Meters are on fabric folding machines. How should we calibrate these meters?

U

Uteesh Dhar - 2006

#1
Caliberation

A very simple question?

We are a textile fabric company. Meters are on the folding machines. How should we caliberate these meters?

At present, we are folding one piece of fabric of 30 yards and then after unfolding the same ,m measure the fabric with normal inch tape available in the market.

How to caaliberate this inch tape too?

If any body helps?
 
Elsmar Forum Sponsor

Jerry Eldred

Forum Moderator
Super Moderator
#2
If your purpose is only to make sure your equipment is accurate enough for your product, I suggest having a tape measure calibrated, and use the tape measure to check your machine. I worked in a laboratory that calibrated tape measures. One detail I found was that some tape measures are very good, and some tape measures are not very good. I would advise purchase of a good tape measure. There are some good quality companies (such as Starrett, Brown & Sharpe, Mitutoyo, and others) that make accurate tape measures.

If the purpose of your tape measure is to keep an auditor happy, you will also need to have some documentation as to how you check your folding machine. If you need more detail, please ask, and I will try to help.

------------------
 
R

Rick Goodson

#3
Jerry makes a very good point about the tape measures. I would suggest one step further. Do the finacial analysis on the cost of calibrating the tape measures versus the cost of scrapping and replacing on a periodic basis. In a past life the replacement option was considerably less expensive than calibrating (verifying) the tape measures
 

Jerry Eldred

Forum Moderator
Super Moderator
#4
I would not have a problem with the possibility of replacing rather than re-calibration in some circumstances.

I might be hesitant about that method in some circumstances. That is, if you discard the tape measure which you purchased with a certification, and replace it with another, you will lose continuity of your measurements.

For example, Let us say you purchase a new certified tape measure, and use it for a year. At the end of the year you discard it, and replace it with another new one of the same type.

You have no way of knowing if the old tape measure ever went out of tolerance, because you have no historical data. The possible way around that would be to obtain some documentation from the manufacturer as to how long it stays in tolerance. And believe me, they can and do go out of tolerance. And some of the manufacturer's certificates are not correct. I won't mention any manufacturer names as that would not be appropriate. But I have attempted to recertify tape measures that had certificates with them, and the tape measures were out of tolerance. And when they are out of tolerance, they must be discarded.

One alternative might be to purchase a 'good' three foot or one meter ruler with a certificated. Make that the standard, and only recalibrate it once every two or three years. Use that to verify the tape measures instead of sending them out. The details will be important, as you will have to properly specify an adequate ruler, and properly specify what level of accuracy you are verifying on the tape measure.

------------------
 
Thread starter Similar threads Forum Replies Date
W Banana jack for Fluke meters...? Calibration and Metrology Software and Hardware 11
G In-House Calibration of Gas Flow Meters General Measurement Device and Calibration Topics 6
M EN ISO 15197:2015 - Glucose meters that were CE marked to EN ISO 2013 Other Medical Device Related Standards 2
V Converting Newtons to Newton Meters APQP and PPAP 4
Ronen E Abbott Voluntary Recall of FreeStyle lnsulinx? Blood Glucose Meters Other Medical Device and Orthopedic Related Topics 0
M Calibration of Digital Panel Meters General Measurement Device and Calibration Topics 4
6 Acceptance Criteria For pH Meters and other Laboratory Equipment Measurement Uncertainty (MU) 16
N Calibration of Water Meters which dispense the water into our mixing vessel General Measurement Device and Calibration Topics 12
R Calibration of Jensen JTM-30 clamp meters (estimation) General Measurement Device and Calibration Topics 4
C Testing Agencies - Gas Meters and Hand Held Analyzers General Measurement Device and Calibration Topics 2
M Calibration of magnetometers (Gauss Meters) General Measurement Device and Calibration Topics 3
A Standard for Calibration of Conductivity Meters General Measurement Device and Calibration Topics 8
T Gage R&R studies for Spectrophotometer & Gloss Meters General Measurement Device and Calibration Topics 32
C Calibrating Welding Machines - Any way to justify not calibrating the meters? General Measurement Device and Calibration Topics 15
V METER CLASSIFICATION - Accuracy values of the various classes of meters General Measurement Device and Calibration Topics 1
Jerry Eldred Calibrating Flow Meters - Rotameters that operate with pressurized air General Measurement Device and Calibration Topics 1
M Fabric Composition Labelling required for Medical Devices? Other US Medical Device Regulations 1
M ISO 10993 for Fabric based Electrode - Finding already FDA Approved Materials Other Medical Device Related Standards 11
F Sampling Plan for Fabric being used in a Medical Product Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
Marc BMW Introduces GINA Concept Car, Covered In Fabric World News 0
T Die Cutter for Kevlar Fabric Manufacturing and Related Processes 10
D What Measurement Equipment is Recommended to Measure Long Pieces of Fabric? General Measurement Device and Calibration Topics 4
B Question about "When PPAP is required" - Tier 2 to GM - Fabric Density on Part APQP and PPAP 24
U Fabric Dying - Sample Technique - Sample size to check evenness of dyeing shade? Statistical Analysis Tools, Techniques and SPC 3
W Experience folding your QC Manual for ASME Section VIII, Division 1 into ISO 9000? Other ISO and International Standards and European Regulations 4

Similar threads

Top Bottom