Medical Device - Change manufacturing plant

#1
Dear all,
I would like to ask you the following issue we are dealing with. Thank you very much in advance for all your support. Appreciate it!

The situation is the following: There is a company developing a new medical device which is being manufactured by a subcontractor in a certified facility. Several tests have been carried out with these devices in order to eventually submit the documentation to a Notified Body. But, before that happens, the subcontractor manufacturing the devices changes its manufacturing plant to a different location, although the manufacturing procedures are more or less the same. Shall the company re-do all the tests that were carried out with the devices from the old plant? What would the company need to demonstrate in order for those studies be considered acceptable by a Notified Body?

pd: Notified Body: the competent authority in charge of evaluating MD in Europe

J.
 
Elsmar Forum Sponsor

yodon

Staff member
Super Moderator
#2
To be sure I understand: you did testing on devices to demonstrate they met requirements (you're not talking about production process validation, right)? Those tests demonstrate product requirements, nothing to do with the facility where they were made.

No, you shouldn't have to repeat those tests. You will need to ensure the new production site can make product which conforms to spec - so it's not just a matter of flipping the switch at the new production site.
 
#3
Thank you very much yodon!

No, we are not referring to the production process validation. We ascertain that if there are some procedures that will slighly vary, we would need to re-validate or validate again such procedures. We are referring to safety and performance tests with animals (i.e. biocompatibility) and even humans (clinical study).
In other words, at which level do we need to demonstrate that the products are the same when manufactured at the old facility and the new one? Is it enough if both comply with the specifications?

thank you very much in advance!
 

yodon

Staff member
Super Moderator
#4
You always (just) need to demonstrate the product meets spec. (Of course, you need sufficient specifications to ensure efficacy.) Your in-process and final acceptance testing should provide sufficient evidence that the devices perform as expected. (And, of course, your in-process and final acceptance tests need to be sufficient to ensure compliance to spec.)
 
#5
You always (just) need to demonstrate the product meets spec. (Of course, you need sufficient specifications to ensure efficacy.) Your in-process and final acceptance testing should provide sufficient evidence that the devices perform as expected. (And, of course, your in-process and final acceptance tests need to be sufficient to ensure compliance to spec.)
thanks again yodon, really helpful.

therefore, if we demonstrate that in-process controls and final acceptance criteria reach the same results, we would have demonstrated that there are no differences in the devices? Could you provide an example of a difference that will jeopardise that both devices are the same?

many thanks, J
 

yodon

Staff member
Super Moderator
#6
That was why I put the 'satisfactory' caveats in. If you don't have sufficient specifications, maybe you allow the new factory to use materials that aren't biocompatibile. Maybe you don't have sufficient performance specs and you accept the new devices but they really don't operate the same. There are probably endless possibilities. You need to look at your specs and confirm they are sufficient enough to always ensure that the products built perform the same. Identify those critical performance specs, do a process FMEA to identify places where you could deviate, and set up proper manufacturing controls to ensure you're not making snowflakes.
 
Thread starter Similar threads Forum Replies Date
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
F Change to Formulation due to Reach (China Medical Device Regulations) China Medical Device Regulations 0
dgrainger Informational EU medical device website change from 'Growth' to 'Health and Food Safety' (6/2020) Medical Device and FDA Regulations and Standards News 0
P Change on Medical Device Classifications - Dental Crowns Other Medical Device Regulations World-Wide 2
D Are medical device companies required to document every change made to their website? Document Control Systems, Procedures, Forms and Templates 2
L Managing design projects: New Medical Device Project or Change Design and Development of Products and Processes 1
R How long to wait for Notified Body Response - Potentially significant change to a medical device CE Marking (Conformité Européene) / CB Scheme 6
S Medical Device - Change / Variations to a Medical Device EU Medical Device Regulations 4
R SaMD - Software as a Medical Device - Software change control form ISO 13485:2016 - Medical Device Quality Management Systems 3
T Change to Country of Origin on Medical Device Labeling Other Medical Device Regulations World-Wide 0
J South Korea, MFDS - Medical Device Change Management Other Medical Device Regulations World-Wide 5
J Class 1 Medical Device ECR (Engineering Change Request) Changes 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
TheMightyWife Address Change - Medical Device Packaging Labelling Requirements EU Medical Device Regulations 7
C Vietnam - Authorized Medical Device Representative Change Other Medical Device Regulations World-Wide 10
M New Medical Device Accessory with New Function - Is this a major change? Other Medical Device Regulations World-Wide 3
H Change of a medical device sterilization dose - New 510(k) necessary? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
C Medical Device Website Change Control ISO 13485:2016 - Medical Device Quality Management Systems 3
M Medical Device Change Requirements - Malaysia Registered Products Other Medical Device Regulations World-Wide 4
MDD_QNA Medical Device PCB component change IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
JoCam Validation of a change to a CE Marked Medical Device EU Medical Device Regulations 4
S Change of Medical Device Company Registered Address - Implications Quality Manager and Management Related Issues 3
M Making a change to an approved Medical Device Other US Medical Device Regulations 9
H FDA/PMA/Medical Device/Before submission/Manufacturing facility change US Food and Drug Administration (FDA) 1
B Class II and Class III Medical Device Amendment for Change in Location? Canada Medical Device Regulations 2
T Class II Medical Device with Software - Change to Computer 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
G Validity for Thailand Medical Device License and can change Distributor thereafter? Other Medical Device Regulations World-Wide 3
D Change in Medical Device Shipping Box - Transport Test Required? Other Medical Device Related Standards 1
G How do I change the Medical Device Product Code in FDA site? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
A Medical Device DHF (Design History File) - One per Design Change? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 17
J Class IIa Medical Device - Can I over label to change an address? ISO 13485:2016 - Medical Device Quality Management Systems 3
A Change Medical Device Class from Class II to Class I 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
M Class I Medical Device - 1st notification and change notification EU Medical Device Regulations 4
D Regulations for Relabeling of Medical Device for Address Change ISO 13485:2016 - Medical Device Quality Management Systems 1
R EU Medical Device Change Decision Tree (when to notify the NB) EU Medical Device Regulations 4
S Medical Device Subcontractor Significant Change EU Medical Device Regulations 9
E Address change on Medical Device labeling 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
K Change Circuit Board to CE marked product on CE Marked Medical Device EU Medical Device Regulations 2
C When to submit a 510(k) for a change to an exisiting medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 29
M Change from Automotive to Medical Device Industry Career and Occupation Discussions 3
A Can a power Supply be an accessory to a medical device, if it is an 'off-the-shelf' product. IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Medical device labelling Date of manufacture US Food and Drug Administration (FDA) 2
W Non Sterile Medical Device Environmental Tests Other Medical Device Related Standards 4
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
K 25-year lifetime of medical device - document storage period EU Medical Device Regulations 1
K Relabeling an existing medical device in the field? Other US Medical Device Regulations 6
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
A Reliable sources for following EU medical device regulatory EU Medical Device Regulations 0
T IVDR Medical device software CE Marking (Conformité Européene) / CB Scheme 8
J Requirements as a Distributor for Incoming Inspection of Purchased Finished Medical Device Medical Device Related Regulations 0
S Microwave medical waste disinfectant - A medical device or not? Other ISO and International Standards and European Regulations 3

Similar threads

Top Bottom