Medical Device Cart
Hi Everyone, Im hoping to solicit clarification on the EUMDR requirements for using a cart with a patient monitor. The cart simply holds the monitor and allows the caregiver to become mobile with the parent device. We are currently looking to have a cart supplier provide us with evidence of compliance to the Machine directive and a CE marking.
Question: Is compliance to the Machinery Directive, CE Marking (from the supplier) and Validation of the cart with the parent device (from the device manufacturer) acceptable to satisfy the EUMDR requirements?
Our organization is considering treating this as a Class 1 and holding the DHF for the cart and essentially owning the cart. Taking ownership of the cart would induce a very large amount of internal compliance work. My gut tells me this is too much and its acceptable that the ownership of the cart reside with the cart manufacturer and not with owner of the parent device.
Help is appreciated
Ronen E
Problem Solver
Moderator
A few thoughts:
1. The machinery directive is unlikely to be applicable unless the cart is motorised.
2. The cart is likely to be an accessory to a medical device, but determining that requires close examination of the intended uses of both the cart and the carried monitor, as expressed by their respective manufacturers of record (i.e. those under whose names the products are placed on the market).
3. It's a little difficult to comment on the situation because it's a little unclear what the supply/commercial chain of events is. Could you please chart the logistic and ownership steps that the cart goes through from being completed at the factory and up to final use?
4. CE marking involves making sure that all the applicable CE marking directives are satisfied.
Watchcat
Agreed. Based on the information provided, what first came to my mind is that, when one company wants to hold the DHF for another company's product, it might be more interested in controlling the DHF than in just holding it. This made me wonder whether changes in the cart might reasonably be expected to affect the performance of the patient monitor, and/or on the suitability of the cart for use with the monitor.
My other thought is that, in simply supporting the cart and enabling the caregiver to move the monitor horizontally, the cart could have essentially the same relationship to the patient monitor as an elevator, which simply supports the cart and enables the caregiver to move the monitor vertically. If the relationship is not the same, perhaps the answer lies in the differences.
A few thoughts:
Thank you for your response :)
Its interesting to hear that you think that the Machinery Directive would only apply if the cart was motorized. The cart is NOT motorized.
We see it as an accessory.
I cannot provide that level of detail. The cart company manufactures the cart to all sorts of other device manufacturers so it is a commercially used product, not unique to our device.
dgrainger
It doesn't appear to be machinery as it isn't driven, nor does it appear to be an accessory to the device as it doesn't seem to enable' the functionality of the device. It's just a stand.
Ronen E
Problem Solver
Moderator
So it seems at face value, but we don't know that. Not enough information was provided. For example, carts sometimes include power supplies. Enable / doesn't enable depends on the definitions made by the manufacturers of record, and those aren't available to us either.
Additionally, the MDR definition contains to options ("or") for a product to be an Accessory to a Medical Device:
specifically enable the medical device(s) to be used in accordance with its/their intended purpose(s)
or
specifically and directly assist the medical functionality of the medical device(s) in terms of its/their intended purpose(s)
The first option doesn't include the word functionality (you wrote: "enable' the functionality of the device"). Is the monitor intended by its manufacturer to be used while sat on the floor? On the patient's bed? On the patient's bedside cabinet? I guess probably not, but the fact is we don't know yet. Moreover, it sounds like the monitor is intended to be relocateable with the patient - so how is it intended to be relocated in the absence of a cart (or similar)? Is it light enough to be carried by the caregiver? We don't know.
Ronen E
Problem Solver
Moderator
If the cart is indeed deemed an Accessory to a Medical Device, all the MDR becomes relevant to it just like to a medical device. The important (but subtle) point is that an object can officially become a medical device (or an accessory to one) without physically changing anything in the object, simply by assigning a certain intended use to it. The entity who owns it (i.e. under whose name it's sold in the EU) when that assignment occurs becomes fully responsible for MDR compliance. In your case it sounds like it can be either the actual manufacturer or your company. It all depends on the intended uses (of both the monitor's and the cart's) as you and them indicate them. It's important to note that the intended use may be changed through implication by some actions (e.g. selling the cart for a specific use), even without changing formal declarations, labelling etc.
How is the actual manufacturer currently handling the issue with their other customers? Are they in compliance with the MDR (or planning that)? If they're fine with taking that responsibility in your case, you can build your case such that you and them would be aligned with the regulation, but for them it might be an added burden. Sounds a little like a zero sum game - either they take the burden or you...
Watchcat
It seems like this sentence should read either:
"The cart company sells the cart to all sorts of other device manufacturers...
or
"The cart company manufactures the cart to all sort of other device manufacturers' specifications..."
If you meant to say the former, then, I agree, it would not be unique to your device. However, this doesn't seem consistent with your original statement about "essentially owning the cart."
It you meant to say the latter, then I would think any cart manufactured to your unique specifications would be unique to your device. In this context, "essentially owning the device" makes more sense to me. In fact, it would then seem that your company might have always "owned" the cart from a legal perspective, and the other company is just your contract manufacturer.
Watchcat
It's just a stand.
Yes, sometimes a cigar is just a cigar, and sometimes "our device is exactly the same as the predicate." But most of the time, it is "Well, except for X." "Oh, and also Y." "Did we forget to mention Z? Oops."
Color me skeptical, especially if something normally so mundane and non-proprietary as the basic supply/commercial chain of events is some kind of corporate trade secret. And why would anyone want to "own the DHF" on "just a stand"? Do they want to own the DHF on "just an elevator," too? Seems unlikely.
J Determining Fuse Rating and Labeling - Medical Device Crash Cart IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Medical device certificate in Australia - ARTG certificate Other Medical Device Regulations World-Wide 0
Q Software as a medical device vs software not sold as medical device: local regulations for sale? EU Medical Device Regulations 4
H Medical device Product Registration Registrars and Notified Bodies 2
A Can a power Supply be an accessory to a medical device, if it is an 'off-the-shelf' product. IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Medical device labelling Date of manufacture US Food and Drug Administration (FDA) 2
W Non Sterile Medical Device Environmental Tests Other Medical Device Related Standards 4
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
K 25-year lifetime of medical device - document storage period EU Medical Device Regulations 1
K Relabeling an existing medical device in the field? Other US Medical Device Regulations 6
A Reliable sources for following EU medical device regulatory EU Medical Device Regulations 0
T IVDR Medical device software CE Marking (Conformité Européene) / CB Scheme 8
J Requirements as a Distributor for Incoming Inspection of Purchased Finished Medical Device Medical Device Related Regulations 0
S Microwave medical waste disinfectant - A medical device or not? Other ISO and International Standards and European Regulations 3
S Registration of Medical Device in Hong Kong - labeling requirements Other Medical Device Regulations World-Wide 0
V Software as medical device (SaMD) replicated for multiple clients through APIs IEC 62304 - Medical Device Software Life Cycle Processes 5
M Is the output of a device a Medical Device? IEC 62304 - Medical Device Software Life Cycle Processes 5
P Do we need to retrospectively use the "MD" symbol (indicating device is a medical device) on labels, e.g. finished devices within expiration date? EU Medical Device Regulations 2
L Medical device registration in Iran Other Medical Device Regulations World-Wide 0
H EU CE marking for Medical Device Class I EU Medical Device Regulations 2
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 4
R Compatibility studies - Medicinal Product and Medical Device Other ISO and International Standards and European Regulations 0
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3
J Medical Device Regulations in Lebanon? Other Medical Device Regulations World-Wide 2
S Medical Device MRI Compatibility EU Medical Device Regulations 3
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
R Components to a finished medical device, MDR requirements Other US Medical Device Regulations 1
J Warnings/Cautions in Medical Device IFU Medical Device and FDA Regulations and Standards News 4
L Medical device HIPAA compliance in encryption Medical Information Technology, Medical Software and Health Informatics 1
E Medical Device - CE marking - Local market notifications EU Medical Device Regulations 2
S Medical Device Registration in Qatar Other Medical Device Regulations World-Wide 1
M Medical device substance based-leachables Other Medical Device Related Standards 2
P Anyone have an Idea on UAE Medical device registeration- Class B with FDA only Other Medical Device Regulations World-Wide 0
F Mobile app regulations - Class II medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
M What are the basics of Medical Device Single Audit Program (MDSAP)? ISO 13485:2016 - Medical Device Quality Management Systems 7
U Medical Device CE Marking - Using a disposable bearing CE Marking (Conformité Européene) / CB Scheme 3
L Medical Device Registration in Macau Other US Medical Device Regulations 1
A Medical Device Registration in the Dominican Republic Other Medical Device Regulations World-Wide 4
Aymaneh Medical Device Cybersecurity Risk Management IEC 27001 - Information Security Management Systems (ISMS) 2
T B2C Medical Device Shipping across the US Other Medical Device Related Standards 0
M Medical Device Registration In Malaysia Other Medical Device Regulations World-Wide 2
N Adding unclassified product to the medical device registration US Food and Drug Administration (FDA) 1
V Sister companies selling same medical device under different names ISO 13485:2016 - Medical Device Quality Management Systems 3
K CE Marking for Class I Medical Device? CE Marking (Conformité Européene) / CB Scheme 8
L Medical device storage conditions ISO 13485:2016 - Medical Device Quality Management Systems 1