SBS - The Best Value in QMS software

Medical Device Internal Test Certificate Example

Q

QACen

#1
Hello,

We manufacture medical devices and a customer is requesting for documentation of the results of our internal testing prior to shipment. I assume that this would be a test certificate or test confirmation type of document. Can anyone share a template?

Thanks!
 
Elsmar Forum Sponsor

Marc

Fully vaccinated are you?
Staff member
Admin
#2
What do you use in-house to document in-house testing such as final acceptance testing?
 
Q

QACen

#3
Yes. I know I can use information from those documents to summarize for the customer. I was just curious to understand what experienced folks in the industry do for the customer. What goes in that document? What are some things that should be listed? The customer was not specific.
 

Marc

Fully vaccinated are you?
Staff member
Admin
#4
I'm not aware of any specific document, but then again - Every company and every product would be different. If you do final acceptance testing I would think that a copy of that report would be sufficient.

You will find a lot of folks here in posts like this will also say you should ask your customer directly specifically what they want. Their generically asking for something like "...documentation of the results of your internal testing prior to shipment..." would be to me final acceptance testing. Ask them if that is what they want, and if they're "not sure" or something like that, offer them a copy of your final acceptance test report(s).
 
P

Pelau

#5
Actually, they want a certificate of conformance - with the proof with the measurements and the method of testing. Concurs with other comments
 
S

SteveK

#6
Attached is the type of template I use for our customers, when requested. It shows an example of a particular standard (in addition to QMS references) the item complies with and was tested to, as per details in the Technical File.

This format has been accepted over many years in our case thus far.

Hopefully this helps.

Steve
 

Attachments

yodon

Staff member
Super Moderator
#7
I'm curious why such a cert would be necessary. As a medical device, it's in a regulated industry and the regulations require that the production be performed in a manner that (reasonably) ensures the product made will comply with its specifications.

I guess if it makes the sale, it's worth it but would customer education be possible?
 
P

Pelau

#8
In addition to the cert...the proof of the actual specifications were achieved and the inspection or tests carried out.
 
Thread starter Similar threads Forum Replies Date
Q Need Outsourced ISO 13485/FDA Internal Auditor for a Medical Device Company Internal Auditing 3
M Use of an Internal Powerstrip in a Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N Internal Audit Process for small Medical Device company Internal Auditing 9
H Internal Auditing Frequency Requirements in a Medical Device Company ISO 13485:2016 - Medical Device Quality Management Systems 14
S Internal Files for Medical Device Ultrasound System Licence Other US Medical Device Regulations 6
V Biocompatibility of Medical Device Internal Parts Other US Medical Device Regulations 9
H Existing cloud based medical device - questions regarding improving the processes IEC 62304 - Medical Device Software Life Cycle Processes 0
K Unused Service Parts in Newly Manufactured Medical Device? Other US Medical Device Regulations 1
B A.I diagnostic software is considered as medical device in FDA? US Food and Drug Administration (FDA) 5
J Medical device repairs (to upholstery) Manufacturing and Related Processes 4
Ed Panek Apple Provides New Medical Device Cautions 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
C Does a medical device active (zinc oxide) needs BPR registration in EU? Other ISO and International Standards and European Regulations 1
DanBOS Cloud Connected Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
dgrainger Informational DRAFT: The Medical Devices (Coronavirus Test Device Approvals) (Amendment) Regulations 2021 UK Medical Device Regulations 0
S Obsolescence of the Medical Device in Various Countries Other Medical Device Related Standards 1
A IT-NETWORK in PEMS Sub-Clause 14.13 for Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
G Medical Device - Borderline/Definition EU Medical Device Regulations 0
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
P European Medical Device Nomenclature (EMDN) and CND EU Medical Device Regulations 3
S Reseller Request to Change UPC on Medical Device via Re-labeling Medical Device and FDA Regulations and Standards News 5
D Hand Held medical device - power supply requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
E Medical device applicability to WEEE Other ISO and International Standards and European Regulations 2
Fjalar Spare parts for discontinued MDD compliant class I medical device EU Medical Device Regulations 4
H Medical Device Label Acceptance Criteria Manufacturing and Related Processes 4
J Calling a medical device a medical device (when it might not be one..) UK Medical Device Regulations 29
B Regulatory Affairs Certification (RAC) Book - Fundamentals of Medical Device Regulations Medical Device and FDA Regulations and Standards News 0
N Medical device name in different countries EU Medical Device Regulations 4
V Medical Device Literature Translation Software ISO 13485:2016 - Medical Device Quality Management Systems 1
Z Over The Air (OTA) updates for medical device Other US Medical Device Regulations 1
H Tukery Medical Device Regulstion Other Medical Device Regulations World-Wide 0
M Medical device certificate in Australia - ARTG certificate Other Medical Device Regulations World-Wide 0
Q Software as a medical device vs software not sold as medical device: local regulations for sale? EU Medical Device Regulations 4
H Medical device Product Registration Registrars and Notified Bodies 2
A Can a power Supply be an accessory to a medical device, if it is an 'off-the-shelf' product. IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Medical device labelling Date of manufacture US Food and Drug Administration (FDA) 2
W Non Sterile Medical Device Environmental Tests Other Medical Device Related Standards 4
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
K 25-year lifetime of medical device - document storage period EU Medical Device Regulations 1
K Relabeling an existing medical device in the field? Other US Medical Device Regulations 6
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
A Reliable sources for following EU medical device regulatory EU Medical Device Regulations 0
T IVDR Medical device software CE Marking (Conformité Européene) / CB Scheme 8
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
J Requirements as a Distributor for Incoming Inspection of Purchased Finished Medical Device Medical Device Related Regulations 0
S Microwave medical waste disinfectant - A medical device or not? Other ISO and International Standards and European Regulations 3
S Registration of Medical Device in Hong Kong - labeling requirements Other Medical Device Regulations World-Wide 1
V Software as medical device (SaMD) replicated for multiple clients through APIs IEC 62304 - Medical Device Software Life Cycle Processes 5
M Is the output of a device a Medical Device? IEC 62304 - Medical Device Software Life Cycle Processes 5
P Do we need to retrospectively use the "MD" symbol (indicating device is a medical device) on labels, e.g. finished devices within expiration date? EU Medical Device Regulations 2
L Medical device registration in Iran Other Medical Device Regulations World-Wide 0

Similar threads