Medical Device changes - Active implants
P
polish1023
#1
Hello,
We are developing an active implant (pre-CE). As we are planning to conduct preclinical studies, the implants we have available at this point have been created by a different manufacturing mold than the one we plan to use to make the implants that are dedicated to humans.
We believe that the change in mold will not affect the safety and performance of the implant and therefore we would like to continue with our preclinical plan and implant the current implant configuration (it will most probably take at least 3 months to manufacture the new mold).
If we plan to submit the preclinical data along with a rationale and bench tests on why these implants are equivalent in terms of safety and efficacy, do you anticipate any problems with the EC or Competent Authorities?
:thanx:
Elsmar Forum Sponsor
planB
Super Moderator
#2
Re: Device changes - Active implants
polish1023,
I am assuming when you talk about preclinical studies you mean activities to support your biocompatibility evaluation? If yes, you may consider demonstrating equivalence by comparing your material characterization data (gained per ISO 10993-18) for the implant when manufactured with the current and future mold, plus a comparative cytotoxicity study (per ISO10993-5): if you end up with an equivalent leachables/extractables profile and an equivalent cytotoxicity potential, you may have a valid rationale for not repeating the tests for all other biological risks.
So in case your rationale is valid, you should be fine for your submission:
Hello,
[...] If we plan to submit the preclinical data along with a rationale and bench tests on why these implants are equivalent in terms of safety and efficacy, do you anticipate any problems with the EC or Competent Authorities?
HTH, Gerhard
#3
In your post you state that "We believe that the change in mould will not affect the safety and performance..."
It may be that the level of concern that a regulatory body will have with the change will depend upon the engineering / scientific rationale to back up your statement. This could be the result of tests performed or strong scientific judgement.
Thread starter Similar threads Forum Replies Date
M Informational New Zealand – Changes to Medical Device Adverse Event Reporting process Medical Device and FDA Regulations and Standards News 0
M Informational TGA Consultation: Proposed changes to medical device essential principles for safety and performance Medical Device and FDA Regulations and Standards News 0
C What if the foreign manufacturer of a medical device changes the Authorized Representative in Saudi Arabia? Other Medical Device Regulations World-Wide 0
E Changes to medical device submitted through a variation? EU Medical Device Regulations 1
M Medical Device News TGA Consultation: Changes to a number of definitions and the scope of the medical device regulatory framework in Australia Medical Device and FDA Regulations and Standards News 0
J Class 1 Medical Device ECR (Engineering Change Request) Changes 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
S Medical Device Changes and How to Assess Their Impacts to Essential Requirements EU Medical Device Regulations 2
shimonv Reporting Medical Device Changes with Annex V/VII EU Medical Device Regulations 1
N Notification of Changes to a Class I Medical Device - Reporting Requirements EU Medical Device Regulations 4
E Status of proposed changes to Medical Device Directive EU Medical Device Regulations 0
Q Looking for documents relating to Medical Device Design Changes Document Control Systems, Procedures, Forms and Templates 5
L Notification and submission requirements for medical device material changes Registrars and Notified Bodies 2
Q Reporting Significant Medical Device Changes to Argentina's ANMAT Other Medical Device Regulations World-Wide 3
somashekar Medical Device Establishment Registration and Listing - Notice of Changes for FY 2013 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
thisby_ Changes in an already certified IEC 60601 Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T How to Report/Notify Medical Device Software Changes to the FDA? US Food and Drug Administration (FDA) 3
P Proposed Changes to the Medical Device Directive (and IVDD) EU Medical Device Regulations 0
E Changes to IFU (Instructions for Use) on a CE Marked Medical Device ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
T Reporting changes to a Class II Medical Device in Canada Canada Medical Device Regulations 3
AnaMariaVR2 Medical Device Establishment Registration and Listing - Notice of Changes for FY 2013 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
T Multiple Changes to a Class III Medical Device - Single License Amendment? Canada Medical Device Regulations 2
R Medical Device Circuit Board Component Changes - Is this rework or not? ISO 13485:2016 - Medical Device Quality Management Systems 8
S What is the best approach for a Class II Medical Device when its branding changes? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
L Changes to existing Medical Device marketed under different name 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
E MDD 2007/47/EEC - Medical Device Directive Labeling Changes EU Medical Device Regulations 44
A ISO 13485 Risk Analysis - Class II Medical Device Engineering Changes ISO 13485:2016 - Medical Device Quality Management Systems 9
M Medical device certificate in Australia - ARTG certificate Other Medical Device Regulations World-Wide 0
Q Software as a medical device vs software not sold as medical device: local regulations for sale? EU Medical Device Regulations 4
H Medical device Product Registration Registrars and Notified Bodies 2
A Can a power Supply be an accessory to a medical device, if it is an 'off-the-shelf' product. IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Medical device labelling Date of manufacture US Food and Drug Administration (FDA) 2
W Non Sterile Medical Device Environmental Tests Other Medical Device Related Standards 4
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
K 25-year lifetime of medical device - document storage period EU Medical Device Regulations 1
K Relabeling an existing medical device in the field? Other US Medical Device Regulations 6
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
A Reliable sources for following EU medical device regulatory EU Medical Device Regulations 0
T IVDR Medical device software CE Marking (Conformité Européene) / CB Scheme 8
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
J Requirements as a Distributor for Incoming Inspection of Purchased Finished Medical Device Medical Device Related Regulations 0
S Microwave medical waste disinfectant - A medical device or not? Other ISO and International Standards and European Regulations 3
S Registration of Medical Device in Hong Kong - labeling requirements Other Medical Device Regulations World-Wide 0
V Software as medical device (SaMD) replicated for multiple clients through APIs IEC 62304 - Medical Device Software Life Cycle Processes 5
M Is the output of a device a Medical Device? IEC 62304 - Medical Device Software Life Cycle Processes 5
P Do we need to retrospectively use the "MD" symbol (indicating device is a medical device) on labels, e.g. finished devices within expiration date? EU Medical Device Regulations 2
L Medical device registration in Iran Other Medical Device Regulations World-Wide 0
H EU CE marking for Medical Device Class I EU Medical Device Regulations 2
A Medical Device Contract Manufacturer - Does the CM need to register with FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 4
R Compatibility studies - Medicinal Product and Medical Device Other ISO and International Standards and European Regulations 0
Similar threads
Top Bottom