Hello,

We want to sell a ventilator manufactured by combining devices bearing a CE marking, consisting of a mask, valve, filter, connectors, and oxygen line tubing. Each of the elements already has his manufacturer's CE marking and the intended use is not changed. Coupled pieces (product) are ready to use after connecting lines with oxygen tank.

In your opinion, according to MDR article 22, could this be considered as a system?

Thanks in advance for your comments.
 
Elsmar Forum Sponsor

Marc

Hunkered Down for the Duration with a Mask on...
Staff member
Admin
#2
Help with this one will be appreciated. Thanks in advance!
 

Asia78

Involved In Discussions
#3
Yes, i agree with your assessment. In this case you want to bring to the market all those devices together and you are creating a system. All related requirements have to be followed
 

Watchcat

Trusted Information Resource
#4
I think the key phrase is in section 1 of Article 22:

"in a manner that is compatible with the intended purpose of the devices or other products and within the limits of use specified by their manufacturers"

I'm skeptical that intended use has not changed. For starters, are all of the individual devices intended for use in patients that require ventilation? Are they all intended to be used to ventilate such patients? I guess it's possible, but it seems unlikely.

Of course Section 1 doesn't say that the intended use has to be the same, only "compatible with." But when someone tells me intended use hasn't changed, and it turns out it has changed a good bit, I'm inclined to think that, when it comes to assessing whether the devices are "compatible with," they are not likely to be on target in that assessment, either.
 
Thread starter Similar threads Forum Replies Date
U MDD Article 12/ MDr Article 22 interpretation EU Medical Device Regulations 2
F Interpretation of MDR 2017/745 Article 23 - CE Mark Requirements EU Medical Device Regulations 8
L EU MDR Article 86 PSUR EU Medical Device Regulations 1
R Requirements for MDR Article 22 Systems EU Medical Device Regulations 1
N EU MDR - Applicability Article 22 Systems and Procedure Packs CE Marking (Conformité Européene) / CB Scheme 4
M Example of statement for procedure pack MDR Article 22? Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 0
S Class I MDR Article 10 – QMS EU Medical Device Regulations 1
R Clinical evaluation without clinical data - MDR Article 61(10) EU Medical Device Regulations 6
A MDR Article 22 applicability - Legal manufacturer EU Medical Device Regulations 6
S Article 61 - MDR Clinical Investigation Requirement EU Medical Device Regulations 5
C Article 16 MDR - Medicine packaged together with Medical Device EU Medical Device Regulations 0
F MDR system/procedurepack (article 22) with device acc. to MDD and MDR CE Marking (Conformité Européene) / CB Scheme 2
N EUDAMED postponement and compliance with Article 120 (3) MDR for Legacy Devices EU Medical Device Regulations 8
R MDR Article 22: Systems and procedure packs Other Medical Device Regulations World-Wide 12
T EU MDR Article 61- Clinical Evaluation EU Medical Device Regulations 2
M Informational EU – MDCG 2019-7 Guidance on Article 15 of the Medical Device Regulation (MDR) and in vitro Diagnostic Device Regulation (IVDR) regarding a “person re Medical Device and FDA Regulations and Standards News 7
E MDR Article 117 / Impact to New Marketing Authorization Applications Other Medical Device Regulations World-Wide 1
C Access to Technical Documentation - MDR Chapter VI Article 61 EU Medical Device Regulations 4
T MDR article 22 - System of 2 CE products and validation / usability? EU Medical Device Regulations 10
T Clarification on MDR - Article 18(d) - Implant Card EU Medical Device Regulations 14
M Is MDR Article 61(5) Unrealistic? EU Medical Device Regulations 9
D Responsible for regulatory compliance (Article 15 of MDR 2017/745) Quality Manager and Management Related Issues 4
S EU MDR - Liability Clause per article 31 EU Medical Device Regulations 22
S EU MDR Annex I Chapter III 23.4 z EU Medical Device Regulations 0
R MDR & depth of review by NB's EU Medical Device Regulations 7
U EU-MDR: complicated supply chain EU Medical Device Regulations 0
C EU MDR - Annex II 6.1 Pre-clinical and clinical data EU Medical Device Regulations 4
A IVD in MDR procedure pack EU Medical Device Regulations 2
Ed Panek IFU Contact Requirements - FDA and MDD/MDR US Food and Drug Administration (FDA) 1
S Are EC type examinations still being conducted under MDR? EU Medical Device Regulations 5
E MDR internal audit Medical Device and FDA Regulations and Standards News 0
R The term "Benefit Risk Ratio" in EU MDR, do I need to present benefit risk analysis as a RATIO Risk Management Principles and Generic Guidelines 4
R MDR, Annex I, 23.1 Interpretation - IFU on the website EU Medical Device Regulations 5
K Two FDA questions regarding UDI (and potentially 803.52 MDR) Other US Medical Device Regulations 0
S EU MDR CAPA - GAP Assessment on CAPA SOPs EU Medical Device Regulations 1
A User interface Language Requirements - EU MDR EU Medical Device Regulations 1
S How is the service life for class I reusable medical devices described in MDR? EU Medical Device Regulations 7
S How is the service life described for class I reusable medical devices in MDR? Elsmar Xenforo Forum Software Instructions and Help 1
P MDR/IVDR Annex III - Technical documentation on PMS EU Medical Device Regulations 1
E Our company is planning to file MDD not MDR next month. Do we require to show chemical characterization report ? CE Marking (Conformité Européene) / CB Scheme 1
S Classification of a product according to MDR EU Medical Device Regulations 3
L Classification under MDR EU Medical Device Regulations 1
I Interpretation of the MDR GSPR 23.4 (u) EU Medical Device Regulations 2
S Converting SOP MDD to MDR Noob ISO 13485:2016 - Medical Device Quality Management Systems 4
K Biological indicator system classification under MDR EU Medical Device Regulations 2
L 'Distributor' definition & MDR/MSR relationship EU Medical Device Regulations 3
A What does this line from MDCG 2020-3 (MDR art. 120 substantial change) mean to you? EU Medical Device Regulations 4
J EU MDR getting closer - Special offer for Elsmar members 10/2020 EU Medical Device Regulations 14
D ISO 14971:2019 vs MDR Annex 1, Requirement #4 - "Manufacturers shall inform users of any residual risks" ISO 14971 - Medical Device Risk Management 5
G MDR translation requirements EU Medical Device Regulations 3

Similar threads

Top Bottom