - Maintenance Records Not Fully Completed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 24
L 21CRF820.70(g)(2) Maintenance Inspection Interval 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
I Tool maintenance and calibration procedure - Calibrating tools/equipment AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
K IATF16949 8.5.1.5 Total Productive Maintenance IATF 16949 - Automotive Quality Systems Standard 9
J FAA Maintenance Requirements - DER Repair to a PMA Part Federal Aviation Administration (FAA) Standards and Requirements 1
K Maintenance requirements for Production Equipment in AS9100D General Auditing Discussions 8
T Part 145 Maintenance Station S/N ESN Character Requirements EASA and JAA Aviation Standards and Requirements 3
M Preventive Maintenance Spreadsheet in Excel Quality Assurance and Compliance Software Tools and Solutions 2
B IATF 16949 Section 8.5.1.5 (Total Productive Maintenance) IATF 16949 - Automotive Quality Systems Standard 19
D Does AS9100 certified OEM need AS9110 if no maintenance is provided? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
N Laboratory Audit Nonconformance - Maintenance of Standards ISO 17025 related Discussions 10
A Technical File Maintenance - Old generation product EU Medical Device Regulations 4
L Seeking help in Basic Maintenance Knowledge Manufacturing and Related Processes 4
M CAPA due to Consumer not following Maintenance Instructions... Nonconformance and Corrective Action 9
S Preparing a Turtle Diagram for Maintenance Department Misc. Quality Assurance and Business Systems Related Topics 2
U Relating maintenance to product quality Manufacturing and Related Processes 5
S ISO 9001 Audit requirement for maintenance department ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
C Service and maintenance in an initial importer facility USA 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
M Total Productive Maintenance vs. Lean, Six sigma, Business Process Reengineering, etc Quality Tools, Improvement and Analysis 9
A Electronic Maintenance Records stored in a machine ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
A How to create a Maintenance procedure based on TS 16949 Requirements IATF 16949 - Automotive Quality Systems Standard 5
N ISO 9001 Quality Manual for an O&M (Operations and Maintenance) Company ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 14
S Predictive Maintenance Format / Example Misc. Quality Assurance and Business Systems Related Topics 4
T Are Customers required to follow a Manufacturers Maintenance Schedule? ISO 13485:2016 - Medical Device Quality Management Systems 3
E Manufacturer Responsibility (vs. User/Customer) for Preventive Maintenance/Repair 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
A Technical Files/Dossier/STED File Maintenance for Notified Body Audiits Other Medical Device Regulations World-Wide 2
C Document Revision for each new Preventative Maintenance Checklist? Document Control Systems, Procedures, Forms and Templates 5
M Where to start Maintenance Function Kaizen Lean in Manufacturing and Service Industries 4
A Assessing a Preventive Maintenance Strategy - Reliability or Maintenance Statistics Reliability Analysis - Predictions, Testing and Standards 2
J Maintenance Man Hour Plan which satisfies EASA Part 145.A.30 (example needed) EASA and JAA Aviation Standards and Requirements 4
Similar threads
Top Bottom