Medical Device Mechanical Test Method Validation
stevoc
Hi
First post on the forum, but have been lurking for a while.
Can anyone help me please regarding Test Method Validation?
I work for a design consultancy and we've been developing a medical device for a client. We've produced devices for verification testing and will now be doing the verification testing in house.
Our client is of the opinion that we need to validate all our test methods - can anyone give me some advice on where to start please? The tests are mainly mechanical - force & torque to operate, pressure to move, sealing pressure, dose delivery accuracy etc.
How do you validate a mechanical lab test? We tried getting a series of technicians to test the same parts repeatedly, but the nature of what we are testing, means there is significant variability - so we can't show that different operators get the same results from the same test samples.
many thanks
Steve
Ronen E
Problem Solver
Moderator
Hi
many thanks
Steve
Hello and congrats for your 1st post :bigwave:
Mechanical test method validation relates closely to GR&R / MSA -- there's a lot of knowledge under this topic in the forums. Further, there's a lot of valuable guidance related to method validation in general, in the pharmacy world guidance (e.g. ICH guidelines), but you might waste (or invest?) too much time there before you find what you need. Repeating a test several times with different techs is a step in the right direction (GR&R) but I advise a bit of self-education and planning before you proceed.
The fact that results are not reproducible / repeatable is a bit alarming -- there might be a real problem with the method or/and its implementation. GR&R methodology includes tools for improving, once you've established that the results are not satisfactory.
Good luck,
Ronen.
somashekar
Your tests must show that the device performs better than the design inputs.
Your design verification must spell out where and how the requirements of the mechanical parameters must be met.
A test protocol must be made and agreed with the customer, and where instruments are used, they must be duly calibrated.
The test set up must be installed based on the protocol (IQ)
The test results must give you consistently exceeding the required mechanical requirement, over the test samples that are taken up for this activity.
Tests can be done by one qualified engineer and you are all done.
W
Waeller
Dear Stevoc,
what you are describing in your question is exactly what's bothering me. We are also developing test methods for medical devices which are just going into production.
So, there is not yet a stable process and further more most of these mechanical tests are destructive. We are also not able to find parts which are similar enough - because there's no test method to judge - to go for an GRR, though. Did you find some useful hints besides the provided links?
To all others,
besides pointing to a GRR is there anything else out there in the quality community which allows for judging a test method to be robust and measuring the values correctly if there are such limitations as described above?
Thanks for any hint
Waeller
badgary
If you have a lot of variation in your test method making it hard to validate and you can't reduce that anymore, but you insist your method works and can detect what it is supposed to detect, then I suggest trying an attribute type data test method validation. Your sample size will be quite significant because its an attribute test (binary results like pass/fail). The criteria to pass the test will be vague most likely due to the nature of attribute data, thus requiring a lot of samples to have any kind of confidence that your test is valid. But often when there is large variation and everything has been done to control it, you can resort to an attribute test method validation. I believe Minitab (their website) has some white papers on this subject that would be creditable. Good luck.:D
W
Waeller
Hi badgary,
thank you for your hints and ideas, this is a possible way but quite an effort.
What we are now implementing as a first guess, whether we are able to measure something reasonable, is something like the probable error approach mentioned by Wheeler.
Waeller
rajwsu
Does a test method certification from a accreditation program can satisfy the requirements for medical device Test method validation (TMV)?
W
Waeller
this might be a possible solution but who is going to check the accredited company if they are doing the right things? Just being accredited doesn't mean being perfect
Best regards
Patric
J GLP accreditation for mechanical tests of medical device? Other Medical Device and Orthopedic Related Topics 1
H Approval list for mechanical and electronics drawings in a Medical Device company ISO 13485:2016 - Medical Device Quality Management Systems 3
A Mechanical Design Procedure - Infusion pump incorporating software - Medical Device Design and Development of Products and Processes 1
M Medical device certificate in Australia - ARTG certificate Other Medical Device Regulations World-Wide 0
Q Software as a medical device vs software not sold as medical device: local regulations for sale? EU Medical Device Regulations 4
H Medical device Product Registration Registrars and Notified Bodies 2
A Can a power Supply be an accessory to a medical device, if it is an 'off-the-shelf' product. IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Medical device labelling Date of manufacture US Food and Drug Administration (FDA) 2
W Non Sterile Medical Device Environmental Tests Other Medical Device Related Standards 4
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
K 25-year lifetime of medical device - document storage period EU Medical Device Regulations 1
K Relabeling an existing medical device in the field? Other US Medical Device Regulations 6
A Reliable sources for following EU medical device regulatory EU Medical Device Regulations 0
T IVDR Medical device software CE Marking (Conformité Européene) / CB Scheme 8
J Requirements as a Distributor for Incoming Inspection of Purchased Finished Medical Device Medical Device Related Regulations 0
S Microwave medical waste disinfectant - A medical device or not? Other ISO and International Standards and European Regulations 3
S Registration of Medical Device in Hong Kong - labeling requirements Other Medical Device Regulations World-Wide 0
V Software as medical device (SaMD) replicated for multiple clients through APIs IEC 62304 - Medical Device Software Life Cycle Processes 5
M Is the output of a device a Medical Device? IEC 62304 - Medical Device Software Life Cycle Processes 5
P Do we need to retrospectively use the "MD" symbol (indicating device is a medical device) on labels, e.g. finished devices within expiration date? EU Medical Device Regulations 2
L Medical device registration in Iran Other Medical Device Regulations World-Wide 0
H EU CE marking for Medical Device Class I EU Medical Device Regulations 2
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 4
R Compatibility studies - Medicinal Product and Medical Device Other ISO and International Standards and European Regulations 0
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3
J Medical Device Regulations in Lebanon? Other Medical Device Regulations World-Wide 2
S Medical Device MRI Compatibility EU Medical Device Regulations 3
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
R Components to a finished medical device, MDR requirements Other US Medical Device Regulations 1
J Warnings/Cautions in Medical Device IFU Medical Device and FDA Regulations and Standards News 4
L Medical device HIPAA compliance in encryption Medical Information Technology, Medical Software and Health Informatics 1
E Medical Device - CE marking - Local market notifications EU Medical Device Regulations 2
S Medical Device Registration in Qatar Other Medical Device Regulations World-Wide 1
M Medical device substance based-leachables Other Medical Device Related Standards 2
P Anyone have an Idea on UAE Medical device registeration- Class B with FDA only Other Medical Device Regulations World-Wide 0
F Mobile app regulations - Class II medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
M What are the basics of Medical Device Single Audit Program (MDSAP)? ISO 13485:2016 - Medical Device Quality Management Systems 7
U Medical Device CE Marking - Using a disposable bearing CE Marking (Conformité Européene) / CB Scheme 3
L Medical Device Registration in Macau Other US Medical Device Regulations 1
A Medical Device Registration in the Dominican Republic Other Medical Device Regulations World-Wide 4
Aymaneh Medical Device Cybersecurity Risk Management IEC 27001 - Information Security Management Systems (ISMS) 2
T B2C Medical Device Shipping across the US Other Medical Device Related Standards 0
M Medical Device Registration In Malaysia Other Medical Device Regulations World-Wide 2
N Adding unclassified product to the medical device registration US Food and Drug Administration (FDA) 1
V Sister companies selling same medical device under different names ISO 13485:2016 - Medical Device Quality Management Systems 3