MDD and Mandatory Standards (IEC/EN 60601-1-4, 60601-1-6 and ISO 14971)

LievenDW

Involved In Discussions
#1
This may be a stupid question, but is it mandatory for the MDD to comply with IEC/EN 60601-1-4, 60601-1-6 and ISO 14971? When I read the MDD official document, these standards are not mentioned literally.
 
Elsmar Forum Sponsor
A

Alexisss

#4
Standards are voluntary. The use of harmonized standards gives presumption of conformity with the related essential principles of the directives, therefore it facilitates the process if you use them.

Please see further discussions in this topic: http://elsmar.com/Forums/showthread.php?p=237965#post237965
Adding to mmantunes... you must show SOMEHOW conformity with the essential requirements (annex 1) of the MDD. One way (presumption as mmantunes says) is by conforming to the harmonized standards. BUT, may be you have another way of proving conformity to the essential requirements, something that your notified body can assess and give u a CE mark for your product (if not class I).
 

Peter Selvey

Staff member
Super Moderator
#7
Just to clarify:

The application of harmonised standards gives you a presumption of conformity (see Article 5), which effectively absolves you of having to explain why you meet the essential requirement. In fact, you don't even need an essential requirements checklist - the list of applied standards and the "risk analysis" is enough. Only where you do not apply the harmonised standard, you are legally required to explain your alternate solution for the applicable essential requirements.

If you find this surprising and unreasonable (as I did), take a careful look at the wording in the MDD itself, for example Annex II, 3.2 (c) 2nd dash.

After digging through EU commission docs dating back to 1984 I found the reason for this, but that's too long and explanation for this post!
 
Thread starter Similar threads Forum Replies Date
Ed Panek Notified Body Substantial Change MDD EU Medical Device Regulations 1
J How to keep MDD certificate valid when legal manufacturer has liquidity problem EU Medical Device Regulations 0
Ed Panek MDD Cleared Private Labelled Devices into EU EU Medical Device Regulations 0
K Section 13.6 Annex 1 MDD 93/42/EEC for Glove EU Medical Device Regulations 0
Ed Panek IFU Contact Requirements - FDA and MDD/MDR US Food and Drug Administration (FDA) 1
U MDD Article 12/ MDr Article 22 interpretation EU Medical Device Regulations 2
E Our company is planning to file MDD not MDR next month. Do we require to show chemical characterization report ? CE Marking (Conformité Européene) / CB Scheme 2
S Converting SOP MDD to MDR Noob ISO 13485:2016 - Medical Device Quality Management Systems 4
silentmonkey Are risks in supply chain and development activities within scope of MDD? EU Medical Device Regulations 4
A First CE-mark class III implantable device under MDD EU Medical Device Regulations 8
Q MDD - distributor EU Medical Device Regulations 1
R MDD x PPE Directive - Statement of Non-Applicability EU Medical Device Regulations 3
D Does Manufacture can submit CE mark application under MDD with NB for his New product after May 2020? EU Medical Device Regulations 3
H MDD Article 12 Labeling for Class IIa Medical Device - Please Advise EU Medical Device Regulations 3
M Requirements for an MDD approved device after 26/05/2021 EU Medical Device Regulations 3
H MDD CM, MDR manufacturer. 2020. Is it ok? EU Medical Device Regulations 1
S Technical File for Surgical Mask in Class 1 MDD Elsmar Cove Forum ToS and Forum Policies 1
S MDR Delay - MDD design Change? (before new MDR DOA) EU Medical Device Regulations 8
P What ASTM std is EC 93/42 MDD EQUIVALENT TO CE Marking (Conformité Européene) / CB Scheme 1
R Applicability of new non-harmonized standards (MDD/MDR) EU Medical Device Regulations 14
N Label Placement Requirement in MDD/MDR EU Medical Device Regulations 1
K Measuring Function MEDDEV 2.1/5 relation with MDD 93/42/EEC CE Marking (Conformité Européene) / CB Scheme 1
A Class medical device (MDD) - Classification help EU Medical Device Regulations 1
Z Economic Operators Responsibilities after DOA for legacy MDD devices EU Medical Device Regulations 2
J MDD to MDR transition - Time required for the implementation of the MDR EU Medical Device Regulations 7
F MDR system/procedurepack (article 22) with device acc. to MDD and MDR CE Marking (Conformité Européene) / CB Scheme 2
I Maintaining Technical Documentation under MDD and MDR EU Medical Device Regulations 1
P Adding "new" medical device to existing CER - under MDD CE Marking (Conformité Européene) / CB Scheme 1
Watchcat Duration of MDD Certification EU Medical Device Regulations 4
Z MDD ER 7.5, "phthalate test"? Aspiration tips EU Medical Device Regulations 3
R Notified Body for MDD 1Q20 - Florida Company Registrars and Notified Bodies 4
S Until when can we CE-mark and sell MDD class 1 devices? CE Marking (Conformité Européene) / CB Scheme 17
D CE Marking Requirements MDD & MDR - new product development covered under same scope EU Medical Device Regulations 1
G Harmonized List - MDD 93/42/EEC Other ISO and International Standards and European Regulations 2
B Accredited laboratory or not under MDD 93/42/EEC IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
T Custom made medical device MDD vs MDR definition CE Marking (Conformité Européene) / CB Scheme 12
Watchcat MDD, MDR/IVDR, Working Groups, Expert Panels, MEDDEVs and Other EU Guidances EU Medical Device Regulations 32
D MDD Certification Validity after May 2020 EU Medical Device Regulations 12
A Compatibility between MDD and MDR devices EU Medical Device Regulations 7
M Device CE certified under MDD / MDR > new GTIN? EU Medical Device Regulations 7
N When is "placing on the market"? MDD certified medical devices EU Medical Device Regulations 32
G Essential Requirements Checklist - (MDD 93/42 / EEC) Misc. Quality Assurance and Business Systems Related Topics 2
K EU MDD Class I Device market access / certification after 26th May 2020 EU Medical Device Regulations 9
V CE Mark under MDD or IVDR - IVD company doing Virtual Manufacturing Other Medical Device Regulations World-Wide 3
JoshuaFroud The withdrawal of Lloyd’s Register MDD & IVDD Notified Body Status CE Marking (Conformité Européene) / CB Scheme 1
A How to manage the QMS system and SOP during the transition from MDD to MDR EU Medical Device Regulations 4
F Class I Device under MDD, reclassified to Class II under MDR Other Medical Device Related Standards 5
S Dual MDD and MDR certification EU Medical Device Regulations 7
K Obligation for Distributor in MDD/MDR CE Marking (Conformité Européene) / CB Scheme 4
D MDD to MDR - Technical File conversion EU Medical Device Regulations 4