Medical Device MRI Compatibility
Relatively new to the regulatory game and still finding my feet somewhat. I am currently updating our some of our Instructions for Use as part of ensuring compliance to MDR. Is it right that we need a section “Use of the device during diagnostic, investigation, evaluations or treatment” I thought I had seen this in the MDR but on reviewing I can’t find it. If that information is required, then I presume it would predominantly relate to using the devices during x-ray or MRI. I guess that we would have to measure against ASTM F2052, is that correct? None of the devices contain metal so would we still have to prove against a standard? Thanks all in advance.
indubioush
From MDR:
REQUIREMENTS REGARDING DESIGN AND MANUFACTURE
14.2 Devices shall be designed and manufactured in such a way as to remove or reduce as far as possible: ....(b) risks connected with reasonably foreseeable external influences or environmental conditions, such as magnetic fields, external electrical and electromagnetic effects
From ISO 14630:
5 Design attributes
The design attributes to meet the intended performance shall take into account at least the following:
n) effects of radiation, electromagnetic and magnetic fields ....
NOTE Particular attention is drawn to the fields used for magnetic resonance imaging (MRI) in respect of patient
safety. The test methods in ASTM F2052, ASTM F2119, ASTM F2182 and ASTM F2213 can be used to evaluate the
safety of an implant in the MR environment.
Hi Indubioush,
Thanks for the info. Do we have to test against the standards or is it permissible to say that from the materials used in construction it is not expected to be affected by MRI?
Scott
indubioush
You will need to have adequate scientific information that supports your MR labeling claim (MR safe, MR conditional). It sounds like you believe your product to be MR safe. To support this claim, you will need to have a documented review of your materials against published literature or other scientific sources. Please do not think that your product is MR safe just because there are no metal components. You need to do a detailed review of every material and configuration. If this review provides adequate information to support the MR safe claim, you may be able to justify not doing the tests.
Please purchase and read ASTM F2503 thoroughly. Also research other MR labeling regulations or guidelines applicable to you.
Note that your design history file needs to show traceability between MR related design inputs, related risk assessment/evaluation, product requirements, and your scientific review document.
M Medical device certificate in Australia - ARTG certificate Other Medical Device Regulations World-Wide 0
Q Software as a medical device vs software not sold as medical device: local regulations for sale? EU Medical Device Regulations 2
H Medical device Product Registration Registrars and Notified Bodies 2
A Can a power Supply be an accessory to a medical device, if it is an 'off-the-shelf' product. IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Medical device labelling Date of manufacture US Food and Drug Administration (FDA) 2
W Non Sterile Medical Device Environmental Tests Other Medical Device Related Standards 4
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
K 25-year lifetime of medical device - document storage period EU Medical Device Regulations 1
K Relabeling an existing medical device in the field? Other US Medical Device Regulations 6
A Reliable sources for following EU medical device regulatory EU Medical Device Regulations 0
T IVDR Medical device software CE Marking (Conformité Européene) / CB Scheme 8
J Requirements as a Distributor for Incoming Inspection of Purchased Finished Medical Device Medical Device Related Regulations 0
S Microwave medical waste disinfectant - A medical device or not? Other ISO and International Standards and European Regulations 3
S Registration of Medical Device in Hong Kong - labeling requirements Other Medical Device Regulations World-Wide 0
V Software as medical device (SaMD) replicated for multiple clients through APIs IEC 62304 - Medical Device Software Life Cycle Processes 5
M Is the output of a device a Medical Device? IEC 62304 - Medical Device Software Life Cycle Processes 5
P Do we need to retrospectively use the "MD" symbol (indicating device is a medical device) on labels, e.g. finished devices within expiration date? EU Medical Device Regulations 2
L Medical device registration in Iran Other Medical Device Regulations World-Wide 0
H EU CE marking for Medical Device Class I EU Medical Device Regulations 2
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 4
R Compatibility studies - Medicinal Product and Medical Device Other ISO and International Standards and European Regulations 0
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3
J Medical Device Regulations in Lebanon? Other Medical Device Regulations World-Wide 2
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
R Components to a finished medical device, MDR requirements Other US Medical Device Regulations 1
J Warnings/Cautions in Medical Device IFU Medical Device and FDA Regulations and Standards News 4
L Medical device HIPAA compliance in encryption Medical Information Technology, Medical Software and Health Informatics 1
E Medical Device - CE marking - Local market notifications EU Medical Device Regulations 2
S Medical Device Registration in Qatar Other Medical Device Regulations World-Wide 1
M Medical device substance based-leachables Other Medical Device Related Standards 2
P Anyone have an Idea on UAE Medical device registeration- Class B with FDA only Other Medical Device Regulations World-Wide 0
F Mobile app regulations - Class II medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
M What are the basics of Medical Device Single Audit Program (MDSAP)? ISO 13485:2016 - Medical Device Quality Management Systems 7
U Medical Device CE Marking - Using a disposable bearing CE Marking (Conformité Européene) / CB Scheme 3
L Medical Device Registration in Macau Other US Medical Device Regulations 1
A Medical Device Registration in the Dominican Republic Other Medical Device Regulations World-Wide 4
Aymaneh Medical Device Cybersecurity Risk Management IEC 27001 - Information Security Management Systems (ISMS) 2
T B2C Medical Device Shipping across the US Other Medical Device Related Standards 0
M Medical Device Registration In Malaysia Other Medical Device Regulations World-Wide 2
N Adding unclassified product to the medical device registration US Food and Drug Administration (FDA) 1
V Sister companies selling same medical device under different names ISO 13485:2016 - Medical Device Quality Management Systems 3
K CE Marking for Class I Medical Device? CE Marking (Conformité Européene) / CB Scheme 8
L Medical device storage conditions ISO 13485:2016 - Medical Device Quality Management Systems 1
F USB powered handheld medical device - Isolation requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
L How to determine / validate Medical Device Storage Conditions ISO 13485:2016 - Medical Device Quality Management Systems 1