MDR reporting and CAPA thoughts? Classifying Complaints on Risk

Jkc3usc12

Involved In Discussions
#1
System I am working with is classifying complaints on risk. If it is high risk and determine to report a CAPA is automatically initiated.

If a medwatch is received a 3500a is filled out no matter and again a CAPA is initiated.

MDR Reporting and CAPA have always been separate decisions for me in the past. I understand they can tie in together just not every time. Also reporting any medwatch received I have never done.

Thoughts?

Thanks
 
Last edited:
Elsmar Forum Sponsor

Marcelo

Inactive Registered Visitor
#2
Usually, reportability is not directly related to risk classification (as it was done before by the regulator), it's simply a decision - does the event fits the criteria? If so, it's reportable, if not, it's not reportable.

As you mention CAPA (which is something that in practice do not exist, because you do not co CA PA or PA CA, you do either PA or correction, which may escalate to CA), maybe the problem is exactly in this distinction (which is a common problem in the US because people are really into the CAPA thing).

Normally, things would be like this:

- Event occurred.
- Is it reportable? If so, report.
- Is correction or any other immediate action necessary? If so, implement.
- What is the root cause of the event?
- Depending on the root cause, a CA may be necessary to prevent recurrence. It may be the case that a CA is not required.
 

Jkc3usc12

Involved In Discussions
#3
A risk matrix for complaints is used to determine risk and which if coded high is reported. If reported a PA or CA is automatically initiated.

I feel the investigation of the root cause would be the determination of the PA or the CA if needed. Not just because it is high risk.

Also your thoughts on receiving a medwatch and automatically initiating a 3500 and another PA or CA.
 

Marcelo

Inactive Registered Visitor
#4
A risk matrix for complaints is used to determine risk and which if coded high is reported. If reported a PA or CA is automatically initiated.
A report has to be submitted (I'm using a manufacturer as an example) when the manufacturer "they learn that any of their devices may have caused or contributed to a death or serious injury. Manufacturers must also report to the FDA when they become aware that their device has malfunctioned and would be likely to cause or contribute to a death or serious injury if the malfunction were to recur."

This has nothing to do with risk classification or analysis. It's either fir the trigger or not.

I feel the investigation of the root cause would be the determination of the PA or the CA if needed. Not just because it is high risk.
If the event already occurred, there can be no PA, as PA if before something happen. The investigation of the root cause (after any correction or other immediate action) would then be used to determine if a CA is required.
 

Jkc3usc12

Involved In Discussions
#5
I agree with you. The risk assessment is to help determine a potentially reportable incident. I just having a hard time saying that every reportable event leads to the capa process and also determining to report is based solely off the risk matrix. Not actually what happened.

I also have a hard time reporting when any medwatch is received which also then goes through the capa process.

I am in the process of trying to steer this away from the direction they are in.

Just seeing if my thoughts are off base?
 

Marcelo

Inactive Registered Visitor
#6
I agree with you. The risk assessment is to help determine a potentially reportable incident. I just having a hard time saying that every reportable event leads to the capa process and also determining to report is based solely off the risk matrix. Not actually what happened.

I also have a hard time reporting when any medwatch is received which also then goes through the capa process.

I am in the process of trying to steer this away from the direction they are in.

Just seeing if my thoughts are off base?
I think you are in the right path.

As I mention before, the problem seems to be in the concept of CAPA. A lot of people thinks that a CA is always required when a problem is found, which is not true. ISO 13485 makes it more clear, because the text in the CFR is misleading.

An correction or other immediate action is required after an NC is detected (see 8.3.2 of ISO 13485).

After that, you need to evaluate the root cause of the NC/complaint (8.5.2 b) of ISO 13485).

Then you need to device if there's a need to escalate to a CA or not (8.5.2 c) of ISO 13485).
 

Jkc3usc12

Involved In Discussions
#7
Yes those steps make sense.

You agree that there can be MDR reporting without CA?

The view of risk dictates everything so one cant go without the other. high risk report and capa. Low risk nether.
 

Jkc3usc12

Involved In Discussions
#9
They view of risk dictates everything so one cant go without the other. high risk report and capa. Low risk nether.

So there is no possible MDR reporting without a CAPA being issued. Its ether both or nether. If the complaint is deemed high risk its reported and a capa.

Same thing with a medwatch received will automatically be reported and a CAPA issued.
 

Marcelo

Inactive Registered Visitor
#10
They view of risk dictates everything so one cant go without the other. high risk report and capa. Low risk nether.

So there is no possible MDR reporting without a CAPA being issued. Its ether both or nether. If the complaint is deemed high risk its reported and a capa.

Same thing with a medwatch received will automatically be reported and a CAPA issued.
Yes, I understood what you said, but to do it this way does not make any sense.
 
Thread starter Similar threads Forum Replies Date
B FDA-Medical Device Reporting (MDR )procedure compliant with 21CFR section 803 US Food and Drug Administration (FDA) 1
T Trend Reporting under the EU MDR EU Medical Device Regulations 44
L MDR (Medical Device Reporting) under 21 CFR Part 803 Other US Medical Device Regulations 5
B Is there a MDR (Medical Device Reporting) equivalent in China ? China Medical Device Regulations 1
C New draft MDR reporting guidance Other US Medical Device Regulations 1
R MDR Reporting for event with RUO instrument 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
S MDR (Medical Device Reporting) Justification - Canned Statements Other US Medical Device Regulations 2
E Post Inspection Regulatory Action by US FDA - 21 CFR 803 - MDR Reporting Other US Medical Device Regulations 4
E Reporting Requirements - Report to US FDA if a MDR event happened outside US Other US Medical Device Regulations 3
L Interpretation of MDR Reportable Event - FDA guidance on Medical Device Reporting 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 16
W MDR Reporting - How to interpret requirements for MDR Other US Medical Device Regulations 14
Q When to File an MDR (Medical Device Reporting) For Incidents in Foreign Countries 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
M MDR Legal requirements EU Medical Device Regulations 0
Ed Panek MDR Audit Comment EU Medical Device Regulations 9
J Minor / Major findings with MDR EU Medical Device Regulations 4
R MDR Clinical Evaluation Plan EU Medical Device Regulations 4
R EU MDR Article 22 applicability of medical device with components. EU Medical Device Regulations 4
K Qualified MDR internal auditor? EU Medical Device Regulations 7
J Clinical Evaluation For Instruments - Class I, Class Ir, Class Im, and Class IIa - EU MDR EU Medical Device Regulations 4
G MDR Translation Requirements in Member States EU Medical Device Regulations 8
R General Obligations of Importers under EU MDR EU Medical Device Regulations 2
V MDR – Article 65 – Clinical investigations on minors EU Medical Device Regulations 6
I MDR Audit: NC per MDR Annex IX Administrative provisions 8. EU Medical Device Regulations 6
R Device with WiFi communication - EU MDR vs. EU RED EU Medical Device Regulations 2
L Clinical Evaluation MDR EU Medical Device Regulations 5
J Common Specifications per MDR EU Medical Device Regulations 2
L Looking for SOP Template for UDI under MDR EU Medical Device Regulations 4
P MDR Article 27 - UDI - Keeping records of UDIs EU Medical Device Regulations 1
Q Class I Medical Device with measuring function (UK MDR 2002 / MDD) UK Medical Device Regulations 3
F EU MDR Quality Plan template CE Marking (Conformité Européene) / CB Scheme 3
K MDR EU Medical Device Regulations 3
B Clinical Investigation in Denmark for MDR EU Medical Device Regulations 2
R EU MDR Remote vs In-Person Audit EU Medical Device Regulations 1
H IVDR & MDR CE Marking (Conformité Européene) / CB Scheme 6
H Update to UK MDR 2002 No. 618 UK Medical Device Regulations 2
A MDR article 54(2) "modifying" EU Medical Device Regulations 1
K MDD to MDR Other Medical Device and Orthopedic Related Topics 4
S EU MDR-Replacement Parts when not manufacture is not yet EU MDR compliant US Medical Device Regulations 2
dgrainger Informational Consolidated UK MDR 2002 UK Medical Device Regulations 0
E MDR Labelling & Languages EU Medical Device Regulations 2
K Subcontractors Providing Services Under MDD or MDR need ISO 13485 from EU Notified Body? CE Marking (Conformité Européene) / CB Scheme 8
D MDR: Active medical device or not? Medical Information Technology, Medical Software and Health Informatics 14
T Justification of use of non-harmonized standards for MDR conformity Other Medical Device Related Standards 12
F The meaning of MDR Article 61 6(b) EU Medical Device Regulations 1
S Repackaging and translation under MDR EU Medical Device Regulations 14
Ed Panek MDR Delays Poll EU Medical Device Regulations 0
S Meaning of Article 61 (10) of the MDR 2017/745 EU Medical Device Regulations 3
T MDR PMS System Vs. Plan Vs. Procedures EU Medical Device Regulations 3
D MDR classification and DoC requirements Medical Device and FDA Regulations and Standards News 3
N Shelf life guidance EU MDR, class 1 and class 2 Medical Devices EU Medical Device Regulations 2