Megger Resistance Tester - How would you control for recalibration? Low Use
How would you handle this?
We purchased a Resistance tester last year for a specific job.
We used it that one time to check about 300 parts.
We have not used it since.
I have it in our recall system, but am not sure what to do at this point.
How stable are these items in storage?
Should I require recal after so many parts?
Recall before next use?
Just want to see how others handle this. Heck we may never use the thing agian.
James
BradM
Re: Megger Resistance Tester - low use - How would you control for recalibration?
Hey, James. I have a customer that has four or five of these I have been calibrating for years. There are a couple of different MFG. of the instruments. They have always been repeatable. The only time I have seen these instruments have problems is when the battery starts running down.
Personally, I might have it checked out after the 300 runs. If you have some certified multimeters, you can buy a high voltage Fluke probe for a $100 bucks or so, and do a quick verification of it if you would like. Basically, you could make it a "Calibrate upon use" instrument.
If you feel pretty certain your're not going to use it, I might sell it. It would be better than keeping up with something you are probably not going to use.
If you store it, I would pull the batteries.
You could make it inactive, tag it and store it. Put you a one year note in Outlook. If in one year it's still sitting in the same spot, I would E-Bay it.
Jerry Eldred
There is a wide range of accuracies/stabilities, depending on what particular type you have. Don't know if I am really adding much to the previous reply, as he seemed to cover it pretty well. But for the sake of getting a "post" in, here's my two cents.
How often you should calibrate it depends on the manufacturer's specs. I have a number of mental pictures of what you may have (something like a Universal brand tester, a Biddle Megger, or an Agilent LCR meter, etc...>>NOTE: No recommendation of brands intended; only to show examples).
The baseline should be the manufacturer's recommended interval. For the Biddle or Agilent, that may be 12 months or something like that. For the Universal tester, it may be much shorter.
As one who has managed numerous calibration programs, if it were in my cognizance, I would recommend making it INACTIVE, and clearly labeled as such. If you have quality compliance requirements, you might strongly consider doing a calibration prior to making it INACTIVE (to assure it remained within specs for the entire time you used it to measure product).
Another alternative (if you expect to continue to use it, AND you use it at a particular resistance value) would be to purchase and certify a check (GOLDEN) standard. That is, a highly accurate and stable reference resistor to periodically check it with. If you use a narrow range of resistance values, you could get resistors to cover that range.
Of course, as I said, the previous poster covered very well most options, so I don't want to repeat what has already been said.
E 60950-1 resistance to abnormal heat test Various Other Specifications, Standards, and related Requirements 0
G Resistance to solve Issues Customer Complaints 8
A ISO 11135:2014, B.1.4, BI resistance x product bioburden ISO 13485:2016 - Medical Device Quality Management Systems 6
R Laminate ESD Mats - Low resistance (RTT) readings Manufacturing and Related Processes 6
T Tests IEC 60601-1 - Grounding Resistance, Leakage Current and Dielectric Rigidity in an automated way IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
L Standard platinum resistance thermometers - sources? General Measurement Device and Calibration Topics 2
J IEC 60601-1 Electrical Safety testing - Class 1 ME externally powered - Earth resistance testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Electrical contact resistance testing in MIL-DTL-5541 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
V Regarding Corrosion resistance test Manufacturing and Related Processes 9
A Wire Holder for Resistance Measurement of Stranded Aluminum Wire Manufacturing and Related Processes 2
0 How to deal with resistance to GDP Document Control Discipline Document Control Systems, Procedures, Forms and Templates 7
M Kaye Instrument 373 RTD Monitor Resistance Curve General Measurement Device and Calibration Topics 2
L How to deal with resistance from auditee(s) Internal Auditing 20
C Japan Regulations - Resistance of Medical Gloves to Permeation Japan Medical Device Regulations 1
H Resistance Temperature Detectors (RTD) Manufacturer's Specifications General Measurement Device and Calibration Topics 3
M Calculating Coefficients for an RTD (Resistance Temperature Detector) Probe General Measurement Device and Calibration Topics 3
J Resistance to Change as a Master Thesis in Quality Assurance Professional Certifications and Degrees 7
A Contact Resistance issue - Gold plating on PCB Manufacturing and Related Processes 3
D Gage R&R on a Load Cell fixture that tests the resistance of a Headrest Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
G OEM resistance to Supplier / PPAP Change APQP and PPAP 5
C Testing for Chemotherapy Drugs Resistance ISO 13485:2016 - Medical Device Quality Management Systems 2
A Resistance Welding Customer Complaints - Cracking and Pitting Root Causes Customer Complaints 4
C Experience with Electronic Resistance Simulation? General Measurement Device and Calibration Topics 2
M Implementation of Quality Circle Proposals? Teams are getting resistance Misc. Quality Assurance and Business Systems Related Topics 1
Jerry Eldred Guildline 9975 DC Comparator Resistance Bridge Manual Needed (Obsolete) General Measurement Device and Calibration Topics 2
M Latest standard for thermocouples & iprt (industrial platinum resistance thermometer) General Measurement Device and Calibration Topics 5
somashekar Optical epoxy for LED having good thermal shock resistance. Manufacturing and Related Processes 2
P Test equipment for resistance welded wire/rebar Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
L Looking for resistance welding FMEA examples FMEA and Control Plans 10
Antonio Vieira PFMEA for Resistance Welding, Winding and Soldering processes FMEA and Control Plans 3
C Resistance Static Testing of Stampings - M.O.ESS.F1B.F185 Various Other Specifications, Standards, and related Requirements 4
S Testing Laboratory For Fire Resistance Test Reliability Analysis - Predictions, Testing and Standards 3
K Resistance welding process - Cold rolled and some zinc plated welding Various Other Specifications, Standards, and related Requirements 2
K Resistance Welder Settings for cold rolled and galvanized steel Manufacturing and Related Processes 3
R Seek Corrosion-resistance Data for Galvanized and Galvannealed steel Manufacturing and Related Processes 1
B How have you overcome resistance to quality? Misc. Quality Assurance and Business Systems Related Topics 5
J What is the Effect of Temperature on OHM (resistance) test? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
D Calibration Standard N24: PFM NO 1841 - Ultrasonic Inspection of Resistance Spot Weld Document Control Systems, Procedures, Forms and Templates 1
J Need Help With Electrical Resistance Problem Manufacturing and Related Processes 6
A Resistance (ohm) Electrical Measurements with a DMM - Verify Shunt? General Measurement Device and Calibration Topics 14
R VisualSPC: Anyone else using this software? Resistance from experienced operators Statistical Analysis Tools, Techniques and SPC 14
I Calibration by correlated parameter - Equipment to measure resistance General Measurement Device and Calibration Topics 3
S Shore A hardness Tester Acceptance Criteria Standard Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 2
B Photovoltaic Tester Calibration General Measurement Device and Calibration Topics 0
P Test Method Validation for Tensile Strength Tester Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
A Tubing Clog Tester - Convert from Air Bubble to Flow Rate Reliability Analysis - Predictions, Testing and Standards 1
F Leak Tester Gage R&R Sheet Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
O Capability or Gage R&R Study for Leak Tester Reliability Analysis - Predictions, Testing and Standards 11
S GR&R for Hardness Tester and Arc Spark Spectrometer Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 15