Iam looking for a reference that will help us identify the atmospheric –Ergonomic conditions that are required to meet “ Class” of Lab.(precision mechanical measurements-Metrology/Calibration Lab)

I have worked in organizations that had huge metrology labs. There are variations in practice.

The most frequently used factors are:
Temperature: 20+/-1 deg C, Humidity:55% Max RH.

I understand there are more to this like:
Clean XXX ppm, vibration, lighting, acoustics, etc.
+ should have anti-room at the entrance.

If there are any web references, standards that point to this information please post.If you are currently following any and/or all of those factors in your lab to your specifications, please post ( if not proprietary).

Thanks,
Govind.
Note:I did a search in the forum database, Iam not sure if this topic was ever discussed.
 
Elsmar Forum Sponsor

Govind

Super Moderator
Staff member
Super Moderator
#3
Marc said:
'Bump'. Any takers on this one???
This is the closest, I was able to find from Google search:
http://www.nist.gov/public_affairs/amlbrochure.htm

This page identifies the environmental controls. Not ergonomic factors.
The environmental controls are too stringent.
I guess there should be other classification of Labs which allow wider tolerance on those control factors.

Regards,
Govind.
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#4
Govind,

Some of the various influences you mention may be site dependent. For example, vibration will have an influence at one location (e.g. California), but may not have nearly as much influence at another location.

For ergonomics, I suggest going to the ergonomics folks for that. The concept would be the same, only the application would be different.

The ppm count deals with clean room approaches. A well designed lab will often be able to achieve class 1000 without too much trouble. If you need something like a class 100 or tighter, then you may have to do some extra work.

Hope this helps.

Hershal
 
G

Graeme

#5
Here are some sources

I don't have any web resources instantly at hand ...

NCSL International does have a couple of recommended practices for calibration laboratory design:
  • RP-7, Laboratory Design
  • RP-14, Guide to selecting standards laboratory environments.
Purchase information is available at their web site, www.ncslinternational.org

There is also some information in The Metrology Handbook, chapter 16. It can be purchased from ASQ Quality Press, or for a higher price from your favorite bookseller.
 

Govind

Super Moderator
Staff member
Super Moderator
#6
Thanks to both Hershal and Graeme.Very useful feedback.

I will go ahead to explore the references you suggested. For the lighting, I remember that there should be a minumim of 400 lux/ test station.In my previous organization, we had an Anti-room and an Air curtain to prevent dust particles entering in. Also we had a small sliding window with a counter for short term transaction of Calibrated Gauges.This helped the temperature control.

Shoe covers were used with the good intention to prevent dust from the shoes. But mostly people forgot to remove and we use to spot those people in the cafeteria. :lol:

If you have any suggestions for building a precision measurement Lab, please let us know.

Regards,
Govind.
 
Last edited:

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#7
Actually, I do have an idea for you. Check with the dimensional lab at NRC (Canada). The room for gage blocks and the like, under Dr. Jim Pekelski.

They recirculate only 1/6 of their air, save energy, achieve better than class 100 without bunny suits, and have change of less than 0.5 degree C per hour worst case.

Tell Jim I said hi.

Hershal
 
Thread starter Similar threads Forum Replies Date
G ITAR Registration for Metrology Laboratory? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
L How to manage a Metrology Department or Calibration Laboratory General Measurement Device and Calibration Topics 3
V Temperature/humidity requirements in internal metrology laboratory General Measurement Device and Calibration Topics 3
C Building a Dimensional Metrology Laboratory to Inspect various Physical Measurements General Measurement Device and Calibration Topics 2
S Standard Equipment List for Stocking a Dimensional Metrology Laboratory General Measurement Device and Calibration Topics 8
D Is Temperature & Humidity monitoring in a Metrology / Test Laboratory required? General Measurement Device and Calibration Topics 28
R Tools used for measuring (New to metrology) General Measurement Device and Calibration Topics 9
C Teaching Metrology - Vernier vs Digital General Measurement Device and Calibration Topics 8
F Online courses - Metrology, standardization, calibration and ISO 17025 Training - Internal, External, Online and Distance Learning 3
F Electrical Metrology (Source/Generation and Measurement) General Measurement Device and Calibration Topics 1
K Dueling Metrology Labs - Two different labs telling you two different things General Measurement Device and Calibration Topics 5
G Looking for a Metrology Course preferably Dimensional and Electronics Training - Internal, External, Online and Distance Learning 5
B New to Metrology and told to complete Gage R&R with AIAG compliant software Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 16
G Metrology Laser Reliability General Measurement Device and Calibration Topics 2
bobdoering Available SPC and Metrology Consultants Paid Consulting, Training and Services 0
O Metrology Devices to Measure the Basic Dimensions of Plastic Specimens General Measurement Device and Calibration Topics 6
M MSA - Validation of Metrology Software such as PCDMIS Calibration and Metrology Software and Hardware 2
J How to measure effectiveness of the Metrology Process General Measurement Device and Calibration Topics 5
D Mad Metrology - A fun Metrology Game General Measurement Device and Calibration Topics 4
N Designing an attribute MSA for a new piece of metrology equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 11
I SPC Statistical Analysis in Metrology General Measurement Device and Calibration Topics 2
Hershal Metrology Careers - Potential good Paying Jobs in the Science of Measurement Field Career and Occupation Discussions 1
P I am looking for some Metrology Procedures Measurement Uncertainty (MU) 2
R Metrology Control Charts for Reference and Working Standards General Measurement Device and Calibration Topics 6
D Correction and Degree of Error - Metrology Definitions and Differences General Measurement Device and Calibration Topics 1
J Requirements for Hardness of Metrology Instrument Parts General Measurement Device and Calibration Topics 3
J Reputable Third Party Inspection (Metrology Lab) In USA? Cam Profile Measurement General Measurement Device and Calibration Topics 5
C Metrology and GD&T Employee Evaluation Test Manufacturing and Related Processes 12
V Has anyone used Empower (Calibration/Metrology software)? Calibration and Metrology Software and Hardware 2
Jim Wynne Great Moments In Metrology History--Underwear Caliper Coffee Break and Water Cooler Discussions 6
S Metrology Logistical Support General Measurement Device and Calibration Topics 1
P Anyone seen Metrology Improvements (P/T, GR&R) as Six Sigma Projects? Six Sigma 3
M Metrology Room Lighting Requirements General Measurement Device and Calibration Topics 7
Hershal A Primer on Metrology General Measurement Device and Calibration Topics 4
M Metrology Technician Certificate - Metrology Technician courses or certifications Professional Certifications and Degrees 3
Hershal New edition of the VIM has been released - International Vocabulary of Metrology General Measurement Device and Calibration Topics 4
M Metrology Related Professional Association Recommendations General Measurement Device and Calibration Topics 5
Q Material, Links & Information on some basic exercise on PC DMIS & CMM Metrology Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
U ISO/IEC Guide 99:2007 - International Vocabulary of Metrology (VIM). General Measurement Device and Calibration Topics 1
N International Symbols in Metrology and Calibration EU Medical Device Regulations 2
L Boeings volumetric metrology procedures D1-8005 Customer and Company Specific Requirements 1
Wayne Searching for Zero - The Definitive Dimensional Metrology Reference Book, Video, Blog and Web Site Reviews and Recommendations 7
L Basic Metrology Overview Slides Needed General Measurement Device and Calibration Topics 2
S Recreating a Metrology System - What will you recommend me to start with? General Measurement Device and Calibration Topics 10
F ISO 10012 and ISO 17025 - Doing research about metrology systems confirmation ISO 17025 related Discussions 7
G The Metrology Handbook - New book in part by 2 Elsmar Covers! Congrats! Book, Video, Blog and Web Site Reviews and Recommendations 0
B Calibration Management Software: MudCats Metrology Suite General Measurement Device and Calibration Topics 14
D Resolution, Discrimination, NDC and the Tower of Babel in Metrology General Measurement Device and Calibration Topics 5
Marc Definition Metrology Terms and Metrology Definitions - A very nice Glossary Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 5
C ISO 80601-2-56 Laboratory accuracy Other Medical Device Related Standards 7

Similar threads

Top Bottom