When your distributor is performing "installation" activities, who is compiling the procedures/forms of IQOQPQ? I guess it would be you providing your distributor all the IQOQPQ forms along with the instrument (shipment).
Upon successful installation, you have to get the installation records from your distributor.
See below excerpts from 21 CFR 820:
Sec. 820.170 Installation.
(a) Each manufacturer of a device requiring installation shall establish and maintain adequate installation and inspection instructions, and where appropriate test procedures. Instructions and procedures shall include directions for ensuring proper installation so that the device will perform as intended after installation. The manufacturer shall distribute the instructions and procedures with the device or otherwise make them available to the person(s) installing the device.
(b) The person installing the device shall ensure that the installation, inspection, and any required testing are performed in accordance with the manufacturer's instructions and procedures and shall document the inspection and any test results to demonstrate proper installation.
Sec. 820.180 General requirements.
All records required by this part shall be maintained at the manufacturing establishment or other location that is reasonably accessible to responsible officials of the manufacturer and to employees of FDA designated to perform inspections. Such records, including those not stored at the inspected establishment, shall be made readily available for review and copying by FDA employee(s). Such records shall be legible and shall be stored to minimize deterioration and to prevent loss. Those records stored in automated data processing systems shall be backed up.
Regards,
Sreenu