Hi,

I work for a medical device manufacturer.
I have led a few FMEAs to date and have run into a frequent issue of determining where to limit potential effects.
For example:
MFG Step Requirement: Deposit an Enzyme layer of 1 (+/-0.1um) on electrode.
Failure Modes: 1) Deposit too thick, 2) Deposit too thin, 3) Step missed.
Effects for any one of those Failure modes could be: Local: Fail Part, Local: Rework. Next Step: Poor adhesion, End: Product does not meet customer needs - Dissatisfied Customer, Product does not meet customer needs - misdiagnosis.

The question that typically comes up is: Should we consider existing prevention and detection control measures when determining potential effects? My belief is that you do not take control measures into account when identifying potential effects. I could be wrong here though and would like some outside opinions before I go too far off course.

Similarly, when determining severity, do you discount end user impacts if you have a great downstream detection or upstream prevention?

Any help here would be greatly appreciated,
 
Elsmar Forum Sponsor
A

amehta44

#2
Hi Julison,

FMEA is a techniques used to determine high-risk product features based on the impact of a failure and the likelihood that a failure could occur without detection.

FMEA in a nutshell requires the following:
1. You analyze the product - including all functions, current performance levels, and definitions of failures of each function.
2. For each function, you'd then identify failures modes and associated effects.
3. You define Severity (S) for each failure mode: assigned as number 1-10, 1= not significant, 10=most severe.
4. You define the probability of occurrence (P): assigned as number 1-10, 1=not likely to occur, 10=highly likely to occur
5. You define the likelihood (D) of detection: assigned as number 1-10, 1=easily detected, 10=hard to detect
6. You calculate Risk Pririty Number (RPN): Severity (S) * Probability (P) * Detectability (D) [Result: 1-1000]
7. You'd then prioritize the failure modes based on RPN. Some organizations use threshold values, above which preventive action must be taken. For example, the organization may require improvement for any RPN exceeding 120.

With that said, lets look at your questions:

"I have led a few FMEAs to date and have run into a frequent issue of determining where to limit potential effects.": You'd include all potential high-risk effects, and let the RPN help you prioritize improvements.

"Should we consider existing prevention and detection control measures when determining potential effects?": You must absolutely account for existing prevention and detection measures without which you cannot have meaningful assignment for the probability of occurrence (P), and/or the likelihood of detection (D).

"Similarly, when determining severity, do you discount end user impacts if you have a great downstream detection or upstream prevention?": The RPN calculation automatically does this for you.

Hope this helps.

Regards.
 
R

randomname

#3
Severity is the impact if it were to happen, independent of controls.

The boundaries of your analysis (internal, external) and a common concern. For example, harm to the patient is a no brainer. But how about impact on a medical provider, or calibration tech, etc.
 
R

randomname

#4
By the way, the impacts and boundaries should have been already discussed during the device risk assessment (hazards and harms) required by the risk management plan (ISO 14971).
 
J

Julison

#5
Hi,

Thank you for the replies.
However, based upon those replies I am still uncertain what the best answer is to the questions I have posed.

Question #1 was:
"Should we consider existing prevention and detection control measures when determining potential effects?":

Answer #1: You must absolutely account for existing prevention and detection measures without which you cannot have meaningful assignment for the probability of occurrence (P), and/or the likelihood of detection (D).

I think I may not have been clear enough in the question.
We do record failure modes, effects, causes and control measures.

The question I have is:

Do we record a potential end effect if downstream we have a control that will detect it with a high degree of certainty.

For instance. The process requirement could be to deposit a glucose sensor that is 1um thick. The failure mode could be that it is >1um thick or <1um thick or absent. Some of the effects could be local: Product fails at that step's inline QC, OR, End User: The customer (hospital) gets an inaccurate result.

IF the inline QC detection is VERY effective, should we even record the potential End User effect "The customer (hospital) gets an inaccurate result "?

Thanks again,

Julison
 
J

Julison

#6
Thanks Randomname.

Your answer was "Severity is the impact if it were to happen, independent of controls."

I think part of my problem is that our severity criteria are based upon how far reaching the impact of the failure mode reaches.

If something is caught at the step that it occurs it has lower severity than if it makes it to the next step or worse case the customer is impacted in a manner that creates harm.

So, our criteria force us to assess internal and external controls and as such we cannot assess the severity "independent of controls."

Not sure how to address this without challenging the company risk criteria.

Ideas???

Julison
 
R

randomname

#7
I would use the highest severity, since that is the actual risk. However, each failure mode could have multiple effects, with a different severity for each.
 

Eamon

Involved In Discussions
#9
Perhaps the following observation is obvious, or perhaps it bears suggesting.

It is often the case that the same failure mode has a greater likelihood of a lower severity effect, as well as a lesser likelihood of a higher severity effect. (This is without considering detectability, about which there may be similar dependencies and considerations.)

The way to document this would be to reproduce the same failure mode on more than one FMEA row, with each row possibly documenting a different effect, and with each row showing a different balance between the risk numbers. In such a case, the RPN on each row of the the "split" failure mode may be acceptable, but a single row combining the highest likelihood and the worst severity would be unreasonably high RPN.

I think this is a reasonable approach to take into account a common situation. I would be interested to know if this is commonly done, or there is an argument against taking this approach.

Eamon

Edit:

I have just realized that the likelihood primarily pertains to the occurrence of the failure mode, not to the occurrence of the effect, so my suggestion may not play well with the semantics of most FMEA methodologies. Nonetheless, there should be a way of capturing such analysis in an FMEA. How should it be done?
 
Last edited:
Thread starter Similar threads Forum Replies Date
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N Medical Device Failure Rate as part of the Risk Management Report ISO 13485:2016 - Medical Device Quality Management Systems 10
Marc FDA: Software Failure Behind 24% of Last Year's Medical Device Recalls Other Medical Device and Orthopedic Related Topics 2
A Medical device labelling Date of manufacture US Food and Drug Administration (FDA) 2
W Non Sterile Medical Device Environmental Tests Other Medical Device Related Standards 4
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
K 25-year lifetime of medical device - document storage period EU Medical Device Regulations 1
K Relabeling an existing medical device in the field? Other US Medical Device Regulations 6
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
A Reliable sources for following EU medical device regulatory EU Medical Device Regulations 0
T IVDR Medical device software CE Marking (Conformité Européene) / CB Scheme 8
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
J Requirements as a Distributor for Incoming Inspection of Purchased Finished Medical Device Medical Device Related Regulations 0
S Microwave medical waste disinfectant - A medical device or not? Other ISO and International Standards and European Regulations 3
S Registration of Medical Device in Hong Kong - labeling requirements Other Medical Device Regulations World-Wide 0
V Software as medical device (SaMD) replicated for multiple clients through APIs IEC 62304 - Medical Device Software Life Cycle Processes 4
M Is the output of a device a Medical Device? IEC 62304 - Medical Device Software Life Cycle Processes 5
P Do we need to retrospectively use the "MD" symbol (indicating device is a medical device) on labels, e.g. finished devices within expiration date? EU Medical Device Regulations 2
L Medical device registration in Iran Other Medical Device Regulations World-Wide 0
H EU CE marking for Medical Device Class I EU Medical Device Regulations 2
A Medical Device Contract Manufacturer - Does the CM need to register with FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 4
R Compatibility studies - Medicinal Product and Medical Device Other ISO and International Standards and European Regulations 0
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3
J Medical Device Regulations in Lebanon? Other Medical Device Regulations World-Wide 2
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
S Medical Device MRI Compatibility EU Medical Device Regulations 3
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
R Components to a finished medical device, MDR requirements Other US Medical Device Regulations 1
J Warnings/Cautions in Medical Device IFU Medical Device and FDA Regulations and Standards News 4
L Medical device HIPAA compliance in encryption Medical Information Technology, Medical Software and Health Informatics 1
M V&V phase: Justification of acceptance criteria (statistical method ) - (Medical Device) Design and Development of Products and Processes 2
E Medical Device - CE marking - Local market notifications EU Medical Device Regulations 2
S Medical Device Registration in Qatar Other Medical Device Regulations World-Wide 1
M Medical device substance based-leachables Other Medical Device Related Standards 2
P Anyone have an Idea on UAE Medical device registeration- Class B with FDA only Other Medical Device Regulations World-Wide 0
F Mobile app regulations - Class II medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
M What are the basics of Medical Device Single Audit Program (MDSAP)? ISO 13485:2016 - Medical Device Quality Management Systems 7
U Medical Device CE Marking - Using a disposable bearing CE Marking (Conformité Européene) / CB Scheme 3
L Medical Device Registration in Macau Other US Medical Device Regulations 1
A Medical Device Registration in the Dominican Republic Other Medical Device Regulations World-Wide 4
Aymaneh Medical Device Cybersecurity Risk Management IEC 27001 - Information Security Management Systems (ISMS) 2
T B2C Medical Device Shipping across the US Other Medical Device Related Standards 0
M Medical Device Registration In Malaysia Other Medical Device Regulations World-Wide 2
N Adding unclassified product to the medical device registration US Food and Drug Administration (FDA) 1
V Sister companies selling same medical device under different names ISO 13485:2016 - Medical Device Quality Management Systems 3
K CE Marking for Class I Medical Device? CE Marking (Conformité Européene) / CB Scheme 8
L Medical device storage conditions ISO 13485:2016 - Medical Device Quality Management Systems 1
F USB powered handheld medical device - Isolation requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
L How to determine / validate Medical Device Storage Conditions ISO 13485:2016 - Medical Device Quality Management Systems 1

Similar threads

Top Bottom