Medical Devices - Ancillary - ISO 10993-17
#1
Dear readers,
we have a device which is an adhesive gel that contains Chlorhexidine (CHX) at 0,12% to be applied to an injured wound. The physical barrier is created by the gel and does not allow bacteria to reach the wound. We would like to say that the CHX has antimicrobial properties and reduces risk of infection. However, the % (0,12% used) is also accepted as preservative in cosmetics for instance.
How will this device be classified? if we claim the antimicrobial properties, the device will fall within rule 14? how can we demonstrate that it is a MD and not a drug? should we present any clinical data comparing the gel with vs gel without CHX?
or if we claim that the CHX is preservative, will the device fall under rule 21?
Does ISO 1099317 apply to the CHX? is CHX considered a leachable?
many many thanks!
Elsmar Forum Sponsor
#3
Dear planB,
thank you very very much for your answer!!!!.
Reading the thread that you provided in your answer, I understand that, if we are able to demonstrate that the CHX acts as a preservative an it is a component of the gel, we will fall under Rule 14.
However, the CHX is likely to "leach" de gel and act on the wound so I think that we are more likely to fall under rule 21. Do you agree?
and hopefully, emphasising that the gel is very important since it acts as a physical barrier to the microorganisms, we hope that the Notified Body will not consider the CHX to have the main action!
thanks again!
marie
planB
Super Moderator
#4
marie,
since you described your device forming a stable physical barrier of compromised skin, the device would not be regarded as being absorbed or dispersed in the body, even when CHX quantitatively "leaches" to the patient; thus, Rule 21 would not apply.
Rather Rule 4 applies, presuming that you just claim preservative properties for CHX. Rule 14 would only apply if you additionally claimed therapeutic modes of action for CHX, turning CHX into an ancillary medicinal substance.
You might also want to to re-confirm your classification with your notified body.
Hope this helps,
#5
Dear planB,
really appreciate your feedback!
Apologies since I think that I mixed up a bit the classification rules in my last post...
Therefore, I'll summarise your input as follows:
- Rule 4 will apply if we just claim that the CHX has a preservative action. --> class IIa or IIb
-Rule 14 will apply if we claim an ancillary action to the CHX (i.e. antiseptic). Even if CHX is present in the device in a % considered to be preservative, we are claiming an ancillary action, thus rule 14 apply --> class III
-Rule 21 will apply if the device itself (the gel) is absorbed or dispersed. Class IIa will apply since the action takes place in. the mouth/oral cavity. From my understanding, if we want to follow under rule 21, we can't claim any antiseptic/ancillary mode of action to the CHX., is that right?
thank you very much gain, really appreciate your support.
marie
planB
Super Moderator
#6
marie,
sorry that I wrongly assumed that your device is non-invasive. Since you clarified that the location of application is the mouth/oral cavity, your device is invasive and thus, either Rule 5 (not Rule 4) / Rule 21 (medical device) apply, or Rule 14 (combination device): You cannot have a purely medical device with a claimed ancillary therapeutic substance.
Hope this helps,
#7
perfect, really helpful!!
and releted to the leachables, I understood that in case that CHX leaches de device, yes, it will be considered a leachable.
I was a little bit confuses because the ISOs, when describing leachables, only talk about "chemicals, antioxidants polymers, additives, etc) but never related to medicinal substances. This is why I thought that medicinal substances will not be considered leachables.
thanks!
Thread starter Similar threads Forum Replies Date
M Europe GMP medical devices - Combination products - Ancillary medicinal substance CE Marking (Conformité Européene) / CB Scheme 4
A Brexit Mandate for EU Authorised Representative for non medical devices CE Marking (Conformité Européene) / CB Scheme 2
A Applicability of Photobiological Safety Evaluation for LED used in medical devices Reliability Analysis - Predictions, Testing and Standards 1
T Google Fit Health datas and Medical devices Application Medical Information Technology, Medical Software and Health Informatics 0
B Procedure packs with non-medical devices EU Medical Device Regulations 1
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
A Customer Approval (Medical Devices) Document Control Systems, Procedures, Forms and Templates 4
D Electrical Medical Devices class I EU Medical Device Regulations 0
J Are complaints applicable to development of medical devices? Customer Complaints 2
P Do we need to retrospectively use the "MD" symbol (indicating device is a medical device) on labels, e.g. finished devices within expiration date? EU Medical Device Regulations 2
Brian Benzarti Medical Devices CLP Regulation Quality Assurance and Compliance Software Tools and Solutions 2
dgrainger Informational Medicines and Medical Devices Act 2021 UK Medical Device Regulations 0
A Sampling plan for in-process QC (medical devices) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
B Free Sales Certificate for Non Medical Devices Other Medical Device Related Standards 2
K Usability for custom made medical devices IEC 62366 - Medical Device Usability Engineering 5
dgrainger Informational MHRA - Regulating medical devices in the UK - Post Brexit - 2021-02 UK Medical Device Regulations 0
L Shelf life of medical devices Other Medical Device Related Standards 4
S How is the service life for class I reusable medical devices described in MDR? EU Medical Device Regulations 7
S How is the service life described for class I reusable medical devices in MDR? Elsmar Xenforo Forum Software Instructions and Help 1
M Brexit The Medical Devices (Amendment etc.) (EU Exit) Regulations 2020 UK Medical Device Regulations 1
P New Global HQ Suggestions for Virtual manufacturing/own brand labelling of medical devices? EU Medical Device Regulations 4
N Chemical Testing on Medical Devices - Solutions in a container closure system (bag) EU Medical Device Regulations 1
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Subcontracting a repair of Medical Devices Canada Medical Device Regulations 0
S Regrind Material in Medical Devices ISO 13485:2016 - Medical Device Quality Management Systems 1
R Product Recall - medical devices in the hospital warehouse Canada Medical Device Regulations 2
P Monolithic application and medical devices ISO 13485:2016 - Medical Device Quality Management Systems 2
N Medical Devices in oxygen rich environment IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Medical Devices Shelf Life ---INDIA Other Medical Device Regulations World-Wide 3
JoCam Warranty of Medical Devices EU Medical Device Regulations 2
R Obsolesce medical devices ISO 13485:2016 - Medical Device Quality Management Systems 1
dgrainger Informational UK - Medicines and Medical Devices Bill 2019-21 UK Medical Device Regulations 0
S Call it refurbish or not? Medical Devices Manufacturing and Related Processes 2
C Registering Medical Devices in Australia and New Zealand - ANZTPA Other Medical Device Regulations World-Wide 0
S Requirements for "Claims Guide" for medical devices Medical Device and FDA Regulations and Standards News 6
C FDA Establishment registration - Buying some medical devices from another manufacturer Medical Device and FDA Regulations and Standards News 5
G Accelerated stability of medical devices Other Medical Device Related Standards 1
A Medical Device Vigilance decision tree for Japan for class 2 devices. Japan Medical Device Regulations 1
K Contamination Control - Class Is medical devices (Clause 6.4.2 ISO 13485:2016 (E)) ISO 13485:2016 - Medical Device Quality Management Systems 12
W Seeking Guidance Verification Test Strategy for Class B Medical Devices IEC 62304 - Medical Device Software Life Cycle Processes 1
G Free Sales certificates for Medical Devices - List of countries EU Medical Device Regulations 4
dgrainger Informational Parliament to vote on postponing new requirements for medical devices -17th April Medical Device and FDA Regulations and Standards News 1
K Old medical devices -> 7.3.7. Design and development validation ISO 13485:2016 - Medical Device Quality Management Systems 1
M Who needs a MDEL? How to buy and sell medical devices on my own Canada Medical Device Regulations 14
D Social Media Feedback process for Medical Devices ISO 13485:2016 - Medical Device Quality Management Systems 6
M Post Market Surveillance for Class II Medical Devices Medical Device and FDA Regulations and Standards News 2
M Special Controls and Class II Medical Devices Medical Device and FDA Regulations and Standards News 18
dgrainger Informational Have your say on the UK's Medicines and Medical Devices Bill Medical Device and FDA Regulations and Standards News 0
L Packaging unsterile medical devices ISO 13485:2016 - Medical Device Quality Management Systems 6
R Expected life span of Medical devices - X-Ray based diagnostic medical devices ISO 13485:2016 - Medical Device Quality Management Systems 6
Similar threads