Measurement Capability vs. Specification Tolerance

Mike S.

Happy to be Alive
Trusted Information Resource
#1
A specification for a widget is 10 +/- 1 units and the specification also requires that "measurement capability meets or exceeds a 4:1 ratio between the product tolerance and the certified accuracy of the measurement device". Measuring device "A" has a certified accuracy of +/- 0.5 units.

Does measuring device "A" meet the specification?

Or does the measuring device need to have a certified accuracy of less than or equal to +/- 0.25 units?

Can you provide a published source that clearly demonstrates the correct technique?
 
Elsmar Forum Sponsor

optomist1

A Sea of Statistics
Trusted Information Resource
#2
Hi Mike,

It appears as though you are asking two questions:
1) What is the required accuracy of the measurment device. i.e accuracy is the difference between observed accuracy and the true accuracy or measurment as referenced to a master part or to a national bureau of standards.
2) What is the resolution of the measuring device - i.e. resolution is the smallest reading that an inspection instrument or porcess can reliably display.
You say a 4 to 1 ratio, I am only familair with a 10 to 1 ratio regarding the spec. so if the spec tolerance is 2 units the instrument must be able to reliably measure to 0.2 of a unit.

Have you performed a MSA on this instrument or set up?

Hope this helps..

Marty
 

Mike S.

Happy to be Alive
Trusted Information Resource
#3
Op -- thanks for the reply, but you're making it more difficult that it has to be. All the info needed is there, the question is in the specifics of how "the 4:1 ratio between the product tolerance and the certified accuracy of the measurement device" is calculated.
 

Ninja

Looking for Reality
Staff member
Super Moderator
#4
'K, so I'm scratching my head thinking to myself..."What's the catch?"

The tolerance is 2 units.
The certified accuracy is 0.5 units.
2 divided by 0.5 equals 4

What am I missing?

Can I provide a published source...ummmm...my 5th grade math book.
And when I have thoughts like that with all the smart folks around here, I figure I'm missing something significant...:eek:

I waited all day to see if there was a post to show me what I was missing but it didn't come.
And I figure you are asking a serious question...So....what's the catch?

Is an auditor or customer or coworker insisting that the tolerance is really 1 instead of 2?
 

Mike S.

Happy to be Alive
Trusted Information Resource
#5
It really does seem simple, doesn't it? But this is not intended as a trick -- just a "sanity check". Okay, here's the issue. I will try to ask the question without biasing the answer:

You could do this one of two ways. The spec. tolerance is 2 units wide (+/- 1). The measurement device certified accuracy is +/- 0.5 units. A 4:1 ratio is required.

You divide the total spec. tolerance (2 units) by 4 and get 0.5. Does that mean that (A) a measurement device that has a certified accuracy of +/- 0.5 is what you need as a minimum or (B) a measurement device that has a certified accuracy of +/- 0.25 is what you need as a minimum?

Does the question seem clear now?
 
N

NumberCruncher

#6
Hi Ninja

I don't agree with your calculation. You have divided a range by a semi range.
Spec = 1/-1 hence range = 1-(-1) = 2
Measurement instrument spec = +/-0.5 hence range = 0.5-(-0.5) = 1
Ratio = 2/1 = 2

It may help to draw a simple diagram. Mark three equidistant lines

11__________



10__________



9___________

Now measure the lines and get a reading of 10+/- 0.5


11__________

10.5.............

10__________

9.5...............

9___________

I think you can see that the instrument uncertainty takes up half of the specification.

Any other contributions from someone who knows for sure?

NC
 

Ninja

Looking for Reality
Staff member
Super Moderator
#7
Ahhh, that's what I was missing...the +/- symbol on the accuracy. Thanks NumberCruncher!
 

Mike S.

Happy to be Alive
Trusted Information Resource
#9
The 4:1 ratio for spec. tolerance vs. measurement accuracy is a fairly common requirement referenced in several standards, from old MIL-STD-45662 to current standards like SAE ARP 9013, although the exact wording varies slightly from place to place. The wording I used is from ARP 9013.

As you can see, it is a simple enough question on the surface...
 
Thread starter Similar threads Forum Replies Date
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
D O Ring capability and measurement - What is the automotive 'norm' for capability studies on O Rings? General Measurement Device and Calibration Topics 3
S Understanding control chart and measurement capability Statistical Analysis Tools, Techniques and SPC 2
M Capability of a Vibration Measurement Gauge Quality Tools, Improvement and Analysis 3
Q JMP Measurement Capability Analysis Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
A Yearly MSA (Measurement System Analysis) & Capability Study - TS 16949 Requirements Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
R Converting Best Measurement (BMC) to Calibration & Measurement Capability (CMC) Nonconformance and Corrective Action 2
T Comparing Two Measurement Systems in terms of Accuracy and Capability Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
N Customer Requirement exceeded my BMC (Best Measurement Capability) Measurement Uncertainty (MU) 3
A Score capability of measurement of gauge Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
B Calculating BMC (best measurement capability) for Fluke 8508A and HP 3458A Measurement Uncertainty (MU) 5
E Measurement System Capability with a Continuous Measurement System Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 22
R BMC (Best Measurement Capability) exceeding Tolerance! Measurement Uncertainty (MU) 8
T Suitability and Capability of Measurement Equipment Capability, Accuracy and Stability - Processes, Machines, etc. 8
S The Best Measurement Capability (BMC) of a Small Calibration Laboratory Measurement Uncertainty (MU) 9
B Test Method Qualification - Measurement Capability - What else is involved? Reliability Analysis - Predictions, Testing and Standards 1
D Measurement Uncertainty effects on Process Capability and Tolerances Measurement Uncertainty (MU) 7
E MSA vs. Measurement Capability Analysis - What are the differences? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 33
M MSA Chicken and Egg - Process Capability and Measurement System Capability Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
S How to Determine a Machine Capability? Shaft Fit Tolerance Measurement Results Statistical Analysis Tools, Techniques and SPC 2
M Measurement Uncertainty in Optical Microscopy Measurement Uncertainty (MU) 1
B ISO 6508 and portable hardness measurement instruments General Measurement Device and Calibration Topics 0
Ron Rompen Measurement of residual magnetism General Measurement Device and Calibration Topics 2
M Surgical angle measurement guide device with an application software Medical Device and FDA Regulations and Standards News 1
M Disabling measurement data during fault conditions IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J Most versatile measurement device General Measurement Device and Calibration Topics 2
P Surface scratch measurement General Measurement Device and Calibration Topics 1
N Drawing tolerance vs. Measurement device General Measurement Device and Calibration Topics 7
Ron Rompen MSA on automated measurement system - Multiple Step Vision System Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
P Thermo-Hygrometer - Measurement of 10DegC @10%RH ISO 17025 related Discussions 4
hogheavenfarm GDT Flatness measurement question Inspection, Prints (Drawings), Testing, Sampling and Related Topics 10
K Software Validation for Measurement Tools used in Process Validation ISO 13485:2016 - Medical Device Quality Management Systems 2
K Best Measurement Systems Demos in California? General Measurement Device and Calibration Topics 1
M Measurement Error - How to determine what is acceptable? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
K Process Mapping - Inputs/Outputs/Detail Activities/Control points/Measurement? Process Maps, Process Mapping and Turtle Diagrams 4
B IEC 60601-2-10 - Accuracy of Pulse Parameters - Required Measurement Uncertainty IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
N Program or application for standard time measurement in industry from video surveillance? Human Factors and Ergonomics in Engineering 1
M Defining frequency of measurement tools callibration Calibration and Metrology Software and Hardware 3
T How to “validate/qualify” a new measurement equipment? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
Bev D Verification and Validation of Measurement Systems Misc. Quality Assurance and Business Systems Related Topics 0
F How do you add accuracies for tools with different measurement devices? General Measurement Device and Calibration Topics 1
M Viable count Measurement ISO Standard Other Medical Device Related Standards 1
J Gauge R&R on a torque measurement Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 11
W Air Quality Measurement Hardware and Software General Measurement Device and Calibration Topics 11
O Dimension Measurement Tool Recommendation General Measurement Device and Calibration Topics 21
G Does pitch/increment/resolution of a ruled scale apply to measurement uncertainty as line item? Measurement Uncertainty (MU) 10
Marco Bernardi Plane measurement with a 3D touch machine like a CMM. Calibration and Metrology Software and Hardware 3
Marco Bernardi Diameters measurement with 3D touch machine like a CMM Calibration and Metrology Software and Hardware 4
R Uncertainty in measurement larger than tolerance Measurement Uncertainty (MU) 2
R Measurement Uncertainties Budget for Thermometer Measurement Uncertainty (MU) 6