Hi All,
Slightly confused by the expectations for the PMS plan, as required by the MDR Article 84
Article 83 lays out that the PMS System shall be integral to the QMS
If the PMS System shall be based on the plan, and the PMS System shall be integral to the QMS surely the hierarchal structure puts the plan at the top, defining what your requirements are, the system below it with all of the relevant processes and procedures that fulfil the system requirements, then the output of those processes is the PMS Report or PSUR, and any subsequent activities in accordance with the processes/procedures, based on the conclusions.
After reviewing the detail required by the PMS Plan as per Annex III, with a few specific examples, this seems to confirm the plan as a much higher level document than a per-device document.
These all seem like very generic requirements that should be fulfilled by integrated into procedures within a business and then applied consistently as per the risk classification of the device in question.
If all of these have to be documented in a plan, for each device, surely it's nearly all going to be standard/static text for every single plan with the only part I can see as being unique to an individual device is any justification for no PMCF requirements.
I see very little value add in this, but would really like to understand how other people have managed this?
Am I missing something fundamental here?
Thanks
TS
The post-market surveillance system referred to in Article 83 shall be based on a post-market surveillance plan, the requirements for which are set out in ►C2 Section 1 of Annex III. ◄ For devices other than custom-made devices, the post-market surveillance plan shall be part of the technical documentation specified in Annex II.
1. For each device, manufacturers shall plan, establish, document, implement, maintain and update a post-market surveillance system in a manner that is proportionate to the risk class and appropriate for the type of device. That system shall be an integral part of the manufacturer's quality management system referred to in Article 10(9).
(b) The post-market surveillance plan shall cover at least:
— effective and appropriate methods and tools to investigate complaints and analyse market-related experience collected in the field;
— methods and protocols to communicate effectively with competent authorities, notified bodies, economic operators and users;
— reference to procedures to fulfil the manufacturers obligations laid down in Articles 83, 84 and 86;
— systematic procedures to identify and initiate appropriate measures including corrective actions;
Thanks