Hello.

Actually I have to deal with discussions about complaint handling...

If a medical device is sent back from customer to the manufacturer because of a defect AFTER the device's specified lifetime has expired - is it a requirement to treat this as a complaint acc. §820.198?

If yes - where can i find a corresponding regulatory requirement?
Any comment is appreciated!

Thanks in advance for your feedback and support!

BR
Aphel
 
Elsmar Forum Sponsor
M

MIREGMGR

#2
Re: Medical Device Handling AFTER AFTER the device's specified lifetime has expired

My understanding is yes...particularly if harm or elevated risk was involved.

Your specified device life affects your obligation to the end user in regard to what you do to make the end user whole, i.e. during device lifetime you might send them a new unit, after device lifetime you might tell them "thank you for the information, here is our price list and ordering information".

Your specified device lifetime also may be part of your documented device performance specification, i.e. this widget will do whatever widgets do with accuracy per ASTM 12345 during its five year specified lifetime from date of sale.

But my understanding is that you have an obligation to record and evaluate the complaint, and report it to FDA if it meets reportability requirements, even if you do not have to take further action.
 

Ronen E

Problem Solver
Staff member
Moderator
#3
In my understanding an evaluation has to take place, because the fact that the complaint was received after the device lifetime expiry doesn't automatically say that the failure occurred at that time (could have gone unnoticed or unreported - for whatever reason - for a while). Even if it failed after expiry, the defect or otherwise the root cause could have been there long before. You have to look into it in order to understand what happened and what are your obligations are as a result.
 
Thread starter Similar threads Forum Replies Date
W Contract Initial Importer and Order Handling for a Class I Medical Device Service Industry Specific Topics 2
C Safe Handling of Non-Device Refurbished Medical Equipment US Food and Drug Administration (FDA) 3
H Medical device Product Registration Registrars and Notified Bodies 2
A Can a power Supply be an accessory to a medical device, if it is an 'off-the-shelf' product. IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Medical device labelling Date of manufacture US Food and Drug Administration (FDA) 2
W Non Sterile Medical Device Environmental Tests Other Medical Device Related Standards 4
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
K 25-year lifetime of medical device - document storage period EU Medical Device Regulations 1
K Relabeling an existing medical device in the field? Other US Medical Device Regulations 6
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
A Reliable sources for following EU medical device regulatory EU Medical Device Regulations 0
T IVDR Medical device software CE Marking (Conformité Européene) / CB Scheme 8
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
J Requirements as a Distributor for Incoming Inspection of Purchased Finished Medical Device Medical Device Related Regulations 0
S Microwave medical waste disinfectant - A medical device or not? Other ISO and International Standards and European Regulations 3
S Registration of Medical Device in Hong Kong - labeling requirements Other Medical Device Regulations World-Wide 0
V Software as medical device (SaMD) replicated for multiple clients through APIs IEC 62304 - Medical Device Software Life Cycle Processes 4
M Is the output of a device a Medical Device? IEC 62304 - Medical Device Software Life Cycle Processes 5
P Do we need to retrospectively use the "MD" symbol (indicating device is a medical device) on labels, e.g. finished devices within expiration date? EU Medical Device Regulations 2
L Medical device registration in Iran Other Medical Device Regulations World-Wide 0
H EU CE marking for Medical Device Class I EU Medical Device Regulations 2
A Medical Device Contract Manufacturer - Does the CM need to register with FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 4
R Compatibility studies - Medicinal Product and Medical Device Other ISO and International Standards and European Regulations 0
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3
J Medical Device Regulations in Lebanon? Other Medical Device Regulations World-Wide 2
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
S Medical Device MRI Compatibility EU Medical Device Regulations 3
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
R Components to a finished medical device, MDR requirements Other US Medical Device Regulations 1
J Warnings/Cautions in Medical Device IFU Medical Device and FDA Regulations and Standards News 4
L Medical device HIPAA compliance in encryption Medical Information Technology, Medical Software and Health Informatics 1
M V&V phase: Justification of acceptance criteria (statistical method ) - (Medical Device) Design and Development of Products and Processes 2
E Medical Device - CE marking - Local market notifications EU Medical Device Regulations 2
S Medical Device Registration in Qatar Other Medical Device Regulations World-Wide 1
M Medical device substance based-leachables Other Medical Device Related Standards 2
P Anyone have an Idea on UAE Medical device registeration- Class B with FDA only Other Medical Device Regulations World-Wide 0
F Mobile app regulations - Class II medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
M What are the basics of Medical Device Single Audit Program (MDSAP)? ISO 13485:2016 - Medical Device Quality Management Systems 7
U Medical Device CE Marking - Using a disposable bearing CE Marking (Conformité Européene) / CB Scheme 3
L Medical Device Registration in Macau Other US Medical Device Regulations 1
A Medical Device Registration in the Dominican Republic Other Medical Device Regulations World-Wide 4
Aymaneh Medical Device Cybersecurity Risk Management IEC 27001 - Information Security Management Systems (ISMS) 2
T B2C Medical Device Shipping across the US Other Medical Device Related Standards 0
M Medical Device Registration In Malaysia Other Medical Device Regulations World-Wide 2
N Adding unclassified product to the medical device registration US Food and Drug Administration (FDA) 1
V Sister companies selling same medical device under different names ISO 13485:2016 - Medical Device Quality Management Systems 3
K CE Marking for Class I Medical Device? CE Marking (Conformité Européene) / CB Scheme 8
L Medical device storage conditions ISO 13485:2016 - Medical Device Quality Management Systems 1
F USB powered handheld medical device - Isolation requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 1

Similar threads

Top Bottom