What is the name or number of the actual calibration procedure of a 0-1" micrometer? Where can I get it?
 
Elsmar Forum Sponsor
T

TeeTV

#3
Our company do customise calibration procedures for Micrometer, to fulfill ISO/IEC 17025 requirements. We were thinking of making them for online sales that you can use it immediately as your in-house procedure. (This type of calirbation is quite standard.) Will you be interested?


TeeTV :)
 

Govind

Super Moderator
Staff member
Super Moderator
#4
dbulak: Not meant to hijack your original thread, apologies.

JStain1,
We dont have that procedure 33K6-4-15-1. We custom developed internal procedure for our micrometers.
Just curious about one point; while measuring master between between the anvils,Is there any mention about "Number ratchet rotations"?
I was told by a senior metrologist from the old school as 2 clicks. About right?
Govind.
 
T

TeeTV

#5
Govind,

We do 2 soft rotations between thumb and the pointing finger when the anvils about touching the object. Remember, soft and slow clicks.


Best regards,
Tee TV
 
J

Jstain1

#6
Micrometers are not expensive "C" Clamps

Most established labs have a procedure in place. The more "robust" labs have access to GIDEP, where there's 1,000's of procedures, so I'm not really not in need of a procedure.

I'm sure I'll get differing opinions on this, but most of the Air force procedures are ISO17025 compliant, with the Issue of uncertainty being the biggest sticking point. I think that even that isn't a big issue.

The rule of thumb for tension is from 2 to 3 gentle turns. I've been through the Air force as well as NAVY and Mitutoyo training, and not suprising I was taught differently at each. :)

So, in general terms I would imagine there may be a market for procedures, but I think it would be a small niche.

Thermometers can get to be a bit involved, given the various standards that are available. I think, IMHO, that this may be very problematic.

Would you offer a "garantee" of sorts that the procedure would be compliant to some standard? if so, then your going to have to resolve this with the various certifying agencys, I.E. A2LA, NVLAP and LAB.
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#7
Just a few notes......

The military procedures (USAF T.O. 33K and USN 17-20 & 17-50 series) may not necessarily be compliant to ISO/IEC 17025 in all details, uncertainty being the main sticky point, but they are extensively validated, and so we do accept them for accreditation.

I believe our colleague organizations (A2LA, NVLAP, and even L-A-B and ACLASS) also accept the military procedures for the same reason. Incidentally, the other AB not mentioned is IAS, the oldest AB in the U.S.

Marc, forgive me for blowing our horn, just want to make sure we are not overlooked.

Hershal
 

Govind

Super Moderator
Staff member
Super Moderator
#8
Jstain,
I did enjoy your humor about"Micrometers are not expensive "C" Clamps" :D pertaining to number of ratchet rotations. Also thanks to both yourself and TeeTv for providing clarity on the number of turns. But both of you did not confirm if this practice is a "thumb rule" or written standard guidelines.

This will matter for digital micrometer calibration as the display start to move back and forth couple of microns depending on the torque through ratchet rotation.I dont see this issue with mechanical micrometers though.

Govind.
 
T

TeeTV

#9
Govind said:
Jstain,
I did enjoy your humor about"Micrometers are not expensive "C" Clamps" :D pertaining to number of ratchet rotations. Also thanks to both yourself and TeeTv for providing clarity on the number of turns. But both of you did not confirm if this practice is a "thumb rule" or written standard guidelines.

This will matter for digital micrometer calibration as the display start to move back and forth couple of microns depending on the torque through ratchet rotation.I dont see this issue with mechanical micrometers though.

Govind.

I haven't seen a international standard guildline on this before (what about others, have you seen one before?), the method I described was from Mitutoyo.


TeeTV
 

bobdoering

Stop X-bar/R Madness!!
Trusted Information Resource
#10
Re: Micometer calibration - Seeking actual calibration procedure of a 0-1" micrometer

Be sure to check the flatness of the anvils for valid calibration - either using gage balls or optical flats. It is one of the most overlooked aspects of micrometer calibration in in-house labs.
 
Thread starter Similar threads Forum Replies Date
C Micrometer and Vernier Caliper simple calibration process General Measurement Device and Calibration Topics 8
G From an ISO 17025 auditor perspective must micrometer calibration check anvil flatness? General Measurement Device and Calibration Topics 4
T A question on Calibration on In-line Laser Micrometer General Measurement Device and Calibration Topics 1
C One Inch Micrometer Calibration Method General Measurement Device and Calibration Topics 2
S Micrometer Calibration Using a CMM Calibration and Metrology Software and Hardware 12
M Calibration of a Micrometer used as a Standard General Measurement Device and Calibration Topics 5
W First Draft Micrometer Calibration Procedure ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 19
C Micrometer Parallelism and Flatness Calibration Requirements and Specs General Measurement Device and Calibration Topics 10
J Calibration of Micrometer Reading up to Three Decimals General Measurement Device and Calibration Topics 3
K Calibration Procedure for ID Micrometer Calibration & Extension Rods General Measurement Device and Calibration Topics 5
P Micrometer Calibration Methods Calibration Frequency (Interval) 11
K Micrometer Calibration - Resolution, Accuracy and Range General Measurement Device and Calibration Topics 16
P How to Calibrate Laser Micrometer when no Calibration Company in the Country? General Measurement Device and Calibration Topics 3
V Calibration Procedure for Thread Pitch Micrometer General Measurement Device and Calibration Topics 3
T In-House Calibration - Digital Micrometer found Out of Calibration General Measurement Device and Calibration Topics 10
C Pin Gauge Calibration using a Micrometer - Accuracy & Calibration Level Question General Measurement Device and Calibration Topics 10
J Calibration of Flute Micrometer General Measurement Device and Calibration Topics 3
apestate Thread micrometer questions, discussion regarding MU and calibration. General Measurement Device and Calibration Topics 3
apestate Micrometer Calibration Procedure - Please critique General Measurement Device and Calibration Topics 20
N Calibration of micrometer by gauge block General Measurement Device and Calibration Topics 5
N Confuse: Calibration using master equipment - Resolution - Micrometer General Measurement Device and Calibration Topics 7
S How often for micrometer calibration? General Measurement Device and Calibration Topics 10
A Outside micrometer anvils parallelism General Measurement Device and Calibration Topics 3
G 0.00005" tolerance on micrometer with 0.0005" resolution? General Measurement Device and Calibration Topics 4
A Depth micrometer rod adjustment? General Measurement Device and Calibration Topics 8
A Micrometer anvil chipped General Measurement Device and Calibration Topics 19
G How to compute or estimate a micrometer or caliper flatness using optical flat General Measurement Device and Calibration Topics 2
G Procedure Creation specifically for dimensional (caliper, micrometer, dti) Document Control Systems, Procedures, Forms and Templates 8
W Standard for Micrometer Anvil Flatness Oil and Gas Industry Standards and Regulations 5
G Uncertainty Budget Examples for Caliper, Micrometer and Dial Gauge Measurement Uncertainty (MU) 3
G Caliper or Micrometer's Dimensional Flatness General Measurement Device and Calibration Topics 5
G Manual Procedure with Uncertainty Budgeting for either a Caliper or Micrometer General Measurement Device and Calibration Topics 6
M Uncertainty Budget for a Micrometer Measurement Uncertainty (MU) 5
M Laser Micrometer Suggestion (if any) for diameters that are +/-.0001 inch General Measurement Device and Calibration Topics 1
M Micrometer Anvil Placement to measure the diameter of Steel Wires Measurement Uncertainty (MU) 6
W Micrometer Storage while not in use General Measurement Device and Calibration Topics 4
D Can anyone share a completed Gage R & R on a Micrometer? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
W Thread Micrometer Standard Tolerances General Measurement Device and Calibration Topics 4
5 Reasonable GRR result for a 0-1" micrometer with a total tolerance of .002" Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 16
S Laboratory to Calibrate Objective Micrometer needed General Measurement Device and Calibration Topics 4
D Procedure on How to Calibrate a External Micrometer General Measurement Device and Calibration Topics 1
J Air Micrometer Basics - What are the internal parts and how do they work Manufacturing and Related Processes 7
M Trigonometric Calculation (V-Anvil Micrometer). I'm not getting it. General Measurement Device and Calibration Topics 6
4 Federal Specification GGG-C-105C - Caliper, Micrometer, Gage, Depth Micrometer General Measurement Device and Calibration Topics 7
D Determining Instrument Purchase based on Accuracy (Micrometer) General Measurement Device and Calibration Topics 3
A Generic or Specific Gauge R&R on Standard Equipment (ie.: Micrometer) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
D What is the correct Working Standard for Digital Micrometer? General Measurement Device and Calibration Topics 21
D Laboratory where we can have parts checked for Thickness with a Laser Micrometer General Measurement Device and Calibration Topics 6
D Laser Micrometer for checking Thickness of Steel - Any Recommendations or Comments General Measurement Device and Calibration Topics 3
M How to Fix the Standard Bar of a Screw Thread Micrometer General Measurement Device and Calibration Topics 9

Similar threads

Top Bottom