MDR PSUR versus Canadian Summary report

bill.Kal

Involved In Discussions
Hi Community

I am trying to figure out our post-market surveillance requirements for our IIa/IIb CE-marked (legacy devices) but in Canada (they also are on the market there already).
I have been reading the latest canadian guidance and I find no significant differences from MDR's PSUR. Since we already have that ready for our MDR submission do you think I could just change the compliance statement and submit the same document? Anyone has any experience/feedback from Health Canada after submitting a PSUR as a Summary Report? Or even better does anyone have any comments on the differences of the 2 documents?
Thank you all for the precious help
 

DannyK

Trusted Information Resource
The Canadian post market surveillance report does not have to be submitted to Health Canada, unless requested or unless the benefit/risks have changed significantly.
It will be verified by your MDSAP auditor.
Health Canada also asks for foreign risk notification.
There is a guidance document for foreign risk notification that you can combine with post market surveillance report.
 

bill.Kal

Involved In Discussions
The Canadian post market surveillance report does not have to be submitted to Health Canada, unless requested or unless the benefit/risks have changed significantly.
It will be verified by your MDSAP auditor.
Health Canada also asks for foreign risk notification.
There is a guidance document for foreign risk notification that you can combine with post market surveillance report.


--
thank you for the valuable feedback! If I understand correctly, what you are saying is that the Summary Report must be part of the periodic MDSAP audits and not a stand-alone independent and direct submission to the canadian authorities?
 

DannyK

Trusted Information Resource
--
thank you for the valuable feedback! If I understand correctly, what you are saying is that the Summary Report must be part of the periodic MDSAP audits and not a stand-alone independent and direct submission to the canadian authorities?

The Summary report and foreign risk notification can be requested during the MDSAP audit.
There is no direct submission to Health Canada. It can be requested by Health Canada.
 
Thread starter Similar threads Forum Replies Date
A PMSR or PSUR for legacy MDD-class I becoming MDR class-IIa EU Medical Device Regulations 8
S PSUR - MDR First Submission Requirement? EU Medical Device Regulations 8
L EU MDR Article 86 PSUR EU Medical Device Regulations 1
P EU MDR and PSUR (Periodic Safety Update Report) EU Medical Device Regulations 46
Ed Panek MDR and Accessories EU Medical Device Regulations 0
N MDR EU Medical Device Regulations 1
R MDR vs. Machinery Directive - Single EU Declaration of Conformity EU Medical Device Regulations 2
M EU MDR Economic operator -Importer EU Medical Device Regulations 2
J MDR Consequences on existing stock for PLM EU Medical Device Regulations 3
A Class Ir device have to be certified under MDR ? EU Medical Device Regulations 2
B Under MDR, are VATS and MICS still considered just class Ir? ISO 13485:2016 - Medical Device Quality Management Systems 1
G Trend analysis according to Article 88, MDR (EU), type of data point. EU Medical Device Regulations 1
S MDD Certificate extension according to 2023/67 MDR extension EU Medical Device Regulations 3
A Will EU MDR Notifying Body CER and CEP Technical documents Other Medical Device Related Standards 1
D What does a contract manufacturer have to do to comply with for MDR? EU Medical Device Regulations 2
G Trend Reporting according to Article 88 MDR, custom made implant EU Medical Device Regulations 5
G MDR critical supplier audit checklist EU Medical Device Regulations 0
G MDR/UKCA Supplier Audits EU Medical Device Regulations 3
C Revising a DoC for a legacy device under MDR/IVDR transitional periods EU Medical Device Regulations 7
D MDR literature review on device without clinical claims EU Medical Device Regulations 7
B EU MDR transitional period extension EU Medical Device Regulations 4
I Classification of device per Annex 8 of the EU MDR EU Medical Device Regulations 4
R Class I devices in the extended MDR transition period EU Medical Device Regulations 17
A MDR- Technical Documents EU Medical Device Regulations 4
P Distribution of "CE" Declaration of conformity under MDR EU Medical Device Regulations 3
E Expired CE with no pending MDR certification EU Medical Device Regulations 4
T SaMD labeling requirements - MDR IEC 62304 - Medical Device Software Life Cycle Processes 3
L Annex of ISO 13485:2016 as harmonized standard (MDR/IVDR) ISO 13485:2016 - Medical Device Quality Management Systems 2
P MDR Reporting Requirements for Devices Already covered in a correction under 21 CFR 806 US Food and Drug Administration (FDA) 0
M Clinical evaluation process of FDA and MDR Medical Device and FDA Regulations and Standards News 2
G MDR CE Expiration Date EU Medical Device Regulations 4
C MDR vs CMDR EU Medical Device Regulations 4
O MDR language requirements EU Medical Device Regulations 12
K Whether the product is an accessory according to the MDR definition EU Medical Device Regulations 3
L Adverse Event Reporting Flowchart (MDR EU) EU Medical Device Regulations 3
Doninina Trend Reporting for medical device class IIb according to MDR EU Medical Device Regulations 4
Ed Panek MDR Delay? EU Medical Device Regulations 2
dgrainger Informational Proposal to address MDR challenges EU Medical Device Regulations 23
goldenguo MDR/IVDR Insurance and Liability for Manufacturers, EU Representative and Importer EU Medical Device Regulations 3
C MDR 2017/744 Article 22 EU Medical Device Regulations 3
K Sales of MDD devices with expired CE that are pending MDR certification EU Medical Device Regulations 4
D New Device Post Market Requirement for MDR EU Medical Device Regulations 3
D EU MDR Derogation EU Medical Device Regulations 0
Doninina Risk management file according MDR or ISO 14971:P2019 ? EU Medical Device Regulations 2
M On-demand EU MDR Compliance toolkit with online training courses, templates & document management system Training - Internal, External, Online and Distance Learning 0
M EU MDR (GSPR Checklist) consolidated with TGA EP Checklist EU Medical Device Regulations 5
H MDR clinical evaluation EU Medical Device Regulations 0
M Possible topic for Master thesis topic - MDR and QMS ISO 13485:2016 - Medical Device Quality Management Systems 4
D MDR technical file: does it need to contain duplicates of controlled documents EU Medical Device Regulations 2
D Does the MDR article 10 (14) needs to be documented word for word in a Procedure? EU Medical Device Regulations 1

Similar threads

Top Bottom