I am a newbie to the uncertainty world. I started working in our lab a couple years ago and noticed that we should be working to ISO standards and not to the 1960’s mil specs. Long story short, I need a little guidance to see if I am on the right trail. We work in inch, so everything on my budget is in inches.
I also made an assumption that a 0-1 micrometer with a .001 resolution could be raised to a .0005 resolution due to the large space between the graduations on the barrel.
I have enclose the budget for people to look at, and to give me some feed back.
 

Attachments

Elsmar Forum Sponsor

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#2
The budget looks reasonable to me.....seems like you took into account at least the major contributors of MU.....

The space between the marks of the barrel might be best left alone.....if you let folks measure using the dead space, you now add parallax error, which will be equal to the resolution error (0.0005"), and so the MU will likely increase.

Hershal
 
R

rich.dittmanjr

#3
Resolution Error

:thanx: Thank you for your feedback Hershal. I feel a little more confident moving forward on other gages now.
The resolution of the micrometer still bugs me. I have read many different articles saying you can use the area between the vernier etch marks as a half increment. I also read you do not do that.
At this plant they normally round to the .001 and don’t estimate the distance between the marks. I feel safe because in this case, the calibration uncertainty based on .0005, would then be safe.
On the other hand, the Manufacturing team should not be using gages that do not meet the 10% resolution rule.
 
R

rich.dittmanjr

#4
Is this right?

I sent a 0-1 inch vernier micrometer with .001 resolution to the calibration lab that we use for other gages. I did this to get a start on setting up our uncertainty and also do a little inter-lab comparison. When I saw what they had for stated uncertainty, I am really questioning the credibility of this lab. I cannot see how you can get +/- .000083 uncertainty from a vernier micrometer. Am I missing something?
 

Attachments

Thread starter Similar threads Forum Replies Date
K Measurement Uncertainty Budget (MU) of for Ultrasonic Thickness gauge Measurement Uncertainty (MU) 1
S Measurement Uncertainty Budget (MU) of Ultrasonic Thickness gage? Measurement Uncertainty (MU) 4
J Measurement Uncertainty Budget for Comparative Dimensional Measurements Measurement Uncertainty (MU) 8
M Contributing Factors in a Supermicrometer Measurement Uncertainty Budget? Measurement Uncertainty (MU) 8
CalRich Measurement Uncertainty for Adjustable Thread Rings but No Budget Made Measurement Uncertainty (MU) 4
N QUERIES On ISO 17025 AUDIT - Measurement Uncertainty Budget, etc. ISO 17025 related Discussions 1
M Measurement Uncertainty in Optical Microscopy Measurement Uncertainty (MU) 1
B IEC 60601-2-10 - Accuracy of Pulse Parameters - Required Measurement Uncertainty IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
G Does pitch/increment/resolution of a ruled scale apply to measurement uncertainty as line item? Measurement Uncertainty (MU) 10
R Uncertainty in measurement larger than tolerance Measurement Uncertainty (MU) 2
K ISO 17025:2017 clause 7.6.2 - Performing calibration of its own equipment shall evaluate the measurement uncertainty ISO 17025 related Discussions 6
G Reporting measurement uncertainty for custom items Measurement Uncertainty (MU) 2
G Can Measurement Machine Bias be part of Uncertainty? General Measurement Device and Calibration Topics 1
D Difference between uncertainty and expanded uncertainty of measurement General Measurement Device and Calibration Topics 1
J Internal Laboratory Scope and Measurement Uncertainty (MU) IATF 16949 - Automotive Quality Systems Standard 1
C MU (Measurement Uncertainty) in an Internal Calibration Laboratory Measurement Uncertainty (MU) 1
W Dimensional Measurement of 10% accuracy and 25% uncertainty standard General Measurement Device and Calibration Topics 6
B Measurement Uncertainty using Minitab Measurement Uncertainty (MU) 2
A Uncertainty Measurement - What this accuracy specification "1% + 10^5": means? General Measurement Device and Calibration Topics 2
Marc Measurement Uncertainty Discussion Sub-Forum General Measurement Device and Calibration Topics 0
C Measurement Uncertainty fluctuates half a tick mark (20 millionths of an inch) Measurement Uncertainty (MU) 1
R MU (Measurement Uncertainty) of a Tensile Tester (e.g. Zwick Roell) Measurement Uncertainty (MU) 4
H Uncertainty Measurement Calculation in Chemical Testing Measurement Uncertainty (MU) 1
S MU (Measurement Uncertainty) for Oscilloscope Calibration Measurement Uncertainty (MU) 2
T Method Repeatability for MU (Measurement Uncertainty) Measurement Uncertainty (MU) 1
S Combined Uncertainty of an Indirect Measurement Measurement Uncertainty (MU) 8
P Measurement Uncertainty and Equipment Accuracy Measurement Uncertainty (MU) 2
N Uncertainty calculation for a voltage measurement Measurement Uncertainty (MU) 3
T Measurement of uncertainty for standard mechanical properties General Measurement Device and Calibration Topics 1
K New Gauges calibration certificate shows deviations from the measurement uncertainty IATF 16949 - Automotive Quality Systems Standard 5
A Calculating Combined Measurement Uncertainty - VDA 5 Measurement Uncertainty (MU) 1
T Uncertainty of Measurement vs. Tolerance Measurement Uncertainty (MU) 5
A Calibration Method and Measurement Uncertainty General Measurement Device and Calibration Topics 5
Geoff Cotton Measurement Uncertainty (MU) of HR-150A Rockwell Hardness Tester Measurement Uncertainty (MU) 1
Jerry Eldred When Does Resolution of Measurement Standard Apply in Uncertainty Calculations General Measurement Device and Calibration Topics 14
E Uncertainty Measurement for Water COD, TDS, TSS & pH Measurement Uncertainty (MU) 1
D General Procedure for Uncertainty Measurement Measurement Uncertainty (MU) 2
S Measurement Uncertainty for Neutral Salt Spray test to BS EN ISO 9227:2006 General Measurement Device and Calibration Topics 4
S Organizing Uncertainty Budgets for different Measurement Equipment Measurement Uncertainty (MU) 4
B Length Measure - Interpreting Calibration Results and Measurement Uncertainty (MU) Measurement Uncertainty (MU) 3
R Procedure for Measurement Uncertainty Estimation for Electrical Measurements Measurement Uncertainty (MU) 2
R Signal Conditioning Card - Calibration Measurement Uncertainty Measurement Uncertainty (MU) 5
K AATCC 81 pH test - How to calculate MU (Measurement Uncertainty) Measurement Uncertainty (MU) 2
optomist1 Definition Measurement Uncertainty - Standard definitions for Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 8
S Uncertainty Estimation in Temperature Measurement by K type Thermocouples Measurement Uncertainty (MU) 7
DietCokeofEvil ILAC-g8:03/2009 - Reporting Results with Measurement Uncertainty General Measurement Device and Calibration Topics 7
N Rounding the Numerical Value of the Measurement Uncertainty Measurement Uncertainty (MU) 2
A Uncertainty of Measurement for Calibration Laboratory? Measurement Uncertainty (MU) 8
C How to interpret Measurement Uncertainty (MU) Measurement Uncertainty (MU) 5
S ISO 19036 - Measurement Uncertainty for Microbiology Measurement Uncertainty (MU) 7

Similar threads

Top Bottom