Medical Device Design verification sample prototype
I have been reading a lot of topics about this subject but can't really understand how it is done..
What should I do (step by step) to calculate the sample size for design verification. We need to do it on few samples but in the audit they told us to come up with a valid statistic argument to justify the sample size.
We are testing an active guidewire (DV tests: Biocompatibility, software, computer safety..)
Thanks for reading this
Ring
John Broomfield
You verify the conformity of your prototype to your design before you then test to validate the design. That way you know exactly what you have tested to see if the prototype works as intended.
From your question it seems that you are seeking information on design validation not verification.
Thank you for your response.
No it is for design verification, validation will come after.
What I am looking for is how to justify (and chose) the sample size for the tests of DVerification
John Predmore
When you do verification, there are four possibilities because of variation and measurement uncertainty. A hypothesis is true and your experiment supports the claim, or else you incorrectly conclude it is false. Or a hypothesis is really false and you prove it false, or else you are unable to show it false. Scenarios 1 and 3 are the outcomes you want. Scenario 2 is called a Type 1 error and scenario 4 is a Type II error. The statistical test you design (such as t-test for difference of means) has a factor in the calculation related to the sample size. As the experimental designer, you decide what level risk you can accept (balanced against the practical concern about excessive cost), and you choose a sample size appropriate for the risk you agree to accept.
Ronen E
Problem Solver
Moderator
This is roughly the traditional textbook spill about the topic; unfortunately when you go into it in greater detail and acknowledge all the assumptions and approximations made along the way, you might realise that there's a lot more to it. After >20 years in developing medical devices I'm still in search of the holy grail (caveat: I didn't directly deal with statistics and experiments the whole time... the other side of it is regulatory and standards, where things are even fuzzier!).
Ronen E
Problem Solver
Moderator
Biocompatibilty testing sample sizes will be best advised by your accredited testing provider (3rd party lab / toxicologist), which in turn will rely on widely accepted test standards.
Software testing, AFAIK, doesn't require a "sample" because at least in theory all instances of the software are identical.
Not sure what you mean by "computer safety".
I do imagine that additional (bench) test types are involved in verifying a guidewire.
Thank you for your answers.
Yes from ISO 11070 mostly (see in link).
Sorry I'm just a rookie and I was told "we haven't a lot of samples for verification so we need to use a small amount and the auditer told us to have a "statiscal rational" to justify this small samples"
Not the easiest task :rolleyes:
Ronen E
Problem Solver
Moderator
No, it's not an easy task, and looking at the list it's also not a small one... Actually it's a full-on mini-project for a statistician or another knowledgeable professional.
Just note that "a sample" in this case is not necessarily identical with a sample of physical device units. Unless it's a disposable device, you are generally allowed to run each unit multiple times to enlarge your sample size.
A Sample Agreement available for Outsourcing Medical Device Design activity? ISO 13485:2016 - Medical Device Quality Management Systems 1
U Medical Device Design finalization testing ISO 13485:2016 - Medical Device Quality Management Systems 2
I Does anybody use Detection in medical device Design FMEA? ISO 14971 - Medical Device Risk Management 18
M Informational US FDA Draft Guidance – Patient Engagement in Design and Conduct of Medical Device Clinical Investigations Medical Device and FDA Regulations and Standards News 0
M Medical Device Design Control Auditor Recommentations General Auditing Discussions 19
L Managing design projects: New Medical Device Project or Change Design and Development of Products and Processes 1
P Software requirement and design specifications - Medical device contains both hardware and software IEC 62304 - Medical Device Software Life Cycle Processes 2
L Software Medical Device - 7.3.8 - Design and Development Transfer ISO 13485:2016 - Medical Device Quality Management Systems 4
I Design & Development, Research and Innovation in the Medical Device framework ISO 13485:2016 - Medical Device Quality Management Systems 3
3 Brasil: OEM Medical Device Business - Product Design Customization Other Medical Device Regulations World-Wide 6
G Design Philosophy - Iran Medical Device Registration Other Medical Device Regulations World-Wide 7
Q Medical Device Design & Development Plan Template wanted Document Control Systems, Procedures, Forms and Templates 1
Q Looking for documents relating to Medical Device Design Changes Document Control Systems, Procedures, Forms and Templates 5
A Sample document on design and development activities plan for a medical device Other Medical Device Regulations World-Wide 1
L Medical Device New Product Design & Development- How long does this process take you? Other Medical Device and Orthopedic Related Topics 1
L Medical Device Manufacturing Equipment Design Other Medical Device Related Standards 2
M Design Inputs - Final Medical Device Specifications 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
M Class II Medical Device Design Control Non Conformance 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
A Medical Device Service Records linked to DHR (Design History Record)? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
A DHF (Design History File) for a Legacy Product (Class iii Medical Device) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
G Medical Device Company - Design Only Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
K Importance of Medical Device Accessory Design Controls Other US Medical Device Regulations 4
H Medical Device Design Control Requirements (820.30) on Existing Products 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
R Medical Device Software Segregation and Detailed Design IEC 62304 - Medical Device Software Life Cycle Processes 10
A Medical Device Electronic Systems Design - Points to be cleared EU Medical Device Regulations 2
A Design Company - CE Mark for a Medical Device Class 2A , Annex VI & Annex VII EU Medical Device Regulations 12
D Class I and Class II Medical Device Accessory Design Control Requirements US Food and Drug Administration (FDA) 3
T 21 CFR Part 820.30 Design Controls for Class 1 Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
K Medical Device Design Validation vs Clinical Evaluation ISO 13485:2016 - Medical Device Quality Management Systems 5
L When does Medical Device New Product R&D stop and Design start? Design and Development of Products and Processes 6
A Medical Device DHF (Design History File) - One per Design Change? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 17
L Creating DHF (Design History File) for Medical Device Systems Design and Development of Products and Processes 8
A How to do Medical Device Design Validation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
B ISO 13485 Class I Medical Device Design Exclusion ISO 13485:2016 - Medical Device Quality Management Systems 10
U Medical Device Design Review Records Requirements Design and Development of Products and Processes 16
A Taguchi Inner and Outer Array Design for Medical Device Testing Quality Tools, Improvement and Analysis 1
J How do I qualify a Label Design for Medical Device Design and Development of Products and Processes 4
D Medical Device Design Verification and Validation Differences 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 16
J Class II Medical Device Design Revision Controls ISO 13485:2016 - Medical Device Quality Management Systems 7
C At what stage does Design Control play a role in Medical Device Development ISO 13485:2016 - Medical Device Quality Management Systems 2
J What things go in Design Control Documentation for a 510k of Class II Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
J Design Control & Rapid Prototyping - Medical Device Software 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
N Medical Device Design and Development Plan Procedures help wanted 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
Ajit Basrur Design Controls for medical device exported from China China Medical Device Regulations 5
S Medical Device Design Verification and Validation - info Other Medical Device and Orthopedic Related Topics 1
T Medical Device Design Quality Plan Design and Development of Products and Processes 6
D Design Verification for special 510(k) - Size modification to an existing medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
S Medical Device Technical Files (Class IIa & IIb) ? Design Development Documentation EU Medical Device Regulations 7
N Design and Development Plan for a Medical Device - Template needed 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
A Medical Device Design Verification Testing - Is ISO 62366:2008 what I'm looking for? IEC 62366 - Medical Device Usability Engineering 16