Measurement Uncertainties Budget for Thermometer
Hello all,
I am trying to create uncertainty budgets for our calibration equipment at our facility. I am new to creating these and was wondering if I could get some advice on the following:
Lets say I want to create an uncertainties budget for a thermometer. I want to use the thermometer in various environments (ovens, water baths, etc). Should I perform repeatability tests for use in both medias (air and water) since they will have differing results depending on it, and add each results to the uncertainty, or would adding both results be over estimating the uncertainty of the thermometer?
Thank you
indubioush
How would a thermometer have different results from air or water?
Usually, statement about equipment accuracy come from the manufacturer.
I was under the impression that equipment accuracies from the manufacturer would be considered a Type B uncertainty, whereas performing a statistical analysis, such as repeatability, would be Type A uncertainty, which is what I am referring to.
I plan on using this thermometer in both air and water mediums. Just curious if I should perform analysis in both mediums or if one is sufficient.
dwperron
You are correct about Type A vs. Type B.
Tell me about the thermometer.
Liquid in glass? If so, is it full or partial immersion?
Or is this a temperature probe connected to a temperature meter?
dwperron
It is a temperature probe connected to a temperature meter.
Doing repeatability in both media is a good idea. The RTD elements in probes come in different sizes, and may not be totally immersed in liquids, so that thermal wicking can occur. Not so much a problem in air unless you are only inserting the probe tip into a chamber.
Combining the uncertainty contributors is a philosophical question. Many normalize the results and combine them RSS, this is legitimate. Others take the more conservative approach of adding the contributors together, giving you a "worst case" uncertainty. In your case I think I would have two separate uncertainties for liquids and air (if there is a difference...) instead of combining them. That would exaggerate your uncertainty budget.
R Looking for Practical Advice in Managing Measurement Uncertainties Measurement Uncertainty (MU) 3
E Oscilloscope Measurement Uncertainties - MU Components for Vertical Deflection Measurement Uncertainty (MU) 3
A Do we still need to Calculate Measurement Uncertainties (MU)? Measurement Uncertainty (MU) 6
C Temperature Measurement Uncertainties - Verify a PRT at 0?C specified ?0.02?C General Measurement Device and Calibration Topics 4
D Micrometer Uncertainty - Best Measurement Uncertainties for O.D. Micrometers General Measurement Device and Calibration Topics 2
B Building my own measurement equipment General Measurement Device and Calibration Topics 3
P Test Method Validation (TMV) for all Measurement Methods in Rec/Inspection Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
J Keyence Measurement Systems General Measurement Device and Calibration Topics 3
Q Measurement Equipment Revocation - Looking for a Disposal Form with Risk Assessment IATF 16949 - Automotive Quality Systems Standard 10
C IVD Metrological traceability without a reference measurement procedure General Measurement Device and Calibration Topics 11
E Any template/ form of Monitoring and Measurement of Processes and product to ISO 13485? ISO 13485:2016 - Medical Device Quality Management Systems 1
A Flexible "Ring" Roundness Measurement General Measurement Device and Calibration Topics 2
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
I QMS monitoring, measurement, analysis and evaluation requirement - Template ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
F Physical Part Measurement General Measurement Device and Calibration Topics 6
B ISO 6508 and portable hardness measurement instruments General Measurement Device and Calibration Topics 0
Ron Rompen Measurement of residual magnetism General Measurement Device and Calibration Topics 2
M Surgical angle measurement guide device with an application software Medical Device and FDA Regulations and Standards News 4
M Disabling measurement data during fault conditions IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J Most versatile measurement device General Measurement Device and Calibration Topics 2
P Surface scratch measurement General Measurement Device and Calibration Topics 1
P Thermo-Hygrometer - Measurement of 10DegC @10%RH ISO 17025 related Discussions 4
hogheavenfarm GDT Flatness measurement question Inspection, Prints (Drawings), Testing, Sampling and Related Topics 10
K Software Validation for Measurement Tools used in Process Validation ISO 13485:2016 - Medical Device Quality Management Systems 2
K Best Measurement Systems Demos in California? General Measurement Device and Calibration Topics 1
K Process Mapping - Inputs/Outputs/Detail Activities/Control points/Measurement? Process Maps, Process Mapping and Turtle Diagrams 4
N Program or application for standard time measurement in industry from video surveillance? Human Factors and Ergonomics in Engineering 1
M Defining frequency of measurement tools callibration Calibration and Metrology Software and Hardware 3
T How to “validate/qualify” a new measurement equipment? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
Bev D Verification and Validation of Measurement Systems Misc. Quality Assurance and Business Systems Related Topics 0
F How do you add accuracies for tools with different measurement devices? General Measurement Device and Calibration Topics 1
M Viable count Measurement ISO Standard Other Medical Device Related Standards 1
W Air Quality Measurement Hardware and Software General Measurement Device and Calibration Topics 11
O Dimension Measurement Tool Recommendation General Measurement Device and Calibration Topics 21
Marco Bernardi Plane measurement with a 3D touch machine like a CMM. Calibration and Metrology Software and Hardware 6
Marco Bernardi Diameters measurement with 3D touch machine like a CMM Calibration and Metrology Software and Hardware 4
O Performance Measurement ISO 9001: 2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
O Monitoring performance - How do I determine performance measurement basis within my organization? Misc. Quality Assurance and Business Systems Related Topics 4
G What information to put on measurement Dimensional Results APQP and PPAP 7
D O Ring capability and measurement - What is the automotive 'norm' for capability studies on O Rings? General Measurement Device and Calibration Topics 3
F Electrical Metrology (Source/Generation and Measurement) General Measurement Device and Calibration Topics 1