With the MDR delayed until May 2021, are able to make a design change to our MDD compliant device? We have a design change project, where all V&V and necessary documentation will be completed by May 2020. Based on Article 120 (3), since the DOA is now May 2021, can we make this design change? Is Article 120(3) saying that we can make design changes to MDD compliant devices prior to the DOA?

Article 120(3)

"By way of derogation from Article 5 of this Regulation, a device with a certificate that was issued in accordance with Directive 90/385/EEC or Directive 93/42/EEC and which is valid by virtue of paragraph 2 of this Article may only be placed on the market or put into service provided that from the date of application of this Regulation it continues to comply with either of those Directives, and provided there are no significant changes in the design and intended purpose. However, the requirements of this Regulation relating to post-market surveillance, market surveillance, vigilance, registration of economic operators and of devices shall apply in place of the corresponding requirements in those Directives."
 
Elsmar Forum Sponsor

Asia78

Involved In Discussions
#7
Yes, you can make significant changes to mdd devices until may 2021.
The new DoA implies also the postponment of all the applicable requirements.
The published MDCG guidances are still valid in their content, but the dates written may be wrong being that they were published before the latest changes to the Mdr
 

Philip B

Involved In Discussions
#8
BSi stopped clients making significant changes to their MDD certificates last year as there would not have been enough time to implement these before May 2020. Not sure what their stance is now the MDR has been put back a year? Best to check with your NB.
 
Thread starter Similar threads Forum Replies Date
M Informational From RAPS – EU MDR/IVDR: US Raises ‘Serious Concerns,’ Urges 3-Year Implementation Delay Medical Device and FDA Regulations and Standards News 3
P Update on NBOG 2014-3 to address MDR/IVDR, or any plan to do that? EU Medical Device Regulations 0
A MDR requirement where unit of use packaging is too small for UDI carrier EU Medical Device Regulations 1
S MDR GSPR Standards EU Medical Device Regulations 1
A MDR - Legacy Device Review Timeframe and Requirements EU Medical Device Regulations 3
J Translation requirements for the statement referred to in ANNEX XIII of MDR EU Medical Device Regulations 0
J EU MDR GSPR 10.4.3 and 10.4.4 EU Medical Device Regulations 2
F Labelling to comply with both FDA and MDR US Food and Drug Administration (FDA) 6
N EU MDR and its impact on existing Registrations EU Medical Device Regulations 1
M Post Market Safety Update Report for devices that will be up-classified under MDR but are currently under MDD EU Medical Device Regulations 3
Y MDR Transition - make "Available on the market" after CE certificate expiration EU Medical Device Regulations 2
C MDR Classification Rule 10 EU Medical Device Regulations 13
C MDR software classification EU Medical Device Regulations 12
Y MDR requirements for Class I accessories EU Medical Device Regulations 2
M MDR legal actions - manufacturers CE Marking (Conformité Européene) / CB Scheme 8
R Products not within the MDR grace period EU Medical Device Regulations 1
shimonv MDR transition checklist EU Medical Device Regulations 0
K Re-packaging under MDR EU Medical Device Regulations 3
Ed Panek Upcoming NB MDD ---> MDR Crunch EU Medical Device Regulations 0
S MDR - System and procedure pack article 22 and all sub processes that apply ISO 13485:2016 - Medical Device Quality Management Systems 0
C Requirements for distributors under MDR: translation EU Medical Device Regulations 0
A Should we assign the PRRC before the date of application of MDR (26 May 2021)? EU Medical Device Regulations 0
N Looking for a recommendation for an EU MDR Importer EU Medical Device Regulations 1
JoCam Rental and MDD versus MDR EU Medical Device Regulations 1
A Can anyone share a Distribution Agreement template under MDR 2017/745? EU Medical Device Regulations 0
A MDR ANNEX XI part A or Part B EU Medical Device Regulations 0
N OBL under the MDR. Which are the obligations of the OEM? EU Medical Device Regulations 0
U Distributor obligations under MDR EU Medical Device Regulations 5
Raisin picker CECP, consultation according to Art 54 MDR EU Medical Device Regulations 1
K Question on MDR classification EU Medical Device Regulations 4
M STED vs MDR EU Medical Device Regulations 0
F UDI-PI required on packaging (MDR) EU Medical Device Regulations 2
M Extension to shelf life of new product under MDR EU Medical Device Regulations 1
J Instructions for use for Class I devices under MDR EU Medical Device Regulations 1
K Importer in EU, but not distribute in the EU - MDR obligations EU Medical Device Regulations 0
K MDR Class 1 Self Certification - Competent Authority or Notified Body? EU Medical Device Regulations 6
M Responsibilities of Importers / Distributors - MDR EU Medical Device Regulations 6
S PSUR - MDR First Submission Requirement? EU Medical Device Regulations 2
R EU-MDR Article 9-common specifications Other ISO and International Standards and European Regulations 2
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 4
JoCam MDR Gap Analysis - Consultant fees EU Medical Device Regulations 10
N Is there a need for clinical test of Class IIa products (for MDR)? EU Medical Device Regulations 2
M Examples of Combination Products - MDR Article 1 (8) and MDR Article 1(9) Medical Device and FDA Regulations and Standards News 3
P EU MDR Translation of Product Name EU Medical Device Regulations 3
N Is this a single integral drug device combination product EU MDR CE Marking (Conformité Européene) / CB Scheme 1
R Components to a finished medical device, MDR requirements Other US Medical Device Regulations 1
K Quality Agreement MDR - The manufacturer is outside of EU Other Medical Device Related Standards 4
M MDR - eIFU - When you can supply an electronic IFU instead of a paper IFU EU Medical Device Regulations 6
M MDR harmonized standards EU Medical Device Regulations 3