Medical Device Class I and Class IIa Differences? Inhalator
G
googgic
#1
What is the different between medical class I and IIa. My device is an inhalator which is classed in the directive as class I, is that any chance it will be class IIa???
Elsmar Forum Sponsor
Al Rosen
Staff member
Super Moderator
#2
You really need to look at annex IX in the MDD to determine your classification. If your inhaler is an active device used to administer medicationit is a class IIa device according to Rule 11 in Annex IX. If you are uncertain, discuss it with a Notified Body.
G
googgic
#3
Thank you Al Rosen,
The device is not for medical purpose but it is used as an alternative to alleviate some sickness such as asthma, bronchitis, breathing difficulty, etc. I read through but still confuse though. I may be stupid, coule you tell me what is the difference between invasive and active device.
Al Rosen
Staff member
Super Moderator
#4
googgic said:
Thank you Al Rosen,
In your first post you stated it was class I. Now you say it is not a medical device. I don't quite understand. If it is not a medical device, you don't have to concern yourself with the MDD and classification. The definitions of a medical device, invasive and active are within the MDD.
'medical device' means any instrument, apparatus, appliance, material or other article, whether used alone or in combination, including the software necessary for its proper application intended by the manufacturer to be used for human beings for the purpose of:
- diagnosis, prevention, monitoring, treatment Or alleviation of disease,
- diagnosis, monitoring, treatment, alleviation of or compensation for an injury or handicap,
- investigation, replacement or modification of the anatomy or of a physiological process,
- control of conception,
and which does not achieve its principal intended action in or on the human body by pharmacological, immunological ormetabolic means, but which may be assisted in its function by such means;
Invasive device
A device which, in whole or in part, penetrates inside the body, either through a body orifice or through the
surface of the body.
Active medical device
Any medical device operation of which depends on a source of electrical energy or any source of power other than that directly generated by the human body or gravity and which acts by converting this energy.
Medical devices intended to transmit energy, substances or other elements between an active medical device and the patient, without any significant change, are not considered to be active medical devices.
You can download a copy of the MDD from here.
G
googgic
#5
:thanx: Thank you Al,
I read through all documents what are hugely gathered on my desk and download and print everything which is possible to make me more clear. I try to understand with my limited English knowledge. I wish I could find this forum last 3-4 month ago when I was wondering. I am not sure my product is medical device or not. It can be and it cannot be. It should be great if you can look through here and please help to let me know whether it is a medical device or not.
I have tried to find this answer for alomost half year and I am really a non expert and novice and I want to do thing right before someone will start charge me for criminal offense. If I do not want it to be medical device which information I should not allow to put on my product.
Thank you for your time so much. I hope your good Karma will be increase. I am a buddhist and I wish you all the best. Good thing will turn around to you as you like to help people, you will never be in trouble.
GOoggic
Tim Folkerts
Super Moderator
#6
Googgic,
I'm no expert on medical devices so I can't comment much in that area, but one bit of business advice - hire an editor for the website who is knowledgeable with medicine and knowledgeable with English. When dealing with alternative medicine, credibility is essential. I tend to be skeptical of alternative medicine, so any excuse - like poor grammar or poor wording - can be enough for me to dismiss the product without really looking into the actual merits. It may not be fair to equate the technical expertise of a person with his ability to write in English, but that is the reality. There were numerous cases throughout the website where the intent was clear, but the actual wording was less than ideal. Part of that may be American English vs. British English, but some of it is definitely incorrect English no matter where you come from.
A few dollars (or pounds or Euros) invested in cleaning up the writing could go a long way toward boosting consumer confidence and sales.
Tim F
G
googgic
#7
thank you Tim for your comment. I am going to see some English people to edit them for me. Sometimes, it is so difficult to explain in English what I would like to let people know about it. But i will try to improve on it.
Al Rosen
Staff member
Super Moderator
#8
GOoggic
I think you should find a lawyer experienced in medical devices to review your website and the claims you make for your inhaler.
R
Roland Cooke
#9
Re: Difference between Medical Device Class I and Class IIa? Inhalator
Thank you Al Rosen,
Hi Googic,
As you are based in the UK, I imagine that you are discussing the requirements for selling into that country. (There are posters from all over the world here, classifications are different in different places!)
Firstly, it sounds as if you do indeed need to come back to us with a clear description of the product and its intended use.
The use of clear English is critical, so I would definitely advise getting some assistance with that - however you may be lucky, we have people from all corners on the globe on this forum, you may find there is a native speaker here that can assist with a translation.
As for whether the device is a medical device or not, it sounds from the limited description as if it most definitely is. It may not cure asthma/bronchitis, but if it improves the patient's clinical condition, it would come under the "alleviation of disease" description.
Again, as already pointed out, the clinical evidence would need to be robust (true for any medical device).
An active device is one that is powered by electricity (etc), rather than by gravity or musclepower.
The classification of your device will depend on whether it is invasive and how long it will be used. See Annex IX as discussed (MDD attached).
Ajit Basrur
Staff member
Admin
#10
Just an add-on to the great advise given by my friend Roland - a new revision of MDD 93/42/EEC is available and is located HERE
Thread starter Similar threads Forum Replies Date
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
H EU CE marking for Medical Device Class I EU Medical Device Regulations 2
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 4
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
P Anyone have an Idea on UAE Medical device registeration- Class B with FDA only Other Medical Device Regulations World-Wide 0
F Mobile app regulations - Class II medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
K CE Marking for Class I Medical Device? CE Marking (Conformité Européene) / CB Scheme 8
D Class IIB Medical Device Accessories - OTS Tablets Other Medical Device Regulations World-Wide 1
A Interpretation of GMP Requirements for class 1 medical device manufacturer (device GMP exempt, only General controls applicable) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
H MDD Article 12 Labeling for Class IIa Medical Device - Please Advise EU Medical Device Regulations 3
A Medical Device Vigilance decision tree for Japan for class 2 devices. Japan Medical Device Regulations 1
Edward Reesor EU Authorized Representative for a Class I Medical Device CE Marking (Conformité Européene) / CB Scheme 11
W Direct to customer export of medical device (class I: prescription lenses + frame) US Food and Drug Administration (FDA) 2
C New Class III medical device application for Health Canada Canada Medical Device Regulations 6
T First 510(k) submission - Class II software as medical device US Food and Drug Administration (FDA) 4
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S How to register class IIA medical device accessories EU Medical Device Regulations 1
U Medical device trial budget - 2 hospitals for Class IIa medical device in Sweden Other Medical Device and Orthopedic Related Topics 2
A Class medical device (MDD) - Classification help EU Medical Device Regulations 1
M Notified Body audits -Class Ia medical device product EU Medical Device Regulations 4
N Can a (class I) medical device be manufactured in a GMP certified site with no ISO13485? EU Medical Device Regulations 18
U Do we need clinical trial data for Class IIa medical device under MDR EU Medical Device Regulations 7
R Countries that accept 510k approvals - Class II nonsterile medical device Other Medical Device Regulations World-Wide 2
B Class III Medical Device Accelerated conditions Other Medical Device Related Standards 1
D FDA Approved Class II Medical Device? Other US Medical Device Regulations 11
N CE Mark for Class 1 medical device CE Marking (Conformité Européene) / CB Scheme 11
D Class II medical device - When should a complaint be closed? Customer Complaints 6
D Software as an accessory to a Class I medical device EU Medical Device Regulations 4
C ETO Sterilised Class II Medical Device - Required Temperature Storage ISO 13485:2016 - Medical Device Quality Management Systems 1
A Does Class 1 Medical Device need to be certified to MDSAP? Canada Medical Device Regulations 5
N Can we take a part from 510k cleared medical device and use it in class I device? Other US Medical Device Regulations 3
J Class 1 Medical Device - Using a UPC over the UDI? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
N Selling a class II Medical Device in Canada in Pharmacies Elsmar Xenforo Forum Software Instructions and Help 10
S Acrylic in Class II b Active Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
L A Taiwan company want to sell Class I medical device (510(k) exempt) on Amazon, should we register with FDA? US Food and Drug Administration (FDA) 4
S Conformity Assessment Route question for Class IIa medical device under MDR EU Medical Device Regulations 3
K Inventory Control - Class II (IIb for CE) medical device ISO 13485:2016 - Medical Device Quality Management Systems 6
S Software Release Note - Class A stand alone software medical device IEC 62304 - Medical Device Software Life Cycle Processes 2
J FDA Class 1 Medical Device UDI requirements Other US Medical Device Regulations 19
L Class IIa Medical Device, CE Mark & 510k - Labeling Content CE Marking (Conformité Européene) / CB Scheme 4
M How is Class I Medical Device Compliance Enforced? EU Medical Device Regulations 7
K Calibration of Dose Area Product (DAP) Meter in our Class IIb X-Ray Medical Device General Measurement Device and Calibration Topics 1
Z Class 2 medical device - Process Validation - Test sterile or non-sterile units? Qualification and Validation (including 21 CFR Part 11) 4
C Medical device manufacturing (class 2 ISO 13485:2016) - Is a Deviation allowed? Other Medical Device Related Standards 5
S Understanding UDI requirements - Class 2 medical device (hearing aids) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
N Per the FDA regulations can a medical device be class I despite that it is sterile? Other Medical Device Related Standards 4
M Include IFU in Class II Medical Device Marketing Brochure? US Food and Drug Administration (FDA) 2
J Steps when changing material for class 1 medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
Similar threads
Top Bottom