I need some explanation if someone can help please.
ISO/IEC 17025:2005 clause 5.10.4.2 "When statements of compliance are made, the uncertainty of measurement shall be taken into account" . What is the interpretation, let's say I have CMM uncertainty .040mm and I am measuring feature that has tolerance +/- .015 deviate .008 from the nominal. Customer remeasure the same feature and result is .042 mm.
Now
#1 Is it appropriate for my lab to make a statement that can hold +/-.015mm tolerance when CMM uncertainty is .040 mm?
#2 Can I tell the customer that should take into account the uncertainty for the difference in the results?
#3 If I measure feature with same tolerance and the deviation is .020mm, Would that be acceptable taking into account .040mm my CMM uncertainty?
Any input would be greatly appreciated. Thanks!!!
 
Elsmar Forum Sponsor
G

George Weiss

#2
Re: Measurment Uncertainty Explaination per ISO/IEC 17025:2005 clause 5.10.4.2

It sounds like you have a CMM uncertainty of +/- 0.40 something, and measuring a feature with +/-0.15 something tolerance, and found it to be + .008 something from nominal. In a great number of other threads it has been suggested that an accuracy/uncertainty of measurement would desirably be smaller, like one quarter to one tenth the attempted test value. You suggest you have an uncertainty of 40, and are testing an item with a test limit of 15. It sounds as if your test is upside down, or your stated question is.
Part of the question was taking uncertainties into account. I suggest looking at one of the many uncertainty documents, and looking at guard-banding to explain the use of the uncertainties a bit.

Uncertainty 8 pages @
http://www.uncertainty.com.au/The%20Drongo's%20Guide%20to%20Uncertainty.pdf
Agilent guardbanding for beginners @
http://www.agilent.com/metrology/pdf/guardband_beginners.pdf
NAMS UKAS M3003 @
http://www.ukas.com/library/Technical-Information/Pubs-Technical-Articles/Pubs-List/M3003.pdf
NIST tech. Note 1297 @
http://physics.nist.gov/Pubs/guidelines/TN1297/tn1297s.pdf
 

new_to_this

Involved In Discussions
#3
Re: Measurment Uncertainty Explaination per ISO/IEC 17025:2005 clause 5.10.4.2

George
Thanks for reply. I am not sure if you understood my questions. I am sorry I should of explain little better. In our scope of accreditation CMM BMC is (40 +.003L)um. We build gages so we measure and certify them. We tell the customer that we can hold +/- .015mm tolerance on the Datum features. I red about the guard bend and to tell you the truth I don't understand how to apply it in this case. I am not un Uncertainty expert so that's why I ask these questions in this forum.
Now, what is the interpretation, with above stated uncertainty,I am measuring feature that has tolerance +/- .015mm.In my measurement feature deviation is .008 from the nominal(gage certification pass)Customer re-measure the same feature and the deviation is .042 mm from the nominal (gage certification fail).
Questions(I think they are simple):
#1 Is it appropriate for my lab to make a statement that can hold +/-.015mm tolerance when CMM uncertainty is .040 mm?
#2 Can I tell the customer that should take into account the uncertainty for the difference in the results?
#3 If I measure feature with same tolerance and the deviation is .020mm, Would that be acceptable taking into account .040mm my CMM uncertainty?

 
J

Joe222

#4
Re: Measurment Uncertainty Explaination per ISO/IEC 17025:2005 clause 5.10.4.2

I need some explanation if someone can help please.
ISO/IEC 17025:2005 clause 5.10.4.2 "When statements of compliance are made, the uncertainty of measurement shall be taken into account" . What is the interpretation, let's say I have CMM uncertainty .040mm and I am measuring feature that has tolerance +/- .015 deviate .008 from the nominal. Customer remeasure the same feature and result is .042 mm.
Now
#1 Is it appropriate for my lab to make a statement that can hold +/-.015mm tolerance when CMM uncertainty is .040 mm?
#2 Can I tell the customer that should take into account the uncertainty for the difference in the results?
#3 If I measure feature with same tolerance and the deviation is .020mm, Would that be acceptable taking into account .040mm my CMM uncertainty?
Any input would be greatly appreciated. Thanks!!!
The above statement simply means that if you say that a gauge is in tolerance, you must consider the measurement uncertainty. Lets say that you were measuring 1 V with a tolerance of ± 0.1 V and your measurement uncertainty was ± 0.025 V. If the piece read 1.09 V, you really don't know that it is in tolerance, it could be a false pass based on your measurement uncertainty.
 
Thread starter Similar threads Forum Replies Date
M Measurement Uncertainty in Optical Microscopy Measurement Uncertainty (MU) 1
B IEC 60601-2-10 - Accuracy of Pulse Parameters - Required Measurement Uncertainty IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
G Does pitch/increment/resolution of a ruled scale apply to measurement uncertainty as line item? Measurement Uncertainty (MU) 10
R Uncertainty in measurement larger than tolerance Measurement Uncertainty (MU) 2
K ISO 17025:2017 clause 7.6.2 - Performing calibration of its own equipment shall evaluate the measurement uncertainty ISO 17025 related Discussions 6
G Reporting measurement uncertainty for custom items Measurement Uncertainty (MU) 2
K Measurement Uncertainty Budget (MU) of for Ultrasonic Thickness gauge Measurement Uncertainty (MU) 1
G Can Measurement Machine Bias be part of Uncertainty? General Measurement Device and Calibration Topics 1
D Difference between uncertainty and expanded uncertainty of measurement General Measurement Device and Calibration Topics 1
J Internal Laboratory Scope and Measurement Uncertainty (MU) IATF 16949 - Automotive Quality Systems Standard 1
C MU (Measurement Uncertainty) in an Internal Calibration Laboratory Measurement Uncertainty (MU) 1
W Dimensional Measurement of 10% accuracy and 25% uncertainty standard General Measurement Device and Calibration Topics 6
B Measurement Uncertainty using Minitab Measurement Uncertainty (MU) 2
A Uncertainty Measurement - What this accuracy specification "1% + 10^5": means? General Measurement Device and Calibration Topics 2
Marc Measurement Uncertainty Discussion Sub-Forum General Measurement Device and Calibration Topics 0
C Measurement Uncertainty fluctuates half a tick mark (20 millionths of an inch) Measurement Uncertainty (MU) 1
R MU (Measurement Uncertainty) of a Tensile Tester (e.g. Zwick Roell) Measurement Uncertainty (MU) 4
H Uncertainty Measurement Calculation in Chemical Testing Measurement Uncertainty (MU) 1
S MU (Measurement Uncertainty) for Oscilloscope Calibration Measurement Uncertainty (MU) 2
T Method Repeatability for MU (Measurement Uncertainty) Measurement Uncertainty (MU) 1
S Combined Uncertainty of an Indirect Measurement Measurement Uncertainty (MU) 8
P Measurement Uncertainty and Equipment Accuracy Measurement Uncertainty (MU) 2
N Uncertainty calculation for a voltage measurement Measurement Uncertainty (MU) 3
T Measurement of uncertainty for standard mechanical properties General Measurement Device and Calibration Topics 1
K New Gauges calibration certificate shows deviations from the measurement uncertainty IATF 16949 - Automotive Quality Systems Standard 5
A Calculating Combined Measurement Uncertainty - VDA 5 Measurement Uncertainty (MU) 1
T Uncertainty of Measurement vs. Tolerance Measurement Uncertainty (MU) 5
A Calibration Method and Measurement Uncertainty General Measurement Device and Calibration Topics 5
Geoff Cotton Measurement Uncertainty (MU) of HR-150A Rockwell Hardness Tester Measurement Uncertainty (MU) 1
Jerry Eldred When Does Resolution of Measurement Standard Apply in Uncertainty Calculations General Measurement Device and Calibration Topics 14
E Uncertainty Measurement for Water COD, TDS, TSS & pH Measurement Uncertainty (MU) 1
D General Procedure for Uncertainty Measurement Measurement Uncertainty (MU) 2
S Measurement Uncertainty for Neutral Salt Spray test to BS EN ISO 9227:2006 General Measurement Device and Calibration Topics 4
S Organizing Uncertainty Budgets for different Measurement Equipment Measurement Uncertainty (MU) 4
B Length Measure - Interpreting Calibration Results and Measurement Uncertainty (MU) Measurement Uncertainty (MU) 3
R Procedure for Measurement Uncertainty Estimation for Electrical Measurements Measurement Uncertainty (MU) 2
R Signal Conditioning Card - Calibration Measurement Uncertainty Measurement Uncertainty (MU) 5
K AATCC 81 pH test - How to calculate MU (Measurement Uncertainty) Measurement Uncertainty (MU) 2
optomist1 Definition Measurement Uncertainty - Standard definitions for Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 8
S Uncertainty Estimation in Temperature Measurement by K type Thermocouples Measurement Uncertainty (MU) 7
DietCokeofEvil ILAC-g8:03/2009 - Reporting Results with Measurement Uncertainty General Measurement Device and Calibration Topics 7
N Rounding the Numerical Value of the Measurement Uncertainty Measurement Uncertainty (MU) 2
A Uncertainty of Measurement for Calibration Laboratory? Measurement Uncertainty (MU) 8
C How to interpret Measurement Uncertainty (MU) Measurement Uncertainty (MU) 5
S ISO 19036 - Measurement Uncertainty for Microbiology Measurement Uncertainty (MU) 7
P Measurement Uncertainty Calculations Spreadsheet help General Measurement Device and Calibration Topics 4
L Measurement Uncertainty using Control Chart Measurement Uncertainty (MU) 2
P Uncertainty Measurement for Colour, Odour, Taste, Turbidity, Total Suspended Solids Calibration Frequency (Interval) 2
T Uncertainty Measurement on Kinematic Viscosity of Fuel Oil General Measurement Device and Calibration Topics 1
N Measurement Uncertainty - When is Normal and when is Triangular Distribution used? Measurement Uncertainty (MU) 3

Similar threads

Top Bottom