Measurement Uncertainty (ISO 17025) - Seeking Lay Person Understanding

cpine

Registered Visitor
I am a little confused. With respect to measurement uncertainty, if an outside lab is performing calibrations, is it necessary to perform them internally.

The reason I ask is I am not quite sure that I completely understand the requirement.

Can someone offer an explanaton that offers the lay person a better understanding.

We are looking at ISO 17025. The company is a distributor and as such has a inspection lab of which once again we source our calibration.


Sincerely,

C. Pine
 

Miner

Forum Moderator
Leader
Admin
Could you clarify your situation? Is your company the gauge user that is outsourcing calibration to a company that subcontracts the actual calibration? Or receiving gauges for calibration that you then subcontract? Or something totally different?
 
R

Randy Stewart

You do not have to calibrate again, however the Uncertainty given by the calibration lab must be reported.
 

cpine

Registered Visitor
Measurement Uncertainty

The situation is that I am sending test equimpent out to be calibrated.

This is a practice that is followed by all corporate business units, and what concerns is that those that are ISO 17025 registered clearly define measurement as being performed. They actually include the formula in the quality manual (which is what is confusing me). If one sources calibration, why include a procedure?

I believe that your answers are correct however I'm looking for ammunition for corporate.

Thank you so much for your time and help


Chuck
 

Jim Wynne

Leader
Admin
cpine said:
The situation is that I am sending test equimpent out to be calibrated.

This is a practice that is followed by all corporate business units, and what concerns is that those that are ISO 17025 registered clearly define measurement as being performed.
You're saying that your company has business units that are 17025 accredited, and "clearly define measurement as being performed?" Did you mean to say "calibration" is being performed?
cpine said:
They actually include the formula in the quality manual (which is what is confusing me). If one sources calibration, why include a procedure?
The formula for what?

cpine said:
I believe that your answers are correct however I'm looking for ammunition for corporate.
You need to help us to help you by clarifying the issue that's concerning you. It sounds like what you're saying is that calibration is being outsourced by business units that are perhaps qualified to do the calibrations internally. What is corporate doing that you need to correct?
 

cpine

Registered Visitor
Let me start over

I do apologize for being less that clear.

Allow me to start over.

Even tho all business units are required to source calibration, the majority of those units state that measurement uncertainty is being performed by each unit in addition to the accrediated lab they are currently using for calibration.

Once again and as always, Thanks for all of your support and guidence

Chuck
 

Jim Wynne

Leader
Admin
cpine said:
I do apologize for being less that clear.

Allow me to start over.

Even tho all business units are required to source calibration, the majority of those units state that measurement uncertainty is being performed by each unit in addition to the accrediated lab they are currently using for calibration.

Once again and as always, Thanks for all of your support and guidence

Chuck
OK, thanks for the clarification. Do they actually calculate uncertainty, and you feel the effort is redundant? If so, and you want to convince the bosses that they should stop, the best way to do that is to estimate how much it costs for them to do it and demonstrate that it's unnecessary because the outside sources are already doing it. Most bosses hate wasted money, but they might just get a glazed, far-away look in their eyes if you start using Greek phrases like "measurement uncertainty."
 

Miner

Forum Moderator
Leader
Admin
cpine said:
I do apologize for being less that clear.

Allow me to start over.

Even tho all business units are required to source calibration, the majority of those units state that measurement uncertainty is being performed by each unit in addition to the accrediated lab they are currently using for calibration.

Once again and as always, Thanks for all of your support and guidence

Chuck

I am still a little unclear about your situation, but hopefully I am on-target with the response.

You appear to be outsourcing the calibration of your gauges, and your corporate requires that you calculate the measurement uncertainty of the gauges.

If this is correct, the following should answer your question.

Measurement uncertainty is a combination of uncertainties from many sources. One part of this would be provided by the external calibration provider. This part is that uncertainty inherent to the gauge itself and how it was calibrated.

However, there are other factors outside of the control of the calibration provider that you must quantify yourself, such as repeatability, reproducibility, effect of temperature deviation from standard conditions, etc. This is probably where your corporate formula comes in.

I have attached an article that addresses the general concept.
 

Attachments

  • Measurement Uncertainty.pdf
    43.8 KB · Views: 699

cpine

Registered Visitor
The End

Thank you everone for your patience and help. I believe thats a wrap.


What a great forum.


Chuck
 
Thread starter Similar threads Forum Replies Date
K ISO 17025:2017 clause 7.6.2 - Performing calibration of its own equipment shall evaluate the measurement uncertainty ISO 17025 related Discussions 6
S Measurement Uncertainty for Neutral Salt Spray test to BS EN ISO 9227:2006 General Measurement Device and Calibration Topics 4
S ISO 19036 - Measurement Uncertainty for Microbiology Measurement Uncertainty (MU) 7
N Measurement Uncertainty Explanation per ISO/IEC 17025:2005 clause 5.10.4.2 ISO 17025 related Discussions 3
B Is Measurement Uncertainty required for a captive lab according to ISO 17025 ISO 17025 related Discussions 1
Q Measurement Uncertainty Calculations and Microbiology - ISO 8199 or ISO 19036 Measurement Uncertainty (MU) 5
J Measurement Uncertainty (MU) - Genetic Testing Lab - Lack of suitable CRM - ISO 17025 ISO 17025 related Discussions 6
Hershal Measurement Uncertainty - Need help with ANS/ISO/IEC 17025 and MU ISO 17025 related Discussions 0
Hershal Measurement Uncertainty Training - ANS/ISO/IEC 17025, ISO/IEC 17020 ISO 17025 related Discussions 0
B Does ISO 17025 Require Point By Point Measurement Uncertainty on Reports? ISO 17025 related Discussions 10
N QUERIES On ISO 17025 AUDIT - Measurement Uncertainty Budget, etc. ISO 17025 related Discussions 1
M Measurement Uncertainty in Optical Microscopy Measurement Uncertainty (MU) 1
B IEC 60601-2-10 - Accuracy of Pulse Parameters - Required Measurement Uncertainty IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
G Does pitch/increment/resolution of a ruled scale apply to measurement uncertainty as line item? Measurement Uncertainty (MU) 10
R Uncertainty in measurement larger than tolerance Measurement Uncertainty (MU) 2
G Reporting measurement uncertainty for custom items Measurement Uncertainty (MU) 2
K Measurement Uncertainty Budget (MU) of for Ultrasonic Thickness gauge Measurement Uncertainty (MU) 1
G Can Measurement Machine Bias be part of Uncertainty? General Measurement Device and Calibration Topics 1
D Difference between uncertainty and expanded uncertainty of measurement General Measurement Device and Calibration Topics 1
J Internal Laboratory Scope and Measurement Uncertainty (MU) IATF 16949 - Automotive Quality Systems Standard 1
C MU (Measurement Uncertainty) in an Internal Calibration Laboratory Measurement Uncertainty (MU) 1
W Dimensional Measurement of 10% accuracy and 25% uncertainty standard General Measurement Device and Calibration Topics 6
B Measurement Uncertainty using Minitab Measurement Uncertainty (MU) 2
A Uncertainty Measurement - What this accuracy specification "1% + 10^5": means? General Measurement Device and Calibration Topics 2
Marc Measurement Uncertainty Discussion Sub-Forum General Measurement Device and Calibration Topics 0
C Measurement Uncertainty fluctuates half a tick mark (20 millionths of an inch) Measurement Uncertainty (MU) 1
R MU (Measurement Uncertainty) of a Tensile Tester (e.g. Zwick Roell) Measurement Uncertainty (MU) 4
H Uncertainty Measurement Calculation in Chemical Testing Measurement Uncertainty (MU) 1
S MU (Measurement Uncertainty) for Oscilloscope Calibration Measurement Uncertainty (MU) 2
T Method Repeatability for MU (Measurement Uncertainty) Measurement Uncertainty (MU) 1
S Combined Uncertainty of an Indirect Measurement Measurement Uncertainty (MU) 8
P Measurement Uncertainty and Equipment Accuracy Measurement Uncertainty (MU) 2
N Uncertainty calculation for a voltage measurement Measurement Uncertainty (MU) 3
T Measurement of uncertainty for standard mechanical properties General Measurement Device and Calibration Topics 1
K New Gauges calibration certificate shows deviations from the measurement uncertainty IATF 16949 - Automotive Quality Systems Standard 5
A Calculating Combined Measurement Uncertainty - VDA 5 Measurement Uncertainty (MU) 1
T Uncertainty of Measurement vs. Tolerance Measurement Uncertainty (MU) 5
A Calibration Method and Measurement Uncertainty General Measurement Device and Calibration Topics 5
Geoff Cotton Measurement Uncertainty (MU) of HR-150A Rockwell Hardness Tester Measurement Uncertainty (MU) 1
Jerry Eldred When Does Resolution of Measurement Standard Apply in Uncertainty Calculations General Measurement Device and Calibration Topics 14
E Uncertainty Measurement for Water COD, TDS, TSS & pH Measurement Uncertainty (MU) 1
D General Procedure for Uncertainty Measurement Measurement Uncertainty (MU) 2
S Organizing Uncertainty Budgets for different Measurement Equipment Measurement Uncertainty (MU) 4
B Length Measure - Interpreting Calibration Results and Measurement Uncertainty (MU) Measurement Uncertainty (MU) 3
R Procedure for Measurement Uncertainty Estimation for Electrical Measurements Measurement Uncertainty (MU) 2
R Signal Conditioning Card - Calibration Measurement Uncertainty Measurement Uncertainty (MU) 5
K AATCC 81 pH test - How to calculate MU (Measurement Uncertainty) Measurement Uncertainty (MU) 2
optomist1 Definition Measurement Uncertainty - Standard definitions for Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 8
S Uncertainty Estimation in Temperature Measurement by K type Thermocouples Measurement Uncertainty (MU) 7
DietCokeofEvil ILAC-g8:03/2009 - Reporting Results with Measurement Uncertainty General Measurement Device and Calibration Topics 7

Similar threads

Top Bottom