Dear all,
Like often here, I'm piggybacking an existing thread

.
So just to clarify on what have been said above - please correct me if I am wrong :
- ASTM F1980 is a "one size fit all" accepted by the regulators - but it is not on the harmonized list?
- There is no other standard (harmonized or not) besides of ISO11607 widely used by the medical device world to specify an accelerated aging protocol.
- Both these standards are dedicated to sterile products yet are used -by lack of an alternative - for non-sterile produy\cts as well.
- None of these standards, and none known standard, does provide a guidance for the parameters to use to specify accelerated aging when the labelled storage conditions are given as a range
Reason I'm asking:
Like OP my product (IIa, non sterile) is labelled with a range for temp and humidity. We have been historically claiming a five years shelf life. The documented evidence, which postdates the claim, is a test report we put together after comparing retention samples and new products with an in-house protocol.
I need now to change supplier for the material of a non-critical part of the product ; new supplier can provide BioC etc... + letter stating they have validated their 2 years shelf life.
I would like to avoid having to decrease the labelled shelf life claim for 2 reasons:
- Inventory / supply management wise - it is much easier for us to manage a longer shelf life
- A change of label would trigger notification/review by our National regulatory body (who is also our NB) and FW to our regisster agents
Without this change of shelf life - the change can be considered non significant (same material, just different supplier)
The new supplier cannot provide 5 years old samples which I could compare to new parts. But if they can provide say 3 years old samples, kept in uncontrolled conditions, would it be acceptable to submit these samples to a "2years scope accelerated aging" then compare with new samples and claim 5 years?
Is it acceptable to follow the ASTM recommendations with using 25C "storage temp" even though the label says -10 to 40?
Does somebody around have any recommendation regarding the best aging protocol (cycles?) for polymer foams and fabric laminates?
I have spent a few hours today reading several threads here and elsewhere about this topic. Altogether, it seems to me that regulators have been in the past couple of years raising up the steam on this subject - very often completely overlooking the "commensurate to the risk" approach. I have seen countless stories where FDA or NBs were asking for documented test reports after refusing written rationale in spite of an obvious very low risk