Medical Gas Pressure Regulators - Anybody had testing conducted at 450 bar? ISO 10524
S
SteveK
#1
Is any one else out there familiar with a specific medical device standard for ‘Pressure regulators for use with medical gases’ – ISO 10524-1:2006? One test for mechanical strength (5.4.10.1) requires that the high pressure side of the regulator shall be capable of withstanding 2.25 times its inlet pressure for 5 minutes. Now a new product of ours is to be used with gas cylinders at 200bar i.e. this means testing at 450bar. Personally I would not be in the same street with something at that pressure. A test laboratory will have to be used and I do know a likely place to approach. The question is, has any body out there actually conducted or requested testing to this pressure level? Bit of a long shot I know - but if you don't ask.
Steve
Elsmar Forum Sponsor
ScottK
Not out of the crisis
Leader
Super Moderator
#2
Re: Medical Gas Pressure Regulators - Anybody had testing conducted at 450 bar? ISO 1
Steve
wow. for those of use that measure in PSI that's 6527.
My company makes regulators for the same market, Steve. We might even be competetors. :mg:
I'm looking at the CGA E-7 spec now which is what we use. CGA requires 4 times.
It can be done hydrostatically. Our designs were proof tested long ago so I don't know if they were done in-house or sent out. We have a pretty high pressure hydrostatic test rig, but it's relatively new.
S
SteveK
#3
Hi Scott,
As you indicate hydrostatic testing would be the safest bet - the standard I quoted does not say specifically test with gas pressure (sort of infers it though). Anyway we do not supply the North American market – so we cannot be competitors there at least!:)
Cannot imagine having to test at 4x!
It's nice to know there is somebody else in the Cove who works in the same field - I thought I'd end up being a bit lonely.:(
Steve
Marcelo
Inactive Registered Visitor
#4
I´ve seen both gas and hydrostatic tests in standards, so i would say that both would be acceptable if there were no explicit requirement.
This is great, you might choose between a potential explosion and a potential flood.
ScottK
Leader
Super Moderator
#5
I'll take the flood any day....
ScottK
Leader
Super Moderator
#6
Hi Scott,
Steve
You're never lonely at the Cove. There's always someone who has a similar issue and misery loves company. :D
Thread starter Similar threads Forum Replies Date
V Cleanliness/Bioburden for Non-Sterile Gas Path Medical Devices Other Medical Device Regulations World-Wide 7
S Medical Gas Change Over Unit - Which Standard to use? Other Medical Device Related Standards 4
G Is Medical Grade Gas always a drug? Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 8
J Does EN ISO 13485 necessary for Class I (non-sterile) medical devices ISO 13485:2016 - Medical Device Quality Management Systems 0
L Directive 2012/19/EU for medical software EU Medical Device Regulations 1
S Sample size considerations in medical process qualification Reliability Analysis - Predictions, Testing and Standards 10
B Medical device disposables - Stored in a controlled environment for 4 months - Effects on Sterilization Medical Device and FDA Regulations and Standards News 1
W MOOPs in Medical Device Charger IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S Specification (Significant digits) - Measurement (Medical device) ISO 13485:2016 - Medical Device Quality Management Systems 3
S Accessory to medical device or not? EU Medical Device Regulations 1
M Looking for medical grade wall adapter with 50Hz power transformer IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
E Microbial Barrier Testing for Medical Devices Qualification and Validation (including 21 CFR Part 11) 2
R Medical equipment selection for refurbishment US Food and Drug Administration (FDA) 1
E Sealing Validation for Medical devices Qualification and Validation (including 21 CFR Part 11) 5
K Question on registration of class 1 medical device in Turkey Other Medical Device Regulations World-Wide 0
A Safety testing of Medical devices ISO 13485:2016 - Medical Device Quality Management Systems 2
A Safety testing for Medical devices to be sold in India ISO 13485:2016 - Medical Device Quality Management Systems 0
G China registration update medical devices China Medical Device Regulations 3
N Class II Medical Device with Metallic Enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
S Export of CE marked medical devices outside of Europe EU Medical Device Regulations 1
D Swiss Documentation Requirements for Class 1 Medical Device Other Medical Device Related Standards 0
A Medical Device including a smartphone - CE mark ? CE Marking (Conformité Européene) / CB Scheme 4
C VOCs as extractable in substance based medical device EU Medical Device Regulations 0
D Correct form to amend medical devices class III (IMDRF vs Form 2016) Canada Medical Device Regulations 0
M Servicing Medical Lasers IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S Mexico medical device registration - for demo Other Medical Device Regulations World-Wide 1
R Gibraltar refurbished medical devices UK Medical Device Regulations 1
L Use of Medical grade plastics in medical devices EU Medical Device Regulations 3
M Medical Devices Regulations and labelling requirements UAE Other Medical Device Regulations World-Wide 0
dgrainger Informational MHRA Medical Devices Regulation Webinar - 24 January 2023, 10:00-11:30 GMT UK Medical Device Regulations 0
S EU Medical Device Regulations EU Medical Device Regulations 6
K Sample size calculation for the medical device with multiple indications. e.g. suture US Food and Drug Administration (FDA) 3
S Registation different version of Medical Devices & Germany competend authority EU Medical Device Regulations 3
Doninina Trend Reporting for medical device class IIb according to MDR EU Medical Device Regulations 4
C Medical Device Gamma Irradiation Validation per VDmax25 (ISO 11137) Qualification and Validation (including 21 CFR Part 11) 1
S Medical device other market Medical Device and FDA Regulations and Standards News 1
drwallice Registration of medical devices EU Medical Device Regulations 2
C Medical Simulators CE Marking (Conformité Européene) / CB Scheme 5
MaHoDie Is it possible to assign medical software to security class A (according to IEC 62304)? EU Medical Device Regulations 8
Stoic Warning letter examples for medical device companies related to the pharma guidance on data integrity? US Medical Device Regulations 5
A 510(k) and DeNovo on same medical device for different indications? US Medical Device Regulations 4
I ISO13485 Folder Structuring for Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 5
S Change in material vendor - Medical device ISO 13485:2016 - Medical Device Quality Management Systems 4
dgrainger Informational Published - Common specifications for the groups of products without an intended medical purpose listed in Annex XVI EU Medical Device Regulations 1
I Medical Device Accessory Responsibility Other Medical Device Related Standards 7
M MDSAP Medical Devices Suppliers Canada Medical Device Regulations 5
C. Tejeda Process validation of rework assembly methods (medical devices) Medical Device and FDA Regulations and Standards News 3
B FDA regulations medical device prescription fulfillment DME question US Medical Device Regulations 0
K What are the differences between the certificates required for home and medical medical equipment? US Food and Drug Administration (FDA) 1
N FDA medical devices US Food and Drug Administration (FDA) 1
Similar threads
Top Bottom